Hi, a question for someone with a multimeter and a climateline hose.
Could you maybe measure the resistances across the 3 contacts for me please?
I've had no problems with rainout for over 2 yrs but am now getting some. I'm not sure if it's the climateline hose or how to check it's working.
I've not checked the voltage at the machine contacts yet, I thought I'd resistance check the heater coil in the tubing first. I'm getting 10 ohms between contacts 1+2 and open circuit between contact 3 and the others.
Have nothing to compare what is normal at the moment.
Any advice?
Many thanks.
G'day Wirdy, welcome to Apnea Board.
My multimeter's a bit old and dodgy but the following are approximately right:
Pin 1 - Pin 2: 5 ohm
Pin 1 - Pin 3: 175
Pin 2 - Pin 3: 175
Those numbers may be off a bit, but there's definitely a current between Pin 3 and the others.
In the States they apparently replace these hoses after 6 months. Mine's about 18 months and still OK, touch wood.
DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
It reads the same every which way, so likely a thermistor rather than a diode.
DeepBreathing
Apnea Board Moderator
www.ApneaBoard.com

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Hi wirdy,
WELCOME! to the forum.!
Hang in there for more responses to your post and much success to you with your CPAP therapy.