Response time to Flow limitations
#1
Response time to Flow limitations
Relative to Resmed Airsense 11 Autoset Elite in Auto mode

Asking for a friend.... 

How quickly does the pressure increase when a flow limitation is observed? 
(I assume that the amount the pressure increases is a function of degree of Flow limitation - but please correct me if that is incorrect)
Post Reply Post Reply






Donate to Apnea Board  
#2
RE: Response time to Flow limitations
In's not only a matter of "response time", I presume. See https://patents.google.com/patent/US20110203588A1/en for all details.
Post Reply Post Reply
#3
RE: Response time to Flow limitations
Here is a statement that may be interpreted and used more easily than that very helpful patent info.

Source of the fragment from a large table: Treatment of sleep-disordered breathing with positive airway pressure devices: technology update - PMC (nih.gov).

   
I have no particular qualifications or expertise with respect to the apnea/cpap/sleep related content of my posts beyond my own user experiences and what I've learned from others on this site. Each of us bears the burden of evaluating the validity and applicability of what we read here before acting on it.  

Of my 3 once-needed, helpful, and adjunctive devices I have listed, only the accelerometer remains operative (but now idle). My second CMS50I died, too, of old age and the so-so Dreem 2 needs head-positioning band repair--if, indeed, Dreem even supports use of it now.



 
Post Reply Post Reply
#4
RE: Response time to Flow limitations
response setting would probably be a factor as well
This website is not secure. Your credentials are being sent in cleartext.
Post Reply Post Reply






Donate to Apnea Board  
#5
RE: Response time to Flow limitations
I believe there are multiple factors including severity of flow limitation, quantity/duration of flow limited breaths, algorithm (Autoset or Autoset for Her), response setting (standard or soft), and current pressure (pressure increases are more aggressive up to 12 cm and then less aggressively after that).

As as side note the Elite model is fixed pressure so I have assumed you mean Autoset (without Elite after it).
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Ramp Time JeffL 4 113 2 hours ago
Last Post: Sleeprider
  Trying to learn what OSCAR data means and I need help understanding my flow limits emmylouwho 7 310 Yesterday, 03:17 PM
Last Post: emmylouwho
  Flow Rate and Flow Limit Graphs ovmeadows 16 1,309 03-26-2023, 05:04 PM
Last Post: Sleeprider
  Pressure vs. Flow Rate Graph enjoysleep59 2 115 03-24-2023, 04:47 PM
Last Post: enjoysleep59
  How can you snore without any flow limits? TexasDon 2 188 03-19-2023, 12:03 AM
Last Post: Phaleronic
  [Equipment] When is it time to let go (mask cushion) StratCat48 2 163 03-18-2023, 08:46 PM
Last Post: bedtimebob
  Flow Rate Value Clarification in OSCAR cmpman1974 8 305 03-18-2023, 10:33 AM
Last Post: cmpman1974


New Posts   Today's Posts






About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.