Rough night!
Hi All,

Still new to all of this. I have been averaging around 5 AHI for the last month. Last night it was 12.9 and I captured a couple of snippets from the sleepyhead software to review. I'm not to familiar with how to interpret it so I thought I'd put it up on the boards for input (any and all is appreciated!). It looks like I am having alot of OA's in these strips. Wonder if I need to up my pressure (it is 8-13 now).

Thanks very much in advance Thanks

Monte


Attached File(s) Thumbnail(s)
           
07-20-2013 10:29 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #2
RE: Rough night!
Welcome

Anything above a 5 for AHI (nightly) is not good.
If you can stay below 5 (most nights) your average should stay below 5.

I see one graph where you are doing 43 AHI. Those OA's in clusters like that are killers. Seriously NOT good.
If I was your doc I would set your low pressure at 13 and high at 18-20...
BUT, first you need to adapt to the therapy so you can get toughened up to handle higher pressures.

I am not a Dr so be sure to work closely with your primary physician on this. Just in case they forget you exist.
If all else post your experiences here and let the other members help you out.

Cheers and hang in there!

Sleep-well

"With ordinary talent and extraordinary perseverance, all things are attainable." - Thomas Foxwell Buxton

Cool
07-20-2013 11:05 AM
Find all posts by this user Post Reply Quote this message in a reply
montehotbike12 Offline

Members

Posts: 12
Joined: Jul 2013

Machine: remstar auto a flex
Mask Type: Full face mask
Mask Make & Model: quattro fx
Humidifier: system one
CPAP Pressure: 9 - 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Ft. Lauderdale , Fla.

Post: #3
RE: Rough night!
(07-20-2013 11:05 AM)Shastzi Wrote:  Welcome

Anything above a 5 for AHI (nightly) is not good.
If you can stay below 5 (most nights) your average should stay below 5.

I see one graph where you are doing 43 AHI. Those OA's in clusters like that are killers. Seriously NOT good.
If I was your doc I would set your low pressure at 13 and high at 18-20...
BUT, first you need to adapt to the therapy so you can get toughened up to handle higher pressures.

I am not a Dr so be sure to work closely with your primary physician on this. Just in case they forget you exist.
If all else post your experiences here and let the other members help you out.

Cheers and hang in there!

Sleep-well


Thanks Shastzi. They want another sleep study but my insurance won't pay. I have a clinicians manual so I can up the pressure myself.

Does anyone know how accurate the machine (Remstar auto a flex - system one) is when differentiating between obstructive apneas and central apneas? I understand that is what the belts are for in sleep studies. I have heard that higher pressures can cause central apneas. I have some hesitation in changing pressure because I don't know if these clusters of apneas are truly obstructive or central.??? Dont-know

Any input on this?

Thanks!

Monte
07-20-2013 12:35 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #4
RE: Rough night!
(07-20-2013 12:35 PM)montehotbike12 Wrote:  Does anyone know how accurate the machine (Remstar auto a flex - system one) is when differentiating between obstructive apneas and central apneas?
Detection accuracy of obstructed airway versus clear airway apneas
http://www.healthcare.philips.com/pk_en/...apneas.wpd
07-20-2013 01:45 PM
Find all posts by this user Post Reply Quote this message in a reply
bwexler Offline

Advisory Members

Posts: 563
Joined: Mar 2012

Machine: PRS 1 BiPAP autoSV Advanced, with Heated Tube humidifier
Mask Type: Nasal mask
Mask Make & Model: Sleepweaver √Član
Humidifier: Resperonics Heated Tube humidifier
CPAP Pressure: Max pressure 25.0 cm H20, Max
CPAP Software: SleepyHead EncorePro

Other Comments: CMS 50i with Software for SpO2

Sex: Male
Location: San Marcos, CA, USA

Post: #5
RE: Rough night!
(07-20-2013 10:29 AM)montehotbike12 Wrote:  Hi All,

Still new to all of this. I have been averaging around 5 AHI for the last month. Last night it was 12.9 and I captured a couple of snippets from the sleepyhead software to review. I'm not to familiar with how to interpret it so I thought I'd put it up on the boards for input (any and all is appreciated!). It looks like I am having alot of OA's in these strips. Wonder if I need to up my pressure (it is 8-13 now).

