ResMed S9 Humidifer 5i Error Message
#1
Hi,

I encountered an error message : "Humidifier Fault, Replace Humidifier" on my ResMed S9 Autoset unit. May I know if anyone has encountered the same, what was the cause, and what was the fix done?

Because of the said error, I am too curious, that I checked the external attributes of the humidfier. I noticed that the contact point of the Humidifier has 4 holes, BUT, only 3 holes has copper into them, while, the lower right hole do not. May I know if this is the same among the users of the same unit? Is this the way it should be?





Thank you for your help.

atta_gurrrrl
Post Reply Post Reply
#2
(07-07-2015, 12:42 AM)atta_gurrrrl Wrote: Hi,

I encountered an error message : "Humidifier Fault, Replace Humidifier" on my ResMed S9 Autoset unit. May I know if anyone has encountered the same, what was the cause, and what was the fix done?

Because of the said error, I am too curious, that I checked the external attributes of the humidfier. I noticed that the contact point of the Humidifier has 4 holes, BUT, only 3 holes has copper into them, while, the lower right hole do not. May I know if this is the same among the users of the same unit? Is this the way it should be?





Thank you for your help.

atta_gurrrrl

Troubleshooting: Displays error message: Humidifier fault, replace humidifier
http://www.resmed.com/us/dam/documents/p...er_eng.pdf

Post Reply Post Reply
#3
Hi atta_gurrrrl,
WELCOME! to the forum.!
Hang in there for more answers to your question about your humidifier.
Much success to you with your CPAP therapy, good luck to you with getting your problem with your humidifier straightened out.
trish6hundred
Post Reply Post Reply


#4
Just had this happen a couple of times. In my case, I had forgotten to fill the water reservoir and awoke from the hot air. I filled it and the fault occurred. Found this in Resmed's documentation. The third one (which I bolded) is my answer. Let it cool before adding water. That fixed it and I did not need to replace the humidifier.

Quote:Displays error message: Humidifier fault, replace humidifier:

Device has been left in a hot environment. => Allow to cool before re-use. Disconnect the power cord and then reconnect it to restart the device.

There is a fault in your H5i heated humidifier. => Discontinue using your H5i and contact your clinician/service provider.

Refilling the humidifier with cold water while humidifier is still hot after therapy. => Allow the humidifier to cool before re-filling.

Filling the humidifier with ice cold water on a warm day or with hot water. => Use room temperature water.
Post Reply Post Reply
#5
After 8'731 hours, my 5Hi humidifier now suddenly displays the error "Humidifier fault, replace humifier". As soon as the error displays, the blower also stops working. Will go to the DME tomorrow, to see about replacing humidifier, or getting a new unit.

;-< gjw
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Thumbsup [CPAP] resmed s9 auto set error message high temperature fault refer to owners manual beebsdaddy49 2 127 03-10-2017, 07:49 PM
Last Post: trish6hundred
  Resmed S9 End of life Message? MarkBart 15 790 03-07-2017, 08:44 PM
Last Post: MarkBart
Sad [CPAP] Res Med S9 help! Error: Settings reset to default values, please contact service pr dexter020281 14 11,159 02-18-2017, 01:22 AM
Last Post: vsheline
  SD Card Error (Nothing new) justMongo 3 240 02-02-2017, 10:22 AM
Last Post: Crimson Nape
  [Equipment] respironics bipap error code hanasdaddy 6 481 12-07-2016, 01:48 AM
Last Post: hanasdaddy
  [CPAP] Error code E33 1captnronrosser 3 521 10-14-2016, 04:10 PM
Last Post: trish6hundred
  [CPAP] Service Required Error Codes - Inner Workings of a CPAP machine SuperSleeper 18 72,642 10-10-2016, 01:48 PM
Last Post: juan345

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.

For any more information, please use our contact form.