Resmed F20 leak rate
#1
What is everyone getting for a leak rake in L/min for the RESMED AirTouch F20
Post Reply Post Reply



Donate to Apnea Board  
#2
RESMED AirTouch F20 what is everyone getting for a leak rate.
My mask is only leaking from the connection of headgear to the mask plate itself this is suppose to be a intentional leak but it is huge. 50 l/min .
Post Reply Post Reply
#3
RESMED AirTouch F20 what is everyone getting for a leak rate.

My mask is only leaking from the connection of headgear to the mask plate itself this is suppose to be a intentional leak but it is huge. 50 l/min .
Post Reply Post Reply
#4
When I'm wearing it is very loud from the amount of air escaping anyone have pics on how the seal to mask is suppose to look. The mask was assembled when i got it maybe it is missing something.
Post Reply Post Reply
#5
Put you mask on try the mask fit feature. The only time i hear mine is if i have a leak.. Yes i dose have air coming out where hose is attached. Try taking it apart use a tooth brush and daw dish soap and clean it. See if that helps. Also post your data that would help too.
Post Reply Post Reply



Donate to Apnea Board  
#6
(11-14-2017, 07:23 AM)tricountytrail Wrote: RESMED AirTouch F20 what is everyone getting for a leak rate.
My mask is only leaking from the connection of headgear to the mask plate itself this is suppose to be a intentional leak but it is huge. 50 l/min .

The design leak rate of the Airtouch F-20 is 48 L/min at 16 cm pressure, so your leak rate may not be out of specification for the mask.  See the user's manual page 17 for further details. https://www.resmed.com/us/dam/documents/...er_eng.pdf
Post Reply Post Reply
#7
The leak rate specification is in the user's manual Page 17 https://www.resmed.com/us/dam/documents/...091_airfit- airtouch-f20_user-guide_amer_eng.pdf

Your machine does not report total leak rate, only unintentional, so you need to adjust that number down if you're looking at machine data or Sleepyhead.
Post Reply Post Reply
#8
Technical specifications
Pressure-flow curve The mask contains passive venting to protect
against rebreathing. As a result of manufacturing
variations, the vent flow rate may vary.
Pressure
(cm H2O)
Flow
(L/min)
3 19
8 32
10 37
12 41
16 48
20 54
24 60
28 66
30 69
32 72
36 77
40 82
Dead space
information
Physical dead space is the empty volume of the mask to the end
of the swivel. Using the large cushions it is 240 mL.
Therapy pressure 3 to 40 cm H2O
Resistance with
Anti Asphyxia Valve
(AAV) closed to
atmosphere
Drop in pressure measured (nominal)
at 50 L/min: 0.2 cm H2O
at 100 L/min: 0.6 cm H2O
Inspiratory and
expiratory
resistance with Anti
Asphyxia Valve
(AAV) open to
atmosphere
Inspiration at 50 L/min: 0.6 cm H2O
Expiration at 50 L/min: 0.7 cm H2O
Post Reply Post Reply
#9
(11-14-2017, 08:36 AM)Sleeprider Wrote:
(11-14-2017, 07:23 AM)tricountytrail Wrote: RESMED AirTouch F20 what is everyone getting for a leak rate.
My mask is only leaking from the connection of headgear to the mask plate itself this is suppose to be a intentional leak but it is huge. 50 l/min .

The design leak rate of the Airtouch F-20 is 48 L/min at 16 cm pressure, so your leak rate may not be out of specification for the mask.  See the user's manual page 17 for further details. https://www.resmed.com/us/dam/documents/...er_eng.pdf

Thanks My Air reports it At 56 L/min AHI 6.7 and it is much louder than my nasal mask which when i use is much quiter leak 2-4 l/min and ahi under 2

Something is going on this is my first full mask
Post Reply Post Reply



Donate to Apnea Board  
#10
tricountytrail Wrote:
RESMED AirTouch F20 what is everyone getting for a leak rate.
My mask is only leaking from the connection of headgear to the mask plate itself this is suppose to be a intentional leak but it is huge. 50 l/min .

The design leak rate of the Airtouch F-20 is 48 L/min at 16 cm pressure, so your leak rate may not be out of specification for the mask. See the user's manual page 17 for further details. https://www.resmed.com/us/dam/documents/...er_eng.pdf

Thanks My Air reports it At 56 L/min AHI 6.7 and it is much louder than my nasal mask which when i use is much quiter leak 2-4 l/min and ahi under 2

Something is going on this is my first full mask
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  High flow rate? - another unwell feeling, O2 dropping and quit breathing MyronH 18 367 06-18-2018, 09:03 AM
Last Post: MyronH
  Unwell feeling, SpO2 drops immediately when falling asleep, large Flow Rate MyronH 4 224 06-13-2018, 07:32 AM
Last Post: MyronH
  [Pressure] First Night on APAP - leak advice Zedoary 25 943 06-12-2018, 10:06 AM
Last Post: ShaunBlake
  Mask Leak rooy1960 16 880 06-03-2018, 01:54 PM
Last Post: sheepless
  questions about flow rate waveform sheepless 14 560 05-29-2018, 12:22 PM
Last Post: sheepless
  Pulse Rate Range While Sleeping Dream Machine 6 210 05-18-2018, 09:45 PM
Last Post: Mosquitobait
  [Equipment] Need help for a mask that will not leak at high pressure dfellows5 4 206 05-18-2018, 06:37 PM
Last Post: tedvpap

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.