#1
ResMed aircurve 10V auto
Have COPD Need help with settings. Started on Aircurve 10V auto by ResMed current settings drop sat way low. Also using 4 L O2. See current settings, any suggestions.
Mode V auto
Max IPAP 10.0
Min EPAP 4.0
PS 5.0
T1 Max 3.0
T1 Min 1.0
Post Reply Post Reply
#2
RE: ResMed aircurve 10V auto
You didn't say what problem you are having, but if you feel like not enough air,then raise the max IPAP, and consider raising EPAPmin if you have obstructive events. Under current settings your range is limited to 9/4 to 10/5 (IPAP/EPAP). Clearly you want more than 1-cm of auto adjustment. If you can download #Sleepyhead, it will help us to further refine your needs.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: ResMed aircurve 10V auto
Hi Kerry Abbott,
WELCOME! to the forum.!
Good luck to you with CPAP therapy.
trish6hundred
Post Reply Post Reply
#4
RE: ResMed aircurve 10V auto
did you do your own settings?
I would think you need more PS and less Ti. If you think the machine has been played with, I would do a factory default in setting and work from there. At least the default settings won't hurt you.

It would be worth the money, to get a doctor/provider to set it up for you. I've been reading up on the bilevels for a month and they are still a mystery. I've got a vpap St coming next week, I intend to use it in S mode without backup, because I run high pressure.

Your s10 Vauto would be different to the s9 and I'm unfamiliar with it, is the Ti auto set? is trigger and cycle auto set?
For a guide, this would be the typical starting settings, for 'manual' settings of obstructive/COPD
https://www.resmed.com/us/dam/documents/...lo_eng.pdf
page 23
min 5 max 13
PS 8
Ti max. 1.0
Ti min. 0.3
rise time. 150
trigger. medium
cycle. high

they want the air in quick and more time for exhale, so the lung is empty before another breath. good gas exchange and to prevent stacking.
mask fit http://www.apneaboard.com/wiki/index.php...ask_Primer
For auto-cpap, from machine data or software. You can set the min pressure 1 or 2cm below 95%. Or clinicians commonly use the maximum or 95% pressure for fixed pressure CPAP, this can also be used for min pressure.
https://aasm.org/resources/practiceparam...rating.pdf
Post Reply Post Reply
#5
RE: ResMed aircurve 10V auto
My mother suffers from COPD and since going on Capp I have been researching Capp therapy for her.  I found this article which may help.  There is much more to the article

http://www.aast.org/Assets/5e4309dd-c952...e-2014-pdf

There are different protocols in which to increase o2 and decrease co2, explained in link
Post Reply Post Reply
#6
RE: ResMed aircurve 10V auto
Great information thank you
Post Reply Post Reply
#7
RE: ResMed aircurve 10V auto
as long as you understand that the machine is adjusted to your lungs and not your lungs to the machine.
You really need a doctor to set your target tidal volume etc for your lungs. It's not plug and play.
mask fit http://www.apneaboard.com/wiki/index.php...ask_Primer
For auto-cpap, from machine data or software. You can set the min pressure 1 or 2cm below 95%. Or clinicians commonly use the maximum or 95% pressure for fixed pressure CPAP, this can also be used for min pressure.
https://aasm.org/resources/practiceparam...rating.pdf
Post Reply Post Reply
#8
RE: ResMed aircurve 10V auto
How did the change of settings go?
Is the min 5 max 13 PS:8 giving you enough air volume, or do you need more max pressure?
You would have seen from reading the forum, that the sleepyhead program here and a cms-50f o2 recorder off ebay, will be a big help to monitor your treatment.
mask fit http://www.apneaboard.com/wiki/index.php...ask_Primer
For auto-cpap, from machine data or software. You can set the min pressure 1 or 2cm below 95%. Or clinicians commonly use the maximum or 95% pressure for fixed pressure CPAP, this can also be used for min pressure.
https://aasm.org/resources/practiceparam...rating.pdf
Post Reply Post Reply
#9
RE: ResMed aircurve 10V auto
forgive me if I am wrong, but I believe min EPAP of 5 and max IPAP of 13, and PS is 8, then there is no adjustment left to the machine, it stays at the same setting all night long. (13-8=5)
Post Reply Post Reply
#10
RE: ResMed aircurve 10V auto
I posted the link and suggested a doctor, other than that I'm guessing. It's only because i've been reading up and bought a bilevel S/T for very high pressure and obesity, that I have any clue for a conversation.

It depends if those numbers were used in the avap s mode or avap auto mode, whether they were fixed or adjustable. For avap you could use the trigger and cycle numbers with min 5cm PS:8 and raise the max pressure to 25cm. The pressure and trigger numbers are just the default for copd. They and somewhere to start. They could be set as the base auto settings or used to manually tritrate from there, as per the link. It really needs a doctor to advise, a copd patient on bipap and 4lt O2 isn't to be played with.

From reading, the settings are adjusted from there, as per the patient requirements. The original settings were wrong and straight to auto with min epap 4cm ps:5, wouldn't be best practice with copd. including the wrong trigger settings. AFAIK, the pressure would only rise from a OA/H and not react to the COPD for volume and respiration rate.

At the end of the day the OP may have 5/17 PS:12 as a base titrate and run auto 5-25cm min PS:12. or OP may max out on 5/25cm PS:20. Or even move to a higher output machine. Fixed NIV 5/40 PS:35 isn't unheard of for copd.
I think like adjusting an apap, you have to get into the target zone first. only with apap it is an initial 4-20cm and epr/flex what you like for the first night.
mask fit http://www.apneaboard.com/wiki/index.php...ask_Primer
For auto-cpap, from machine data or software. You can set the min pressure 1 or 2cm below 95%. Or clinicians commonly use the maximum or 95% pressure for fixed pressure CPAP, this can also be used for min pressure.
https://aasm.org/resources/practiceparam...rating.pdf
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Rant and AirCurve 10 vs AirCurve 11 question Whacked 11 298 03-25-2024, 03:14 PM
Last Post: Sleepster
  [Treatment] Johnny O's ResMed AirCurve 10 VAuto for severe sleep apnea (61 AHI) and OSA, CSA JohnnyO 23 553 03-21-2024, 05:02 PM
Last Post: JohnnyO
  Plmnb's New Therapy ResMed AirCurve 10 ST-A Plmnb 204 8,492 03-19-2024, 06:31 AM
Last Post: Plmnb
  Trouble adapting to ResMed AirCurve 10 ASV JanetW 6 205 03-19-2024, 01:58 AM
Last Post: stevew168
  Aircurve 10 ST-A and Lumis 150 VPAP ST-A just a UK branding difference? SingleH 37 2,233 03-18-2024, 07:51 AM
Last Post: SingleH
  Oscar not recording CSR and RE with Aircurve 10 Vauto rmfisher132 1 87 03-17-2024, 04:33 PM
Last Post: Crimson Nape
  new machine Resmed Aircurve 10 Vauto Uzupan 32 1,347 03-10-2024, 07:29 AM
Last Post: Uzupan


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.