[News] RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
"Sam, how can I be 100% certain that my very own gently used and meticulously maintained PR DreamStation or System One is shedding particles that I will surely inhale deeply into my lungs because I can't block them all reliably with an AB filter?"

The above quote is from my recent post # 1,014 on page 102 of this long discussion. The reason why the certainty of exposure to particles is 100% is that Philips finally explained what is happening on July 8 when they gave "Supplemental clinical information for physicians and providers for specific CPAP, Bi-Level PAP, and mechanical ventilator devices".

Philips wrote, "1. Foam degradation: Despite a low complaint rate of 0.03% (i.e. 3 in 10,000) in 2020 [1], Philips has determined from user reports and lab testing that the PE-PUR foam may slowly degrade - through a process called hydrolysis - and produce particulates which may enter the device’s air pathway where they could be ingested or inhaled by the user of impacted Continuous  Positive Airway Pressure (CPAP), Bi-Level Positive Airway Pressure (Bi-Level PAP) and mechanical ventilator devices."

First some terminology, then some basic science to make Philips' admission above more user friendly. When Philips writes, "foam degradation" they are referring specifically to the issue of particles. The phrase they use for the off-gas problem is "chemical emissions". Hydrolysis is the ordinary term for any chemical reaction in which a molecule of water breaks one or more chemical bonds.

In order for any chemical reaction to occur, molecules must bump into each other. How are molecules of water able to contact your machine's foam? The water molecules are in the air, and we know them as humidity. If the relative humidity where you have your xPAP is above 0 (trust me, it is) then your foam is exposed to water molecules. Bumping occurs, and the foam degrades resulting in the infamous particles. Philips says so.

Philips wrote that the foam "may slowly degrade". They get away with writing such vague phrases because it is impossible to test every xPAP model under every combination of temperature, relative humidity and machine settings. They don't admit it, but the foam also may rapidly degrade.

What is useful about this Philips admission is that we have clues for assessing, in a relative way, our personal risk. Common sense tells us that:
1. The higher the room's humidity, the more water molecules, the more bumping, the more reactions, the more foam particles.
2. The more air passing through the device, the more water molecules pass through and touch the foam, the more chemical reactions, and you know the rest. My 950P produces a continuous 9.0 cm of H2O. I would have less exposure if the setting was lower, and more exposure if the setting was higher.

Temperature is another variable in your personal assessment of your relative risk. The hotter the environment, the faster the molecules move, the more bumping they do, etc. The rule of thumb is that a chemical reaction's rate doubles with every 10° C (about 18° F) increase of temperature. The relevant temperature is the temperature inside your xPAP's blower unit, which likely is warmer than the temperature inside the room. You may ask, "Doesn't the air flowing through the unit keep it cooled?" That's a great question. I don't have all the data.

Underneath your System One or DreamStation you'll probably find a sticker that says something like "12V - 5.0A" which means the machine operates at 12 volts and normally (assuming no electrical short) draws a maximum of 5 amps of electrical current. 5 amps of 12 volt electrical current is 5 X 12 = 60 watts of electrical power. Exactly the same energy use as an old school 60 watt incandescent light bulb. However, much of the xPAP's maximum power demand is from the humidifier. I have a watt meter, so I know that my 950P's power supply draws 0.27 watts. I don't use the humidifier. When the blower unit is turned off (including no display on the LCD), the unit's power use fluctuates around 2.23 watts. When the unit is delivering continuous 9.0 pressure to the mask the power consumption is about 3 watts when I exhale, 13 watts when I inhale, and 6 watts when I'm not breathing. On average I figure about 7.5 watts. Regardless of the electrical device, 3.412 BTUs (British Thermal Unit) of heat is produced per hour by one watt. One BTU is enough heat to raise the temperature of 1 pound of water by 1° F. The blower unit's case is plastic which is a great insulator and a lousy conductor of heat. Moreover, the enclosure is very small. The heat accumulating inside tends to stay there when the blower motor isn't running. The blower motor is the source of most of the heat. The second law of thermodynamics tell us that heat inside the xPAP  blower unit will always move from the warm blower to the cooler foam until both are at the same temperature.

