RE: New User
Hi, and welcome.
Your upper pressure setting is substantially higher than where your pressure chart says you like to be. However, it’s only the one night you’re showing, and there might be a lot more variance.
Generally, clustered events, if they are otherwise sporadic AND of the “open airway” or “central” apnea (see the summary index at far left of all the charts), mean you are either dropping your chin and being aroused enough to hold your breath, or you are just forgetting to breathe, something that we see in some people using PAP.
You have clustered events, so you might do better if you were to try a soft foam cervical collar. It would help to reduce the arousals or real obstructions if they happen as well.
I believe you will be asked to provide an expanded view of one or more of those clusters showing at most a minute before those events. Would you mind terribly?
Contrite, and wiser.
RE: New User
Do you have an example of the clusters of events? Are they obstructive or Lear airway? With the BiOAP Pro at 22/17 (PS 5.0) pressure your events appear to be mostly CA. You seem a bit over-ventilated with a tidal volume of 740 mL which probably accounts for the CA events. We don’t know how you arrived at your current pressure and pressure support but it is relatively high although not unheard of. A closer examination of flow rate or your titration study might offer some insights. I assume you have not experimented with lower pressure which might be possible from what I’m seeing in this example.
RE: New User
Thanks so much. What kind of detail would you like me to provide and where does it reside in the data. I will check into the collar. I was also told that I have substantial periodic limb movement during the night. Not sure what all that means and the tech did not address it. I guess the arousals could come some from that also.
RE: New User
Thanks so much. Most of the events are clear airway. The pressure setting was designated by the VA tech. I had two in hospital sleep studies done and each time the results indicated the high pressure was needed to reduce the events substantially. The VA does not give you much info so you are pretty much on your own. That is why I went to this forum. I could not believe that something as critical as this would be a set it and forget it. I have not tampered with any of the pressure settings. I would not know where to begin.
RE: New User
O.K. changed the settings as recommended. I will repost with the new data. Thanks again.
RE: New User
Interesting that the vents are nearly all central when what we did was reduce pressure and pressure support which would more likely increase obstructive events and decrease centrals. I think the more important question is how did the lower pressure feel to you? I'm doubtful the setting changes is what caused the increase in CAI. You have slightly lower tidal volume and minute vent which, unless you're a giant is getting closer to normal. If you're okay with the change, i would continue for second night.
RE: New User
The lower pressure felt more comfortable from the beginning. I usually have to concentrate early on to get used to the high EPAP. I will let it ride and see what happens.
Thanks,