ResScan says doing great, pulse oximeter says not so good!!
#1
ResScan says doing great, pulse oximeter says not so good!!
I use a dental device, but recently decided it was not quite cutting it so went back for round two of CPAP http://www.apneaboard.com/forums/Thread-...IATE-INPUT I am still using the dental device which can only help and with which I have no compliance problems at all. I think I will be able to live with the Quattro FX full face mask.

Here is what ResScan said about last night:

Apnea index .2
Obstructive .1
Central .1
Hypopnea .3
AHI .5
Leaks under 20 the whole night.

So outstanding, right?

No so fast.....

Pulse oximeter said

Event index of 11.3

(Event is defined as a drop in O2 of at least 4% for at least ten seconds.)

There were 66 events 94-90% range. 11 events 89-85% range.

Oxygen below 88% .3 minutes.

Graphs show one dip to about 85% O2 with a simultaneous pulse surge to 100. I believe this was what woke me up about 4:00 AM.

It had been suggested by my sleep doctor and sleep dentist that the pulse oximeter gives false readings due to moving your hand. I tested this by recording a few minutes of waving my hand around as hard as I could. The pulse oximeter works much better than it was given credit for. No matter how hard I sling my hand, it does not corrupt O2 readings, it does not show heart rate spikes, it DROPS a few pulse beats of pulse.

I wonder how common is a false sense that all is hunky-dory based on the CPAP reporting alone?

I am running the CPAP at very conservative pressures right now with compliance my primary goal, but it looks like I will be ramping it up based on the pulse oximeter, and the goal of no awakenings.

I did have one night that the pulse oximeter was much happier with -- reporting 2.3 events.

Don in Austin
Post Reply Post Reply
#2
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 10:22 AM)Don in Austin Wrote: Pulse oximeter said Event index of 11.3
(Event is defined as a drop in O2 of at least 4% for at least ten seconds.)

There were 66 events 94-90% range. 11 events 89-85% range.

Oxygen below 88% .3 minutes.

Graphs show one dip to about 85% O2 with a simultaneous pulse surge to 100. I believe this was what woke me up about 4:00 AM.

spending only .3 minutes below 88% really is doing good enough.

correlate the oximeter report with sleepyhead, bearing in mind that o2 dips will be half to a minute later. you'll probably see an apnea and a corresponding dip.

xpap isn't perfect, but it's keeping you in a good enough range.

my friend with congestive heart failure spent an hour and a half under 88%, even with his ASV. so now he's also got 2lpm of o2 supplementation, because his heart isn't pumping enough blood at night to keep everything going, even with the breathing.

consider yourself lucky and doing well enough Smile

Post Reply Post Reply
#3
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 11:16 AM)diamaunt Wrote:
(07-10-2014, 10:22 AM)Don in Austin Wrote: Pulse oximeter said Event index of 11.3
(Event is defined as a drop in O2 of at least 4% for at least ten seconds.)

There were 66 events 94-90% range. 11 events 89-85% range.

Oxygen below 88% .3 minutes.

Graphs show one dip to about 85% O2 with a simultaneous pulse surge to 100. I believe this was what woke me up about 4:00 AM.

spending only .3 minutes below 88% really is doing good enough.

correlate the oximeter report with sleepyhead, bearing in mind that o2 dips will be half to a minute later. you'll probably see an apnea and a corresponding dip.

xpap isn't perfect, but it's keeping you in a good enough range.

my friend with congestive heart failure spent an hour and a half under 88%, even with his ASV. so now he's also got 2lpm of o2 supplementation, because his heart isn't pumping enough blood at night to keep everything going, even with the breathing.

consider yourself lucky and doing well enough Smile

Things could certainly be much worse. But it would be nice not to wake up in the middle of the night, and I still have some daytime sleepiness. Just as an academic issue, there does seem to be a conflict between ResScan says near perfect, the reality is not too bad but could be better. I was at only 8 for max inhale pressure, (thanks to assistance of dental device) so there is room for improvement.

Don in Austin

Post Reply Post Reply
#4
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 11:28 AM)Don in Austin Wrote:
(07-10-2014, 11:16 AM)diamaunt Wrote:
(07-10-2014, 10:22 AM)Don in Austin Wrote: Pulse oximeter said Event index of 11.3
(Event is defined as a drop in O2 of at least 4% for at least ten seconds.)

There were 66 events 94-90% range. 11 events 89-85% range.

Oxygen below 88% .3 minutes.

