[Pressure] Pressure Settings Can not be Controled
Author Message
schipper06811 Offline

New Members

Posts: 2
Joined: Jul 2013

Machine: unsure
Mask Type: Nasal pillows
Mask Make & Model: unsure
Humidifier: unsure
CPAP Pressure: unsure
CPAP Software: Not using software

Other Comments:

Sex: Undisclosed
Location:

Post: #1
Angry Pressure Settings Can not be Controled
I think I may have a Problem. I have a Resmed S9 Autoset.
Set S9 on Cpap Mode with pressure setting of 11. avg pressure =>14
Set S9 on Auto Mode with Pressure 8 / 12 avg pressure=>14
Set S9 on Auto Mode With Pressure 15/18 avg pressure= <15
I have been using CPAP for 12 years and just received the S9 5 months ago. It is nice to finally see the numbers.

My problem is if a min, or max, setting is dialed in should not the avg. pressure be within those min. or max. setting????

I am trying to dial in a good pressure setting because leak rates are high 28, AHI are in the 16's,etc., Centrals are all above 8's.
Seems like I can not change pressure as all avg pressures are in the 13's or 14's.

You name it I have tried it, for example I am taping my mouth now for 1 week to try and reduce leaks. No changes. I have tried all sequence of pressure settings in both auto and cpap mode, changed masks, set EPR, Humidity, hoses, two types, unplugged and plugged machine back in, removed card for 1 week, put back in etc.etc.etc..

Am I correct to think that the avg pressure reading should be less than the highest setting, and more than the minimum setting, and that when set on cpap mode the avg pressure should be about the same as the cpap setting?????? Help. THE DME or The Docter aren't.
12-01-2013 04:54 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,430
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #2
RE: Pressure Settings Can not be Controled
Hi schipper06811,
WELCOME! to the forum!
Hang in there for suggestions and best of luck to you.

trish6hundred
12-01-2013 05:21 PM
Find all posts by this user Post Reply Quote this message in a reply
me50 Offline

Advisory Members

Posts: 2,559
Joined: Aug 2013

Machine: resmed S9 VPAP Auto
Mask Type: Nasal mask
Mask Make & Model: Wisp and Silicone Head gear
Humidifier: resmed H5i
CPAP Pressure: 18/8
CPAP Software: SleepyHead

Other Comments:

Sex: Undisclosed
Location: USA

Post: #3
RE: Pressure Settings Can not be Controled
(12-01-2013 04:54 PM)schipper06811 Wrote:  I think I may have a Problem. I have a Resmed S9 Autoset.
Set S9 on Cpap Mode with pressure setting of 11. avg pressure =>14
Set S9 on Auto Mode with Pressure 8 / 12 avg pressure=>14
Set S9 on Auto Mode With Pressure 15/18 avg pressure= <15
I have been using CPAP for 12 years and just received the S9 5 months ago. It is nice to finally see the numbers.

My problem is if a min, or max, setting is dialed in should not the avg. pressure be within those min. or max. setting????

I am trying to dial in a good pressure setting because leak rates are high 28, AHI are in the 16's,etc., Centrals are all above 8's.
Seems like I can not change pressure as all avg pressures are in the 13's or 14's.

You name it I have tried it, for example I am taping my mouth now for 1 week to try and reduce leaks. No changes. I have tried all sequence of pressure settings in both auto and cpap mode, changed masks, set EPR, Humidity, hoses, two types, unplugged and plugged machine back in, removed card for 1 week, put back in etc.etc.etc..

Am I correct to think that the avg pressure reading should be less than the highest setting, and more than the minimum setting, and that when set on cpap mode the avg pressure should be about the same as the cpap setting?????? Help. THE DME or The Docter aren't.

Where are you getting this data from? On CPAP mode, the pressure is a set pressure, for example, 12 and that pressure is constant throughout the time you are using the machine.

When in AUTO, there is a pressure range and your pressure should fall within that range depending on what is happening and what is needed while you are asleep.

If you aren't using software, I would encourage you to download it and take a look at your data that way.
12-01-2013 05:32 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PollCat Offline
Apnea Board Polling Administrator
Monitors

Posts: 1,576
Joined: Feb 2012

Machine: ResMed S9 Autoset
Mask Type: Nasal pillows
Mask Make & Model: ResMed AirFit P10
Humidifier: H5i Cleanable with ClimateLine
CPAP Pressure: 10 - 18
CPAP Software: SleepyHead

Other Comments: CPAP-ing since 1995; Type 2 Diabetes; Chronic Asthma; Gout

Sex: Male
Location: Westfield, IN

Post: #4
RE: Pressure Settings Can not be Controled
I agree with me50. You need to get and use either the SleepyHead or ResScan monitoring software. Also, have you obtained the Clinician's manual for your machine? What was your prescribed pressure?


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA.
INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-01-2013 11:29 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,991
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #5
RE: Pressure Settings Can not be Controled
It sounds like your machine is somehow stuck in CPAP mode at a pressure of 14. Can you tell us what's on the on-screen menu when you're breathing? If what I'm thinking is right it should just report 14 during both inhales and exhales.

