% SPONT C and %SPONT T - question?
I do not use software. But my machine does give some data, and for each set of data, it lists DAY | Week | month | six months | year.

Two data sets that I do not know what they mean, or why I get the results that I do are:

"% SPONT C"
"% SPONT T"

For the above two data sets the result is always " 0% " - for both?? Is this good or bad, and/or, what does this mean?

Also, my leak rate bounces between 0.70L/s and 0.38L/S - is this acceptable, or is this throwing my other results off? I get just a single number for this data, so am not clear if it is the highest leak, or the average leak. After inspecting mask hose chamber and headgear I can not find any issues, and this nasal mask fits well with no leaks when awake.

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-15-2013 11:16 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaulaO2 Offline
Wiki Editor
Moderators

Posts: 8,065
Joined: Feb 2012

Machine: S9 Autoset
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i
CPAP Pressure: 14-20
CPAP Software: SleepyHead

Other Comments: Ehlers-Danlos Syndrome, Hypermobility Type; chronic sarcasm

Sex: Undisclosed
Location: western NC, USA

Post: #2
RE: % SPONT C and %SPONT T - question?
Found this on another forum:

spont t
Quote:Spontaneous trigger rate is the rate at which your own breathing caused the transition from exhale to inhale (EPAP-to-IPAP). Ideally your body spontaneously triggers this machine transition 100% of the time (that's not always possible). But the machine's timing rate seems to be rushing each of your own inhale transitions instead of allowing for a natural, spontaneous transition.

spont c
Quote:Spontaneous cycle rate is the rate at which your own breathing caused the transition from inhale to exhale (IPAP-to-EPAP). Ideally your body spontaneously cycles this machine transition 100% of the time (again, not always possible). But the machine's timing rate seems to be rushing each of your own exhale transitions instead of allowing for a natural, spontaneous transition.

PaulaO2
Apnea Board Moderator
www.ApneaBoard.com


Breathe deeply and count to zen.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
10-15-2013 01:37 PM
Find all posts by this user Post Reply Quote this message in a reply
Peter_C Offline

Advisory Members

Posts: 1,308
Joined: Sep 2013

Machine: System One DS960HS
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro "MED"
Humidifier: Attached
CPAP Pressure: Auto mode-script is:15/3/3
CPAP Software: EncoreBasic Other Software

Other Comments: Severe obstructive sleep apnea 1999/Just got an ASV machine

Sex: Male
Location: Akron, Ohio USA

Post: #3
RE: % SPONT C and %SPONT T - question?
(10-15-2013 01:37 PM)PaulaO2 Wrote:  Found this on another forum:

spont t
Quote:Spontaneous trigger rate is the rate at which your own breathing caused the transition from exhale to inhale (EPAP-to-IPAP). Ideally your body spontaneously triggers this machine transition 100% of the time (that's not always possible). But the machine's timing rate seems to be rushing each of your own inhale transitions instead of allowing for a natural, spontaneous transition.

spont c
Quote:Spontaneous cycle rate is the rate at which your own breathing caused the transition from inhale to exhale (IPAP-to-EPAP). Ideally your body spontaneously cycles this machine transition 100% of the time (again, not always possible). But the machine's timing rate seems to be rushing each of your own exhale transitions instead of allowing for a natural, spontaneous transition.

Thank you for the explanation.
OK - so that's what it means. But what about the fact that my machine reads ZERO PERCENT for both - for DAY, WEEK, or MONTH?

Does that mean that my machine is broken, or I am broken? There is no setting to alter this (I think).

There is MIN IPAP, and MAX IPAP, but the max only goes to 4.0 seconds. I read that it stays in IPAP until either: I stop inhaling, or the max of 4 seconds goes by.

My machine stay in EPAP until it senses that I start to inhale, then switches to IPAP - I must be missing something here???

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-15-2013 03:21 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PaulaO2 Offline
Wiki Editor
Moderators

Posts: 8,065
Joined: Feb 2012

Machine: S9 Autoset
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i
CPAP Pressure: 14-20
CPAP Software: SleepyHead

Other Comments: Ehlers-Danlos Syndrome, Hypermobility Type; chronic sarcasm

Sex: Undisclosed
Location: western NC, USA

Post: #4
RE: % SPONT C and %SPONT T - question?
Not a clue. It could be you've not got it set to show that. Or you aren't switching on your own at all and the machine is doing it for you. Or you are doing it all the time and it is doing it 0%.

This is where using software may help some.

