Settings help for an S9 VPAP Auto
Hello, I have been using an S9 Autoset for over a year now and it is working well, but I was recently given an S9 VPAP Auto and I want to try it out. I tried to get the instructions from RewsMed but I'm not a healthcare professional so I was blocked. Is there a link somewhere that I can use? Thanks, Jim

P.S. It came without it's own SD chip so can I move my chip to the new machine, or should I get a new chip for this machine?
(This post was last modified: 05-19-2016 05:53 AM by okrobie.)
05-19-2016 05:50 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
robysue Offline
Wiki Editor
Advisory Members

Posts: 1,245
Joined: Oct 2013

Machine: PR Dreamstation BiPAP Auto
Mask Type: Nasal pillows
Mask Make & Model: Swift FX
Humidifier: PR Dreamstation humidfier
CPAP Pressure: min EPAP = 4; max IPAP = 9;
CPAP Software: SleepyHead EncoreBasic EncorePro

Other Comments: Papping since September 2010

Sex: Female
Location: Buffalo, NY

Post: #2
RE: Settings help for an S9 VPAP Auto
okrobie,

You should be able to the clinical manual for your machine from the appropriate part of this web site.

You will need to think about the settings you want to use on the S9 VPAP Auto. Keep in mind that the combination of EPAP, IPAP, and PS settings replaces the Pressure and EPR settings.

To most closely match your current S9 AutoSet settings, you would use the following settings on the VPAP:
  • PS on the VPAP = EPR on the AutoSet
  • Min EPAP on the VPAP = Min Pressure - EPR on the AutoSet (provided this number > 4; otherwise Min EPAP = 4)
  • Max IPAP on the VPAP = Max Pressure on the AutoSet

I'd leave all the TiMin, TiMax, Trigger, and Cycle settings at their defaults to start with.

The biggest difference between the VPAP and the AutoSet is that PS can be set to something larger than 3. If you want to experiment with that, start slowly with a PS = 4. Keep in mind that for a (very) small number of PAPers, a large PS can have a tendency to trigger some CAs. So if your CAI goes up, cut back on the PS.

Questions about SleepyHead?
See my Guide to SleepyHead
05-19-2016 07:25 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Dream Station Ramp and Pressure Settings knowoneuno 27 565 Yesterday 09:02 PM
Last Post: Russatrice
  Newbie - need advice on settings and machine tmc712 18 391 12-06-2016 09:04 PM
Last Post: tmc712
  Dream Station Settings Autumnseer 15 563 11-26-2016 07:01 AM
Last Post: Weird Tolkienish Figure
  Who generally is in charge of your CPAP settings? Mezzrow 9 485 11-12-2016 01:47 AM
Last Post: blueelement2016
Question [Equipment] S9 VPAP AUTO QT2U 0 160 11-05-2016 12:27 AM
Last Post: QT2U
  do I have the right machine (and settings) JaminV 4 380 10-31-2016 10:05 PM
Last Post: trish6hundred
  Dreamstation heated tube/humidifier initial settings. NorthernGuy 9 748 10-28-2016 09:21 AM
Last Post: Perchas

Forum Jump:

Who's Online (Complete List)