It may be that your equipment provider left directions on the SD card to set your machine's pressure, so no matter what changes you make, as soon as the machine turns on it reads the directions and sets the pressure to 14, as prescribed by the doctor.

I know that PRS1 machines can do this. Not sure about ResMeds.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-01-2013 11:45 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
jgjones1972 Offline

Preferred Members-2

Posts: 516
Joined: Nov 2012

Machine: ResMed S9 AutoSet
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro FX
Humidifier: ResMed H5i
CPAP Pressure: 18 - 20
CPAP Software: ResScan SleepyHead

Other Comments: CMS50I PulseOx

Sex: Male
Location: Ohio - USA

Post: #6
RE: Pressure Settings Can not be Controled
Are you sure you are just looking at the previous night's average? If you are getting the numbers from the screen you need to make sure it is set to 1 day or you will be getting an average over multiple days. Mine defaults to a 30 day average and I always have to dial it back to 1 day.
12-02-2013 12:05 AM
Find all posts by this user Post Reply Quote this message in a reply
tkeating4 Offline

Members

Posts: 7
Joined: Nov 2013

Machine: Resmed S9 Autoset EPR
Mask Type: Nasal mask
Mask Make & Model: Mirage Activa LT
Humidifier: H5i w/ClimateLine
CPAP Pressure: 10-15
CPAP Software: ResScan

Other Comments:

Sex: Male
Location: Canada

Post: #7
RE: Pressure Settings Can not be Controled
(12-02-2013 12:05 AM)jgjones1972 Wrote:  Are you sure you are just looking at the previous night's average? If you are getting the numbers from the screen you need to make sure it is set to 1 day or you will be getting an average over multiple days. Mine defaults to a 30 day average and I always have to dial it back to 1 day.

I would bet this is the issue. By default it does show the 30 day avg. You probably know this but... you can hold the "I" and Double Check Marks buttons down when looking at the Sleep Report to get into an additional reporting mode where you can change it to the 1 day report.
12-02-2013 08:54 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
schipper06811 Offline

New Members

Posts: 2
Joined: Jul 2013

Machine: unsure
Mask Type: Nasal pillows
Mask Make & Model: unsure
Humidifier: unsure
CPAP Pressure: unsure
CPAP Software: Not using software

Other Comments:

Sex: Undisclosed
Location:

Post: #8
RE: Pressure Settings Can not be Controled
(12-02-2013 08:54 AM)tkeating4 Wrote:  
(12-02-2013 12:05 AM)jgjones1972 Wrote:  Are you sure you are just looking at the previous night's average? If you are getting the numbers from the screen you need to make sure it is set to 1 day or you will be getting an average over multiple days. Mine defaults to a 30 day average and I always have to dial it back to 1 day.

I would bet this is the issue. By default it does show the 30 day avg. You probably know this but... you can hold the "I" and Double Check Marks buttons down when looking at the Sleep Report to get into an additional reporting mode where you can change it to the 1 day report.

Just checked to make sure I set it for one day. Yeak, it was set for 30 days. I suppose that maybe it goes to a default when unplugged.
Thank you for the check list. Ihave all the suggestions listed above in the post covered (using ResMed software).

Thanks for the double check reminder, now to see if I can start bringing some numbers down. Time to see what the next issue will be.
Thanks again.
12-02-2013 12:54 PM
Find all posts by this user Post Reply Quote this message in a reply
robysue Offline
Wiki Editor
Advisory Members

Posts: 1,226
Joined: Oct 2013

Machine: PR Dreamstation BiPAP Auto
Mask Type: Nasal pillows
Mask Make & Model: Swift FX
Humidifier: PR Dreamstation humidfier
CPAP Pressure: min EPAP = 4; max IPAP = 9;
CPAP Software: SleepyHead EncoreBasic EncorePro

Other Comments: Papping since September 2010

Sex: Female
Location: Buffalo, NY

Post: #9
RE: Pressure Settings Can not be Controled
(12-01-2013 04:54 PM)schipper06811 Wrote:  I am trying to dial in a good pressure setting because leak rates are high 28
Where is that leak number coming from? The S9's LCD? Is it the 95% leak rate?

Does Mr. Red Frowny show up or does Mr. Green Smiley show up in the short version of the sleep quality report?
12-04-2013 12:33 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Newbie - need advice on settings and machine tmc712 11 160 Yesterday 09:54 PM
Last Post: Sleeprider
  Dream Station Ramp and Pressure Settings knowoneuno 15 319 12-02-2016 06:50 PM
Last Post: Sleeprider
  Dream Station Settings Autumnseer 15 488 11-26-2016 07:01 AM
Last Post: Weird Tolkienish Figure
  Who generally is in charge of your CPAP settings? Mezzrow 9 463 11-12-2016 01:47 AM
Last Post: blueelement2016
  Which disturbs sleep most? High Pressure or Changing Pressure Rcgop 7 1,249 11-10-2016 07:37 PM
Last Post: Sleep2Snore
  do I have the right machine (and settings) JaminV 4 369 10-31-2016 10:05 PM
Last Post: trish6hundred
  Dreamstation heated tube/humidifier initial settings. NorthernGuy 9 706 10-28-2016 09:21 AM
Last Post: Perchas

Forum Jump:

Who's Online (Complete List)