PaulaO2
Apnea Board Moderator
www.ApneaBoard.com


Breathe deeply and count to zen.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
10-15-2013 07:28 PM
Find all posts by this user Post Reply Quote this message in a reply
Peter_C Offline

Advisory Members

Posts: 1,308
Joined: Sep 2013

Machine: System One DS960HS
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro "MED"
Humidifier: Attached
CPAP Pressure: Auto mode-script is:15/3/3
CPAP Software: EncoreBasic Other Software

Other Comments: Severe obstructive sleep apnea 1999/Just got an ASV machine

Sex: Male
Location: Akron, Ohio USA

Post: #5
RE: % SPONT C and %SPONT T - question?
Hmm - I hope others might chime in with more info - I am still here and sleeping, breathing - so that's a plus, right Smile)

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-16-2013 12:23 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Peter_C Offline

Advisory Members

Posts: 1,308
Joined: Sep 2013

Machine: System One DS960HS
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro "MED"
Humidifier: Attached
CPAP Pressure: Auto mode-script is:15/3/3
CPAP Software: EncoreBasic Other Software

Other Comments: Severe obstructive sleep apnea 1999/Just got an ASV machine

Sex: Male
Location: Akron, Ohio USA

Post: #6
RE: % SPONT C and %SPONT T - question?
Anyone have any thoughts on my leak rate? Is this OK, high, low? I don't really know what is acceptable, and what isn't?

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-16-2013 10:39 PM
Find all posts by this user Post Reply Quote this message in a reply
zonk Offline

Advisory Members

Posts: 7,908
Joined: Feb 2012

Machine: A10 AutoSet
Mask Type: Nasal mask
Mask Make & Model: Activa LT
Humidifier: Integrated /ClimateLineAir
CPAP Pressure: 9/13
CPAP Software: ResScan

Other Comments: CPAP since Nov 2010

Sex: Male
Location: Australia

Post: #7
RE: % SPONT C and %SPONT T - question?
(10-16-2013 10:39 PM)Peter_C Wrote:  Anyone have any thoughts on my leak rate? Is this OK, high, low? I don't really know what is acceptable, and what isn't?
Hi Peter
The number you get from the machine screen is 95% percentile ... for 95% of the time you were at or below this number
ResMed consider leak rate above 0.4 L/s (24 L/m) is associated with patient discomfort, disturbed sleep, and reduced efficacy of treatment
You need the software for detailed data, not sure if your machine uses SD card or smart card (smart card require special card reader)
10-16-2013 11:09 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Peter_C Offline

Advisory Members

Posts: 1,308
Joined: Sep 2013

Machine: System One DS960HS
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro "MED"
Humidifier: Attached
CPAP Pressure: Auto mode-script is:15/3/3
CPAP Software: EncoreBasic Other Software

Other Comments: Severe obstructive sleep apnea 1999/Just got an ASV machine

Sex: Male
Location: Akron, Ohio USA

Post: #8
RE: % SPONT C and %SPONT T - question?
(10-16-2013 11:09 PM)zonk Wrote:  
(10-16-2013 10:39 PM)Peter_C Wrote:  Anyone have any thoughts on my leak rate? Is this OK, high, low? I don't really know what is acceptable, and what isn't?
Hi Peter
The number you get from the machine screen is 95% percentile ... for 95% of the time you were at or below this number
ResMed consider leak rate above 0.4 L/s (24 L/m) is associated with patient discomfort, disturbed sleep, and reduced efficacy of treatment
You need the software for detailed data, not sure if your machine uses SD card or smart card (smart card require special card reader)

Thanks for the answer. I am not sure what my machine uses either. There is a cover in back, but no slot for a memory stick, nor any stick in place, so I suspect it is for some kind of attachment.

I would assume that if my water chamber, hose, or mask had any leaks, they would be rather easy to find? I've been using my machine so long that it isn't so much that I open my mouth, but rather air simply escapes between my lips.

The odd part is that my leak rate is bout the same if not higher if I switch to my FFM mask (which of course, is not as comfortable).

I guess what I am wondering is how to pin down whatever leaks I have using only the tools I have to work with? Buying some kind of reader cable is not in the cards due to my fixed income, I do have spare new parts, so I guess I could simply start swapping things out? But am also wondering if there is anything I am missing - something else I should be checking?

AHI tends to bounce a bit, but stays under 10, and my AI hangs around 1.3 or so, so I am pleased with those numbers (if accurate??).