Thanks very much in advance Thanks

Monte

You don't show leak rate in these samples. If your mask leaks that can cause major problems.
If have no leaks I would raise the upper limit of pressure incrementally until you do not see plateaus at the upper limit.
Than consider raising the lower limit a bit.
You and I need to learn the terms in all the different parameters displayed by Sleephead Software so we can make sense out of our reports.
I firmly believe I care more about my results and health than any doctor or DME.

Bob
07-21-2013 02:19 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
montehotbike12 Offline

Members

Posts: 12
Joined: Jul 2013

Machine: remstar auto a flex
Mask Type: Full face mask
Mask Make & Model: quattro fx
Humidifier: system one
CPAP Pressure: 9 - 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Ft. Lauderdale , Fla.

Post: #6
RE: Rough night!
(07-21-2013 02:19 PM)bwexler Wrote:  
(07-20-2013 10:29 AM)montehotbike12 Wrote:  Hi All,

Still new to all of this. I have been averaging around 5 AHI for the last month. Last night it was 12.9 and I captured a couple of snippets from the sleepyhead software to review. I'm not to familiar with how to interpret it so I thought I'd put it up on the boards for input (any and all is appreciated!). It looks like I am having alot of OA's in these strips. Wonder if I need to up my pressure (it is 8-13 now).

Thanks very much in advance Thanks

Monte

You don't show leak rate in these samples. If your mask leaks that can cause major problems.
If have no leaks I would raise the upper limit of pressure incrementally until you do not see plateaus at the upper limit.
Than consider raising the lower limit a bit.
You and I need to learn the terms in all the different parameters displayed by Sleephead Software so we can make sense out of our reports.
I firmly believe I care more about my results and health than any doctor or DME.

Bob


Thanks for the reply. Here is an expanded view to include leak rate.


Attached File(s) Thumbnail(s)
   
07-21-2013 03:32 PM
Find all posts by this user Post Reply Quote this message in a reply
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #7
RE: Rough night!
If this a typical reports, you need to show the reports to your sleep physician
Leak rates showing are minimal for full face mask "total leak rate"
I believe, you can change some setting in SleepyHead to show unintentional leak vs total leak

I have no idea if "system one resistance" settings makes any difference, reistance setting for Respironcis full face masks "X1", but from what have read ... resistance setting "0" for non-Respironcis masks
07-21-2013 04:38 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
montehotbike12 Offline

Members

Posts: 12
Joined: Jul 2013

Machine: remstar auto a flex
Mask Type: Full face mask
Mask Make & Model: quattro fx
Humidifier: system one
CPAP Pressure: 9 - 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Ft. Lauderdale , Fla.

Post: #8
RE: Rough night!
Just wanted to thank everyone for their input and help. Last PM was MUCH better. After looking at some of my graphs over the past couple of weeks I noticed most of my OA's were in clusters. I wondered if these might be times when I was sleeping on my back so I sewed a tennis ball into the back of a t shirt last PM. AHI was 2.32 last night compared to 12.5 and 8.9 the previous 2 nights (it has never been less than 5 since I started CPAP therapy).

Here is a clip from last PM.


Attached File(s) Thumbnail(s)
   
07-22-2013 09:13 AM
Find all posts by this user Post Reply Quote this message in a reply
DreamDiver Offline

Preferred Members-2

Posts: 77
Joined: Jul 2013

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: Quattro FX Full Face
Humidifier: ResMed S9 H5i
CPAP Pressure: 10.4 | 11.8
CPAP Software: ResScan SleepyHead EncoreBasic EncorePro

Other Comments:

Sex: Male
Location: Georgia

Post: #9
RE: Rough night!
(07-21-2013 03:32 PM)montehotbike12 Wrote:  Thanks for the reply. Here is an expanded view to include leak rate.
Congratulations on your first AHI below 5. Way to go!