I like to think that it doesn't get terribly hot inside my 950P, but I won't know until I insert a temperature probe. When the unit is first turned off the blower motor is not instantly cool. With no air moving the temperature inside will climb.

Back to the personal assessment, for most of us there is more exposure to particles in the summer than there is in the winter when homes typically are a few degrees cooler. Homes tend to be more humid in the summer also. The higher the setting the harder the blower motor works and the more heat it produces. There's a double whammy with the higher pressure setting causing both more heat and more water molecules touching the foam.

The final variable is time. The more hours of xPAP you get, the more exposure you get.

DreamStation users, your device does the same physical work as the older System One equivalent model, and DreamStation I assume has a richer feature set. So DreamStation likely draws comparable power and produces comparable heat, maybe a little bit more.

Back to our $20 science word, hydrolysis. PE-PUR foam has been in commercial use since about 1957. I've found proof that researchers knew as long ago as 1993 that hydrolysis causes PE-PUR foam to degrade. We don't know yet what Philips knew in 2009 when they selected PE-PUR foam. But we know for a fact that they could have known about PE-PUR foam degradation and arguably should have researched the material's known life cycle.
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
Great writeup Sam,

 It made me think of two things to do..  change my intake filter more often to avoid any extra labour (heat) by the blower , and figure a way to keep the unit as cool as possible.  Perhaps a laptop "base" with the built-in cooling fans ??
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
"They don't admit it, but the foam also may rapidly degrade."

and you know this...how?

You've done no testing. You have no access to test results. Your write up, as you say, is based on "a personal assessment". Well, it's good that you're being up front. in this case, you're pulling assessments out of thin air.

You claim that it's been known since 1993 that hydrolysis (humidity, in other words) will cause pre pur foam to rapidly degrade--and yet it's been used in millions upon millions of medical grade devices over the years. that's quite a charge!

You've "found proof". I doubt it. These machines are used throughout the world. Are you saying that all regulators world wide have been bought off to not bring to light this horrible health hazard used by millions for many decades? I mean, these things would be far worse than smoking a pack of cigarettes a day if what you say is true.

Pre pur foam has to pass medical grade specs before approval by all those world wide agencies. apparently it's been used in hospital devices for decades. And yet no one's blown the whistle over all those years? People certainly did for cigarettes, for asbestos, for DDT and a hundred other things.

No. Your claim begs credulity. I think that we need to look to evidence based research before passing judgement.

You certainly have every right to your personal assessment! Me too.
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
"You've "found proof". I doubt it."

It is hard to do damage control when the facts are all against you.

Be sure to read the dates in the footnotes.

https://www.iiconservation.org/archive/w...pub_id=139
Post Reply Post Reply
PHILIPS DEFECTIVE FOAM
(08-07-2021, 07:29 PM)SingingSam Wrote: ... researchers knew as long ago as 1993 that hydrolysis causes PE-PUR foam to degrade.
... could have known about PE-PUR foam degradation and arguably should have researched the material's known life cycle.

Excellent research, scientific evidence and deduction.  Investigations that are proceeding, as a result of the class-action lawsuits against Philips, will likely produce similar findings.  

At least one U.S. Senator is putting additional pressure on Philips to produce immediate results for users:
Quote:"Bridgeport man affected by Philips breathing machine recall says company is not doing enough to address crisis"
NEWS 12 New York
August 7, 2021

A Bridgeport man, one of thousands of Americans affected by a breathing machine recall, says the manufacturer is not doing enough to address the crisis it created.

Joe Furino says his health has been failing ever since Philips recalled its ventilation and sleep apnea devices back in June.
The FDA recalled the machine due to concerns it could release toxic chemicals into the lungs of patients.

Senator Richard Blumethal tells News 12 he plans to contact the manufacturer (Philips) and demand it takes immediate action.

< TV News Video >

Bridgeport man affected by Philips breathing machine recall says company is not doing enough to address crisis (news12.com)
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
(08-07-2021, 09:17 PM)hegel Wrote: "They don't admit it, but the foam also may rapidly degrade."

and you know this...how?

