Pressure, Flow, and Leaks
Author Message
vsheline Offline

Advisory Members

Posts: 1,907
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #11
RE: Pressure, Flow, and Leaks
(05-25-2016 12:08 AM)chill Wrote:  I've seen a lot of posts that say these CPAP machines increase pressure in response to leaks. That does not make sense to the part of my brain that remembers a bit of my university physics. If there is a leak pressure should drop, the same as punctured tire. The machine would increase flow (volume of air) to compensate for the leak and maintain therapy pressure (up to the point where it can't provide more flow which I think is the "large leak" point).

Right, except I think the threshold for "Large Leak" is usually much lower than where the machine can't provide more flow.

(05-25-2016 04:05 AM)FrankNichols Wrote:  Flow rate is whatever the flow rate needs (up to the capacity of the turbine) to be to maintain the pressure.

That is my understanding.

The pressure which is being maintained is the target Mask Pressure. The blower pressure is maintained at a higher pressure than the target Mask Pressure, by the amount of pressure loss across the hose.

The pressure dropped crossed the hose is a function of the airflow rate (which is measured by the blower) and of the tube type (diameter and length). This is why the correct tube type should be entered into the machine's settings by the equipment provider or by the user.

If we use the fat tube (not slimline) and if our Mask Pressure is fairly low, and if unintended Leak is fairly low, I think not much error would be caused by the tube type being incorrectly set.

But if we use the narrow tube, and if unintended Leak is very high, or if our Pressure is very high (leading to high intended leak through the mask vent holes), then the Mask Pressure may be surprisingly much lower than what the machine thinks/reports it is, if the tube type is incorrectly set.

(05-25-2016 04:05 AM)FrankNichols Wrote:  I have read (meaning I don't know what they do) that machines increase pressure (flow) in response to leaks ...

Usually, people who think "machines increase pressure (flow) in response to leaks" don't understand the difference between pressure and flow.

Pressure is measured in units of force per unit area, for example pounds per square inch.

Flow is a rate and is measured in units of volume per unit time, for example Liters per minute).

(05-25-2016 05:00 AM)justMongo Wrote:  There are two definitions of flow that one must consider. There is the absolute flow from the generator. And, there is the flow in/out of the person's airway.

Right. The "Flow" waveform in ResScan and SleepyHead represents the estimated rate of flow in our airway (lungs).

(05-25-2016 05:00 AM)justMongo Wrote:  There is apparently a middle ground where the leak causes the machine to be unable to properly detect the flow to/from the person. Thinking the flow has diminished, the machine raises pressure.

I have noticed in my Flow waveforms that during periods of high Leak the Flow plotted by the machine is usually much lower than when there is not high leak. I think the Flow estimated during high leak is surely much lower than reality.

So if high leakage made it seem like the Flow has dropped more than 40% or 50%, perhaps the machine could misinterpret the simple high leak as being, instead, an hypopnea accompanied by high Leak.

And you are saying, in other words, that during times of high leakage perhaps the machine may incorrectly think an hypopnea is occurring and, in response, may eventually raise the Mask Pressure? Seems like this may be possible, but I think neither ResMed or Philips Respironics has confirmed their machines could respond that way, so I'm not sure.

And if an ASV or other ventilator class of machine is used, during times of high leakage perhaps Pressure Support may be raised immediately in an attempt to end the apparent hypopnea? Again, it seems like this may be possible, but I'm not sure.

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 05-29-2016 02:09 AM by vsheline.)
05-29-2016 12:32 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
SuperSleeper Offline

Administrators

Posts: 9,961
Joined: Feb 2012

Machine: PR System One REMstar Auto (DS560)
Mask Type: Nasal pillows
Mask Make & Model: ResMed Mirage Swift II
Humidifier: none
CPAP Pressure: 12.5 - 18.5 cmH20 (auto range)
CPAP Software: SleepyHead

Other Comments: Have diabetes Type II

Sex: Male
Location: Illinois, USA

Post: #12
RE: Pressure, Flow, and Leaks
chill - would you mind if I moved this thread from the Advisory Members forum into the Main Forum?

SuperSleeper
Apnea Board Administrator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.

05-29-2016 10:43 AM
Find all posts by this user Post Reply Quote this message in a reply
chill Online

Advisory Members

Posts: 850
Joined: Mar 2016

Machine: DreamStation X500
Mask Type: Full face mask
Mask Make & Model: Walk-in Mask Closet
Humidifier: DreamStation
CPAP Pressure: 13-16
CPAP Software: SleepyHead

Other Comments: CPAP True Believer

Sex: Male
Location: Vancouver, Canada

Post: #13
RE: Pressure, Flow, and Leaks
I have no objections, there is a lot of good information here -- more than I expected! I posted here originally as my focus was on "I want to tell people the correct thing".
05-29-2016 12:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Source of Resmed S9 EPR reverse flow whistle! neal 5 224 12-01-2016 07:22 PM
Last Post: klv329
  Is it worth dropping pressure to get over leaks rich2905 19 1,254 11-16-2016 02:38 PM
Last Post: PaytonA
Question CA's, Mask type, and leaks/vent Rcgop 10 361 11-15-2016 03:02 PM
Last Post: robysue
  Flow Rate Interpretation nadprok 2 216 11-13-2016 06:08 PM
Last Post: nadprok
  Which disturbs sleep most? High Pressure or Changing Pressure Rcgop 7 1,239 11-10-2016 07:37 PM
Last Post: Sleep2Snore
  Leaks (Too high?) sleepytimegal 9 416 11-06-2016 09:58 PM
Last Post: sleepytimegal
  Move alot and getting mask leaks HELP christinaleigh 10 514 11-06-2016 03:15 AM
Last Post: holden4th

Forum Jump:

Who's Online (Complete List)