[Pressure] High pressure after bathroom visit
#11
RE: High pressure after bathroom visit
(02-17-2020, 08:20 AM)Sleeprider Wrote: Roamer, you have not included the flow limits chart, but we know that that is what is driving up your pressure.  For example at 01:15 we can see the flow rate pinching in as flow limitation increases.  Your pressure settings are on the default 4-20 and you're using exhale pressure relief (EPR) at 2.  We can help you further reduce flow limitation by increasing minimum pressure and EPR.  You need to change your settings to 7-20 and turn on EPR Full Time at 3.  This will give you 3-cm of pressure relief on exhale which will be both more comfortable and boost your respiratory volume a bit, it will also reduce flow limitation so your machine's pressure will be more stable.

I think you posted this before my last post with the FLow Limit chart. 

I am getting to different suggestions as to the EPR setting of 2 or 3. Can someone clarify? I saw a doctor online say that he sets everyone for 3.
Post Reply Post Reply
#12
RE: High pressure after bathroom visit
I have no issue with EPR=3, That was going to be the next change I was going to suggest. The 2 was to correct the mis-setting expecting to go to 3 after 1 night.
Post Reply Post Reply
#13
RE: High pressure after bathroom visit
I am going to change one thing at a time. Changed EPR from 2 to 3. We will see how a few nights of that goes before changing limits. Thanks to all for your help. Smile
Post Reply Post Reply
#14
RE: High pressure after bathroom visit
A purpose of setting min pressure to a min of 7 is to give the EPR of 3 room to work. Your actual min pressure will be 4
This is a paired change
Post Reply Post Reply
#15
RE: High pressure after bathroom visit
Roamer321,

If you want to start at 4 with an EPR of 3, you need to set your Minimum to 7.
EPR of any setting (1, 2, or 3) cannot subtract from 4, so EPR is not doing anything yet.
Set the machine to EPR 3 full time and Min pressure to 7.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#16
RE: High pressure after bathroom visit
(02-17-2020, 11:57 PM)bonjour Wrote: A purpose of setting min pressure to a min of 7 is to give the EPR of 3 room to work. Your actual min pressure will be 4
This is a paired change

OK thanks. I get it now. I will change the min pressure tonight.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  High flow limit, erratic breathing and resp rates jclarke131000 13 210 9 minutes ago
Last Post: OpalRose
  [Pressure] High pressure Tomarovsch 1 106 04-12-2024, 04:56 PM
Last Post: PeaceLoveAndPizza
  Do you stop your Airsense machine when you go to the bathroom during nighttime? dharma2255 10 302 04-12-2024, 10:23 AM
Last Post: dharma2255
  Extremely high CA readings on the ResMed Airsense 11 Soleil211 1 120 04-08-2024, 11:37 AM
Last Post: Sleeprider
Question [Diagnosis] First post-CPAP visit to Sleep Doc - What to talk about? FlyAU98 7 314 03-30-2024, 05:58 PM
Last Post: FlyAU98
  Phillips Dreamstation 2 High Disconnects mrsleepyhead2001 2 173 03-22-2024, 04:20 PM
Last Post: mrsleepyhead2001
  Recently started CPAP -- very high proportion of Clear Airway Events matt2 0 168 03-17-2024, 02:48 PM
Last Post: matt2


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.