Graphs show one dip to about 85% O2 with a simultaneous pulse surge to 100. I believe this was what woke me up about 4:00 AM.

spending only .3 minutes below 88% really is doing good enough.

correlate the oximeter report with sleepyhead, bearing in mind that o2 dips will be half to a minute later. you'll probably see an apnea and a corresponding dip.

xpap isn't perfect, but it's keeping you in a good enough range.

my friend with congestive heart failure spent an hour and a half under 88%, even with his ASV. so now he's also got 2lpm of o2 supplementation, because his heart isn't pumping enough blood at night to keep everything going, even with the breathing.

consider yourself lucky and doing well enough Smile

Things could certainly be much worse. But it would be nice not to wake up in the middle of the night, and I still have some daytime sleepiness. Just as an academic issue, there does seem to be a conflict between ResScan says near perfect, the reality is not too bad but could be better. I was at only 8 for max inhale pressure, (thanks to assistance of dental device) so there is room for improvement.

Don in Austin

I would suggest that you simultaneously scroll through the flow rate and flow limitation graphs when the flow rate is zoomed in enough where you can see individual breaths (about a 10 minute window). Look for high spots in the flow limitation graph and FLAT spots in the flow rate graph that are roughly correlated to the times of the O2 desats if you know them.

You may be having respiratory events that are just a bit too short to show up as "flagged" events in ResScan, but are still long enough to cause you problems with the O2 levels. You might also see some serious flow limitations, which occur when the airway is compromised, but not compromised enough to be scored as an OA or an H. But some people do have O2 desats with prolonged flow limitations.

If you can't figure out a way to zoom in on the appropriate curves in ResScan, you might want to try SleepyHead instead and read through my Sleepy Head Guide
Questions about SleepyHead?  
See my Guide to SleepyHead
Post Reply Post Reply
#5
RE: ResScan says doing great, pulse oximeter says not so good!!
sadly, the human body is a deeply flawed conglomeration of evolutionary results.

unfortunately, (evolutionarily speaking) sleep apnea tends to hit *after* genes have already been passed on, else it would likely have died out way back when.
Post Reply Post Reply
#6
RE: ResScan says doing great, pulse oximeter says not so good!!
I've worn my CMS50F some nights. It appears to overstate events.
I find that under the 4% drop criterion, I have more events than reported by my APAP.
Further, most do not seem to correlate with events recorded by my APAP.
I tend to run at a low baseline O2 wise. Likely Pickwickian Syndrome; as I have never smoked.
Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#7
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 01:25 PM)justMongo Wrote: I've worn my CMS50F some nights. It appears to overstate events.
I find that under the 4% drop criterion, I have more events than reported by my APAP.
Further, most do not seem to correlate with events recorded by my APAP.
I tend to run at a low baseline O2 wise. Likely Pickwickian Syndrome; as I have never smoked.

I am most concerned with the O2 drops accompanied by heart rate surge and awakening.

Don in Austin

Post Reply Post Reply
#8
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 01:55 PM)Don in Austin Wrote: I am most concerned with the O2 drops accompanied by heart rate surge and awakening.

Don in Austin

did you have an apnea at that time, perhaps something else precipitated the awakening.
Post Reply Post Reply
#9
RE: ResScan says doing great, pulse oximeter says not so good!!
The magic number I watch is the time below 88%. Due to lung scarring, my nighttime average hovers very close to the 89-91% area, and I score any night that my 88% time is under 3 minutes (lately, it has stayed under 1 minute).

I completely ignore what the Oximeter calls an 'event', but rather watch the graph itself, and the nightly totals.
*I* am not a DOCTOR or any type of Health Care Professional.  My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
Post Reply Post Reply
#10
RE: ResScan says doing great, pulse oximeter says not so good!!
(07-10-2014, 03:09 PM)Peter_C Wrote: The magic number I watch is the time below 88%. Due to lung scarring, my nighttime average hovers very close to the 89-91% area, and I score any night that my 88% time is under 3 minutes (lately, it has stayed under 1 minute).

I completely ignore what the Oximeter calls an 'event', but rather watch the graph itself, and the nightly totals.

that's how I read 'em too... Smile
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Idea [CPAP] New to CPAP, Good So Far! TWINDAWN 2 72 Yesterday, 04:30 AM
Last Post: UnicornRider
  Pulse oximeter recommendations? p123 15 1,220 03-26-2024, 04:57 PM
Last Post: UnicornRider
  Need help to find a good Luna G3 pressure. johnathanb76 13 204 03-22-2024, 11:06 AM
Last Post: Sleeprider
  [Equipment] ResMed AirSense 11 compatible pulse oximeters timtaller 63 25,803 03-16-2024, 11:27 AM
Last Post: e727147
  Good AHI but Low Tidal Volume and Minute Ventilation erolalper 1 148 03-15-2024, 07:57 AM
Last Post: BoxcarPete
  Looking for wrist oximeter to check O2 storywizard 0 90 03-13-2024, 05:29 PM
Last Post: storywizard
  25 years of cpap, and still not working good Johan van der veen 2 178 03-08-2024, 10:15 AM
Last Post: Johan van der veen


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.