P10 Vent Rate
Hi: I recently changed to a Resmed Airfit P10 and I have adjusted very well to a nasal pillow mask from my F10 FFM earlier. My unintentional leaks are less than 2Lpm all night. My AHI is below 0.5 every day and my RERAs have dropped to 0.00 on most days.

My question is this:
My start pressure is 7.5cm H20. At this pressure my P10 mask should vent at a rate of nearly 28Lpm. My sleepyhead graph shows that it only vents at 19L-20L per minute. P10 manual says that its designed vent rate at 4cm h2o is 20Lpm. And some rebreathing may occur at low pressures. So should I be concerned that my mask is venting even less than 20Lpm?

Should I try and increase the vent rate by any creative tricks. I have tried brushing the vents vigorously but it didn't move the vent rate on SH.

The reason I am concerned about this is that even though I may wake up with an AHI of 0.0, I feel like I am not feeling refreshed in the morning.

Thanks.

Started APAP 4-20, Closed range to 7.5-14, then straight 8.0 w/ Aflex 3
RDI always below 1. But sleep much much better at straight pressure.
Started on F10, Tried Quattro Air successfully. Finally settled on P10.
02-04-2015 07:10 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,400
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #2
RE: P10 Vent Rate
There is no need to do anything to modify the mask. You have a slightly lower vent rate, but if you were rebreathing excess CO2, you would automatically start hyperventilating.

Download the Sleepyhead software http://sleepfiles.com/SH/index.html?
Take a look at your detailed daily sleep data for clues on why you're not refreshed. You can have zero AHI and a high RERA or flow limitations which are very disruptive to good sleep. If you need help interpreting this, you can post the data here and you'll get help. The DS560 does record these events even though they do not show up on your machine's screen.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
02-04-2015 08:44 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,400
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #3
RE: P10 Vent Rate
Regarding vent rates, lets assume a tidal volume of 600 mL per breath and 20 breaths per minute. That would be 12 Liters of air per minute. With a design vent rate of twice that volume, you are not rebreathing any appreciable volume of expired air. I guarantee you do not have anywhere near that tidal volume or respiratory rate during sleep, but you'll have to check your Sleepyhead numbers to know.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
02-04-2015 08:54 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
AshSF Offline

Advisory Members

Posts: 356
Joined: Oct 2014

Machine: PRS1 D560TS
Mask Type: Nasal pillows
Mask Make & Model: ResMed P10, Bkup: Airfit F10 & Quattro Air
Humidifier: PRS1 60 series with Heated Tube
CPAP Pressure: 8 - 8 w/ Aflex 3
CPAP Software: SleepyHead

Other Comments: Using APAP mode with fixed pressure

Sex: Male
Location:

Post: #4
RE: P10 Vent Rate
I do check my RERAs using my PRS1 560 machine on SH 0.9.8.1. I am posting my last night's report and the summary report since I started using P10.

My RERAs are 0.0 on every night except one night.
My pressure doesn't spike too much. Looks like PRS1 just keep doing the hunt and peck all night.
My Flow Limitations have mostly disappeared since going to P10.
My VS and VS2 have greatly reduced.
Most nights, I have 1 or 2 events total.
I am also sleeping more than 7 hrs each night continuously.
So I think I have optimized most of the variables here. So why the feeling not refreshed. Earlier on my F10, I used to have total leaks in 40+Lpm range and my AHI was < 1.5, RERAs were < 1.5 but I was much more fresh in the mornings.

Any insights into what else may be going on. All I can think of is the total leak rate being close to minimum designed for the mask. Should I be looking at anything else (TV, MV, RR?).

Heres my summary report since I started P10
[attachment=1259]

Here is last night graphs
[attachment=1260]

Just saw Sleeprider's comment. My stats for P10 days for
TV: W-Avg: 515ml
MV: W-Avg: 6.9Lpm

Started APAP 4-20, Closed range to 7.5-14, then straight 8.0 w/ Aflex 3
RDI always below 1. But sleep much much better at straight pressure.
Started on F10, Tried Quattro Air successfully. Finally settled on P10.
02-04-2015 09:21 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,400
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #5
RE: P10 Vent Rate
Data looks excellent...much better than mine. You could chase out the snores with another 0.5 cm pressure but I don't think that your answer lies in machine tweaks or even AHI. You seem to have done as much as possible to clean up sleep apnea and related arousals.

Have you recently had a full wellness exam and done comprehensive blood work including male hormone levels, vitamin D? I got kind of a surprise last summer in that regard, and it explained a lot. Lots of other physical and mental stress can impose on our sense of well-being and energy levels.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
02-04-2015 09:43 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
AshSF Offline

Advisory Members

Posts: 356
Joined: Oct 2014

Machine: PRS1 D560TS
Mask Type: Nasal pillows
Mask Make & Model: ResMed P10, Bkup: Airfit F10 & Quattro Air
Humidifier: PRS1 60 series with Heated Tube
CPAP Pressure: 8 - 8 w/ Aflex 3
CPAP Software: SleepyHead

Other Comments: Using APAP mode with fixed pressure

Sex: Male
Location:

Post: #6
RE: P10 Vent Rate
I got my blood work done pre-cpap. Everything was in great shape including Vit D, B12, All hormones and all panels.

Looking at the data prior to P10, When I was on F10 FFM, my RDI was averaging 1.1 and I was feeling great all day.
On the P10, my RDI is averaging 0.24 but I am not feeling so great.

After looking at all data points, the leak rate is the only one that is very different. Total Leak rate on F10 used to average 37Lpm. On the P10, its 20.3Lpm.

Started APAP 4-20, Closed range to 7.5-14, then straight 8.0 w/ Aflex 3
RDI always below 1. But sleep much much better at straight pressure.
Started on F10, Tried Quattro Air successfully. Finally settled on P10.
02-05-2015 11:34 AM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,857
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #7
RE: P10 Vent Rate
(02-04-2015 08:54 PM)Sleeprider Wrote:  Regarding vent rates, lets assume a tidal volume of 600 mL per breath and 20 breaths per minute. That would be 12 Liters of air per minute. With a design vent rate of twice that volume, you are not rebreathing any appreciable volume of expired air. I guarantee you do not have anywhere near that tidal volume or respiratory rate during sleep, but you'll have to check your Sleepyhead numbers to know.

That's called minute volume. I run 15 BPM and a TV of 500ml.
My minute volume is 7.5 LPM.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
02-05-2015 11:45 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [Equipment] low heart rate BrownFox292 4 246 11-24-2016 03:34 AM
Last Post: holden4th
  Leak rate abrun7 8 494 11-22-2016 09:02 PM
Last Post: Sleepster
Question CA's, Mask type, and leaks/vent Rcgop 10 379 11-15-2016 03:02 PM
Last Post: robysue
  Flow Rate Interpretation nadprok 2 220 11-13-2016 06:08 PM
Last Post: nadprok
  Resmed AirFit P10 Acceptable Leak Rate jimchi 1 369 10-30-2016 09:30 PM
Last Post: Sleeprider
  95% Rate Has Risen Significantly holden4th 13 676 10-25-2016 04:04 AM
Last Post: holden4th
  1st night with P10 leak rate? christinaleigh 10 613 10-24-2016 06:47 PM
Last Post: christinaleigh

Forum Jump:

Who's Online (Complete List)