[Symptoms] not enough air flow
Author Message
calliopy2 Offline

Members

Posts: 12
Joined: Aug 2015

Machine: Phillips Respironics sytem one aut
Mask Type: Nasal mask
Mask Make & Model: wisp nasal mask
Humidifier: Respironics Remstar system one auto Aflex with hummidifier
CPAP Pressure: 11.5 to 18
CPAP Software: SleepyHead

Other Comments: ramp at 5 min at 9

Sex: Female
Location: california

Post: #1
Angry not enough air flow
New to cpap. Using for one month with no problems until last week. I feel short of breath and cant get enough air. When I start to fall asleep I wake up gasping. The airflow on the nasal pillows feel reduced. I checked the tubing and connection for leaks, cleaned the filter and tubes but still no increase in air flow. Has anyone else had this problem? need help
08-09-2015 12:07 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #2
RE: not enough air flow
Welcome

You're in the right place, calliopy2!

The folks here will help you fine tune that new machine so you can get the most out of it.

I think everyone will want to know the exact model # of your Respironics machine
for starters as well as your mask make & model.
Then folks here can help you figure out what pressure(s) you are set for and what to do next.
(Doctors do not have time to help out beginners much so that is why this website is here!)

Hang in there and never quit! Smile

"With ordinary talent and extraordinary perseverance, all things are attainable." - Thomas Foxwell Buxton

Cool
08-09-2015 12:12 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,845
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #3
RE: not enough air flow
As Shasti said, more info on your machine and present settings is needed to narrow down a cause.
Things that come to mind:
1) Using an Auto machine and having too low a starting pressure range. Some noobe's get started at 4 cm-H2O which is generally too low.
2) Using the ramp feature. Most people do better with the ramp turned off.
3) Too small nasal pillow cushion.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
08-09-2015 12:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,430
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #4
RE: not enough air flow
Hi calliopy2,
WELCOME! to the forum.!
Are you using ramp? if so, you might either decrease the time it runs or turn it completely off. Let us know what your pressure is and what kind of machine and mask you have so that people can better assist you.
Hang in there for more responses to your post and much success to you with your CPAP therapy.

trish6hundred
08-09-2015 05:32 PM
Find all posts by this user Post Reply Quote this message in a reply
car54 Offline

Advisory Members

Posts: 108
Joined: Apr 2014

Machine: PRS1 Model 560P
Mask Type: Full face mask
Mask Make & Model: Airfit F10
Humidifier: PR system one
CPAP Pressure: 9-11
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Milwaukee, WI

Post: #5
RE: not enough air flow
I felt like I could not breath because the pressure was making it hard to exhale. On my system one I put it in CFLEX mode which lowers the pressure while you exhale. It did the trick. It is also called expiratory pressure relief

Brad
08-09-2015 07:51 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Shastzi Offline

Advisory Members

Posts: 1,174
Joined: Dec 2012

Machine: ResMed S9 Autoset
Mask Type: Full face mask
Mask Make & Model: FitLife Total face mask
Humidifier: F&P HC150 with Hybernite heated hose.
CPAP Pressure: 15cm-20cm H2O (auto)
CPAP Software: SleepyHead Other Software

Other Comments: CMS50-F wearable Oximeter; Software: SPO2 Assistant

Sex: Undisclosed
Location: Florida, USA.

Post: #6
RE: not enough air flow
Usually it takes a while to acclimate to pressure as high as 10cm & above.