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-17-2013 11:44 PM
Find all posts by this user Post Reply Quote this message in a reply
vsheline Offline

Advisory Members

Posts: 1,907
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #9
RE: % SPONT C and %SPONT T - question?
(10-15-2013 03:21 PM)Peter_C Wrote:  
(10-15-2013 01:37 PM)PaulaO2 Wrote:  Found this on another forum:

spont t
Quote:Spontaneous trigger rate is the rate at which your own breathing caused the transition from exhale to inhale (EPAP-to-IPAP). Ideally your body spontaneously triggers this machine transition 100% of the time (that's not always possible). But the machine's timing rate seems to be rushing each of your own inhale transitions instead of allowing for a natural, spontaneous transition.

spont c
Quote:Spontaneous cycle rate is the rate at which your own breathing caused the transition from inhale to exhale (IPAP-to-EPAP). Ideally your body spontaneously cycles this machine transition 100% of the time (again, not always possible). But the machine's timing rate seems to be rushing each of your own exhale transitions instead of allowing for a natural, spontaneous transition.

Thank you for the explanation.
OK - so that's what it means. But what about the fact that my machine reads ZERO PERCENT for both - for DAY, WEEK, or MONTH?

Does that mean that my machine is broken, or I am broken?

Hi Peter_C,

I think there are 4 settings which can affect the % SPONT C and % SPONT T.

On your machine, what are the values for the following settings?
1. TiMin (min time machine stays in inhalation)
2. TiMax (max time machine stays in inhalation)
3. Trigger sensitivity
4. Cycle sensitivity

ADDED:

Clinical Guide for Auto 25 says:

Trigger/Cycle: VERY LOW/LOW/MED/HIGH/VERY HIGH

Trigger: "Sets the level of inspiratory flow above which the VPAP changes from EPAP to IPAP."
[Actually, the wording here in the Clinical Guide seems misleading, because a picture is included which shows clearly that the VERY HIGH setting causes the earliest (most sensitive) transition from EPAP to IPAP, when the inspiratory flow has not yet risen very much and is still low. The VERY LOW setting produces the lowest sensitivity to the start of inhalation, delaying the transition from EPAP to IPAP until the amount of inspiratory flow is larger.]

Cycle: "Sets the level of inspiratory flow below which the VPAP changes from IPAP to EPAP."
[The Clinical Guide includes a picture which shows clearly that the VERY HIGH setting causes the earliest (most sensitive) transition from EPAP to IPAP, when the Inspiratory flow has dropped a little but is still fairly high. A setting of VERY LOW delays the transition from EPAP to IPAP until the inspiratory flow has dropped to a low amount (but has not yet dropped to zero, so this is toward the very end of inhalation and before exhalation.]

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 10-20-2013 03:04 PM by vsheline.)
10-20-2013 04:21 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Peter_C Offline

Advisory Members

Posts: 1,308
Joined: Sep 2013

Machine: System One DS960HS
Mask Type: Full face mask
Mask Make & Model: Mirage Quattro "MED"
Humidifier: Attached
CPAP Pressure: Auto mode-script is:15/3/3
CPAP Software: EncoreBasic Other Software

Other Comments: Severe obstructive sleep apnea 1999/Just got an ASV machine

Sex: Male
Location: Akron, Ohio USA

Post: #10
RE: % SPONT C and %SPONT T - question?
Prior to last night, the 'trigger' and cycle' were on their default settings of 'MED'. I then changed them before going to sleep last night to 'LOW" (both of them).

Settings:
MODE: VAUTO
Max IPAP: 25
Min EPAP: 7
Press SUP: 6
Ti Max: 4.0s
Ti Min: 0.3s
EXHALATION: MED
SMARTSTART: ON
Mask: STANDARD
Leak Alert: OFF
TRIGGER: Low
CYCLE: Low
Tube Length: 3m
MaxSettle: off

Data from last night (10/19/2013)
USEAGE: 9:48hrs

Item | DAY | WEEK |
LEAK: 0.46L/s 0.48L/s
VT: 420-860 360-780
Resp Rate: 10-15 10-15
MV: 5.2-10.3 5.0-9.6
PRESS: 13.4 13.2
AHI: 19.6 11.2
AI: 4.4 2.3
%SPONT T: 0% 0%
%SPONT C: 0% 0%

I noticed no real change, other than having one of those nights where you don't even notice the air flow in your mask.

*I* am not a DOCTOR or any type of Health Care Professional. My thoughts/suggestions/ideas are strictly only my opinions.

"Only two defining forces have ever offered to die for you. Jesus Christ and the American Soldier. One died for your Soul, the other for your Freedom."
10-20-2013 11:17 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Forum Jump:

Who's Online (Complete List)