(12-28-2012, 02:10 AM)bensonlee Wrote: Hi Folks,
It is my first week with S9 AutoSet and I am trying to download the data so that I could report back my progress to my doctor (seeing high leak rate: median of 84 L / min).
Hi bensonlee and welcome to apnea board
SH is for your own benefit ... your doc more likely to be familiar with Rescan
SH is not recognized as medical or legal document
the median leak rate at 84 L/min is too high and at this level ... the S9 cannot analyzed data correctly
before anything else you need to solve the leak issue
for most folk can take several masks to find the right one
full face masks can be a challenge and especially for newbies
try nasal pillows or nasal masks (with chinstrap) if you can
have you had titration study or you,re doing at home with Autoset
the reason asking, your pressure range set at 4-14
maybe the doc want to see the data to adjust the range