If your leak line is this consistent from night to night, (seems very good on both nights), then it looks like you have that that under control. The upper line is the total leak, including expected leak from the mask. The lower line is the amount you're leaking over expected leak. If the leak line on either of these graphs were mine, I'd be content.

Image A:
[Image: attachment.php?aid=438]

Image B:
[Image: attachment.php?aid=439]

It looks like you have changed your pressure regime from the first night (lowest pressure 8.0 in Image A) to the second night (lowest pressure 9.0 in Image B). I wonder if that might be part of the reason why your AHI is lower. It looks like almost all your obstructives became hypopneas.

Image C:
[Image: attachment.php?aid=437]
In Image C , it looks like as soon as you hit 12.0cm pressure, everything became stable and you didn't have many events. If you took a closer look (30 minute span) at the 4:45 to the end of the night on image C, I suspect you would see relatively consistent breathing. Of course this is only an hour of information, so it helps to look at the entire night graphs you've provided too.

If these were my graphs, I'd try a few more nights on the current regime and see what the numbers look like, since this the first-ever 2.3 AHI, with a pressure line that is fairly flat across the night. It is my understanding the body takes a while to adjust to a new pressure regime.

At some point after a few days, I might consider dialing up my lower pressure setting a notch to see if I could get rid of those hypopneas, since most of my apneas would be obstructive and centrals are minimal.

Of course, I'm not a doctor, so these are merely opinions. Since you seem to be self-titrating, you are in charge of your own therapy. Good luck withe the tennis balls! I hope this has been helpful.

dreamdiver.org
(This post was last modified: 07-22-2013 02:38 PM by DreamDiver.)
07-22-2013 02:35 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
montehotbike12 Offline

Members

Posts: 12
Joined: Jul 2013

Machine: remstar auto a flex
Mask Type: Full face mask
Mask Make & Model: quattro fx
Humidifier: system one
CPAP Pressure: 9 - 16
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Ft. Lauderdale , Fla.

Post: #10
RE: Rough night!
Dreamdiver

Fantastic analysis! As you suggested I checked some other nights and the leak rate seems low. You are correct!....I changed my pressure from 8 to 9- I will keep it there for a few nights and see how it goes...then as you suggested I will bump it up again to see if the hypopneas improve.

I looked at the time from 4:45 AM until I woke up (Image C) and again you were correct....breathing pattern was great.

Thanks again for your help friend!

Monte
07-22-2013 03:26 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  1st night with DreamWear mask - Best night ever pwgphoto 34 5,335 08-30-2016 08:36 PM
Last Post: HoustonMIke
  [Treatment] Seeing significantly different AHI numbers night to night galenaforest 4 821 02-25-2016 02:27 AM
Last Post: Dagmar1
  Changed to ResMed Aircurve 10 VAUTO and Dreamwear mask - bit of a rough start Sacktime 15 2,323 02-04-2016 11:53 AM
Last Post: _cy_
  One night = 0 leaks, next night 30 L.....??? GrammaBear 3 738 01-28-2016 01:39 PM
Last Post: justMongo
  Rough Night on Titration - When Does the Suffocation feeling go away? linn 13 1,572 11-14-2015 03:57 PM
Last Post: linn
  First night behind me - 2nd night update - mostly Centrals? worn_out_in_lebanon 19 4,467 03-21-2015 11:11 AM
Last Post: retired_guy
  0.0 Events Last Night - Night #7 Sparkle 4 1,794 01-20-2015 09:47 PM
Last Post: Sparkle

Forum Jump:

Who's Online (Complete List)