(02-28-2020, 11:43 AM)Geer1 Wrote: Lets see some zoomed in examples, when doing so post 2 min durations and show the flow rate chart and mask pressure chart.
Around 1:05 and 3:05 caught my eye. These seem to indicate reduced breathing leading up to arousals. If you think there is anything else that looks odd/interesting post an example.
One thing that caught my eye was that your inspiration time was only 1 second and your tidal volume is on low side. What is your Timin and Timax set to? Do you ever feel like the machine is trying to supply or reduce pressure to early or late?
02-28-2020, 01:31 PM (This post was last modified: 02-28-2020, 01:33 PM by Geer1.)
RE: Possible UARS Sufferer?
I wondered about that... Increase Timax back to 2.
The machine runs off your spontaneous breathing effort. All you have done is force the machine to stop supplying pressure prematurely which is encouraging you to exhale sooner and this is also causing the low tidal volume.
I havent seen anything to indicate you need to be using Timax. What was probably bothering you and what you probably need to adjust is your cycle sensitivity. A higher cycle sensitivity will reduce the pressure sooner but still relies on your spontaneous breathing rather than trying to force your breathing to end sooner using ticontrol.
Edit: Trigger sensitivity slowriter mentions is the opposite and controls how soon the machine starts to supply pressure. You can try adjusting it if you are having issues with that as well.
(02-28-2020, 01:31 PM)Geer1 Wrote: I wondered about that... Increase Timax back to 2.
The machine runs off your spontaneous breathing effort. All you have done is force the machine to stop supplying pressure prematurely which is encouraging you to exhale sooner and this is also causing the low tidal volume.
I havent seen anything to indicate you need to be using Timax. What was probably bothering you and what you probably need to adjust is your cycle sensitivity. A higher cycle sensitivity will reduce the pressure sooner but still relies on your spontaneous breathing rather than trying to force your breathing to end sooner using ticontrol.
Edit: Trigger sensitivity slowriter mentions is the opposite and controls how soon the machine starts to supply pressure. You can try adjusting it if you are having issues with that as well.
Good Stuff. I'll change timax back to 2 and tune in my cycle. I'll report back in a week or so.
You tidal volume and minute vent do look a bit low, I am 5’ 9” and my tidal volume is similar to yours of 320 which is low but my respiration rate is higher giving a minute vent of 6l which is normal for my size. Have you used an SPO2 reader to check O2 levels mine are within normal range
(03-04-2020, 07:18 AM)jaswilliams Wrote: You tidal volume and minute vent do look a bit low, I am 5’ 9” and my tidal volume is similar to yours of 320 which is low but my respiration rate is higher giving a minute vent of 6l which is normal for my size. Have you used an SPO2 reader to check O2 levels mine are within normal range
I use a beddr sleep tracker and my SPO2% is is in the mid 90's.