You've done no testing. You have no access to test results. Your write up, as  you say, is based on "a personal assessment". Well, it's good that you're being up front. in this case, you're pulling assessments out of thin air.

Well, it certainly degrades rapidly from exposure to ozone. See: https://www.youtube.com/watch?v=DbHcZUFRDqg

Aside from that, there have been a fair number of people who took apart their DreamStations to find that the foam had gotten oily/sticky/gooey. I posted a pic of one of them a looong ways back in this thread. So causes aside, there's at least multiple angles of proof that the foam *can* and *does* degrade within the lifetime of the machine, in some cases, somehow.
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
"So causes aside, there's at least multiple angles of proof that the foam *can* and *does* degrade within the lifetime of the machine, in some cases, somehow."

Actually it would be extremely hard to find a single recalled machine that can be proven to be free of foam degradation. Philips has warned that not seeing particles is not evidence that the degradation has not occurred.

It is far, far closer to the truth to say that the foam degrades in all cases than to say that it degrades in some cases.

As for the "somehow" we already know how. Hydrolysis. Philips said so.
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
Hegel seems to be challenging the science on the grounds that if it were true, the substance would have been banned. 

These machines are used throughout the world. Are you saying that all regulators world wide have been bought off to not bring to light this horrible health hazard used by millions for many decades? I mean, these things would be far worse than smoking a pack of cigarettes a day if what you say is true.”

I would like to point out the simplistic reasoning that bribery and corruption is the only motive for regulators to do or not to do something. What about ignorance, gullibility and all the other flaws that plague us humans. 

Moreover, context is important. We do not know what Philips provided the regulators in the way of test results or changes or special operating conditions that might have tipped the balance of their decisions. 

Sam said that “the foam also may rapidly degrade" in response to Philips’ PR-speak that the foam "may slowly degrade". You asked: “And you know that…how?” I do not understand why you think proof is needed for saying something MAY happen. 
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
(08-08-2021, 09:56 AM)BobbieM Wrote: Hegel seems to be challenging the science on the grounds that if it were true, the substance would have been banned. 

No he's not.
Post Reply Post Reply
RE: RECALL THREAD-- IMPORTANT PHILIPS DREAMSTATION & SYSTEM ONE USERS
It's always good to see a healthy discussion of issues like this one, with different views.  

I, for one, am uncomfortable taking risks that I am unable to quantify.  Particularly where my health is concerned.  I know with certainty that I need a CPAP and that doing without is bad for my health.  It is even potentially fatal.  I know that Philips has recalled these machines and that there is an unquantified but very real risk that my continued use of the machine could have serious health consequences.  

There may, be ways to mitigate the risk of continued use, but they are also unquantifiable. As I have already stated, I have opted for a new (Resmed) machine.  My insurance company didn't like the looks of the unquantified risks either.  They chose - quickly - to pay their portion of the cost.  I know that I am blessed to have had that option.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  NOTICE: Philips Discontinues Numerous Devices and Halts Dreamstation Sales in the U.S. WakeUpTime 48 3,846 Yesterday, 04:49 PM
Last Post: WakeUpTime
  WW4B Therapy thread WW4B 49 4,073 Yesterday, 04:31 PM
Last Post: WW4B
  IanD - Therapy Thread IanD 28 2,268 03-26-2024, 08:18 PM
Last Post: IanD
  Foam Removal - Philips System One peewee44 47 10,619 03-26-2024, 05:46 AM
Last Post: jojopuppyfish
  NOTICE: Philips Respironics Recall Replacement Tally RayBee 183 29,064 03-24-2024, 06:30 PM
Last Post: Phaleronic
  Marcin's Therapy Thread marcinpiskala 20 967 03-23-2024, 04:38 PM
Last Post: Sleeprider
Information Alaxo Hybrid Stent with CPAP - Therapy Thread G. Szabo 29 1,213 03-22-2024, 10:51 PM
Last Post: G. Szabo


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.