The feeling will go away after some time when you slowly get used to it.
When you wake up and think your machine is running, but not working, check your exhaust vents and you'll
likely feel the air blowing out.
Then you'll know you have adapted.
Wink

"With ordinary talent and extraordinary perseverance, all things are attainable." - Thomas Foxwell Buxton

Cool
08-10-2015 03:41 PM
Find all posts by this user Post Reply Quote this message in a reply
ArthurLaurent Offline

Members

Posts: 11
Joined: Jul 2015

Machine: ResMed AirCurve 10 ASV
Mask Type: Full face mask
Mask Make & Model: ResMed
Humidifier: ResMed Internal
CPAP Pressure: 10-25 (ASV)
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Virginia

Post: #7
RE: not enough air flow
If you think the reduced airflow is due to a mechanical (rather than software) problem, in addition to the cleaning and looking for tube leaks you've already done, check the connections between (1) the tubing and the mask, and (2) the tubing and your machine. Also check to make sure there are no kinks or sharp corners in your tubing. I hope you have success!
08-11-2015 09:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
OpalRose Offline

Monitors

Posts: 3,167
Joined: Nov 2014

Machine: PR System One REMstar Auto 560 with A Flex
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10 For Her
Humidifier: REMStar heated humidifier with heated hose
CPAP Pressure: Auto Cpap 10-13 AFLEX 1
CPAP Software: SleepyHead EncoreBasic

Other Comments: Started CPAP Therapy October 23, 2014

Sex: Female
Location: Northeast Ohio, USA

Post: #8
RE: not enough air flow
(08-09-2015 12:07 PM)calliopy2 Wrote:  New to cpap. Using for one month with no problems until last week. I feel short of breath and cant get enough air. When I start to fall asleep I wake up gasping. The airflow on the nasal pillows feel reduced. I checked the tubing and connection for leaks, cleaned the filter and tubes but still no increase in air flow. Has anyone else had this problem? need help

Hi calliopy2,
Welcome to the forum. Please filled out your profile.

What pressure are you using and if you are using the ramp, turn it off.
If your pressure is too low, you will feel as if you are not getting enough air.

Also, what nasal pillow mask are you using. You can try a size larger pillow to see if that helps.

OpalRose
08-12-2015 06:36 AM
Find all posts by this user Post Reply Quote this message in a reply
calliopy2 Offline

Members

Posts: 12
Joined: Aug 2015

Machine: Phillips Respironics sytem one aut
Mask Type: Nasal mask
Mask Make & Model: wisp nasal mask
Humidifier: Respironics Remstar system one auto Aflex with hummidifier
CPAP Pressure: 11.5 to 18
CPAP Software: SleepyHead

Other Comments: ramp at 5 min at 9

Sex: Female
Location: california

Post: #9
RE: not enough air flow
To all, thanks for all the help. Sorry for the response delay. I was hospitalized for pneumonia and just got discharged today. Having a hard time finding out info. The bill of sale lists each part by number not names and on the machine, the supply company has placed their plaquard info probably hiding what I am looking for. Did find out the problem. Inside the tubing is another flimsy soft plastic coil which was twisted. In the hospital, I had no problems with the cpap. I suspect, like my wheelchair, I must have a walmart quality machine. I did take the hospital tubing and nasal mask. They appear to be of a better quality
08-14-2015 06:03 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Source of Resmed S9 EPR reverse flow whistle! neal 5 225 12-01-2016 07:22 PM
Last Post: klv329
  Flow Rate Interpretation nadprok 2 216 11-13-2016 06:08 PM
Last Post: nadprok
  [Equipment] Modding Resmed FX pillow mask, no turbo air flow DaveDog 9 345 10-23-2016 08:09 PM
Last Post: PaulaO2
  Any way to determine REM sleep from Flow Rate? Roy289 18 1,125 08-31-2016 10:35 AM
Last Post: o1bob037
  Bizarre Sleepyhead Reading for Flow Limitation holden4th 7 465 07-24-2016 01:58 PM
Last Post: vsheline
  Help Interpreting Flow Rate Graph nadprok 15 1,205 06-20-2016 06:11 PM
Last Post: Sleeprider
  Pressure, Flow, and Leaks chill 12 811 05-29-2016 12:20 PM
Last Post: chill

Forum Jump:

Who's Online (Complete List)