System One no longer heating hose
I couldn't say exactly when it happened, but it's gotten very noticeable this winter with really cold temps. At some point though my System One stopped heating the tube.

I've gone into the clinical menu and the options are locked and can't be changed.

I thought maybe my tube was bad, but I just got a new one that I tried and it also doesn't work. So I've tried three tubes including a brand new one and the options remain locked.

Does this mean the unit is not detecting the hose is heatable and won't unlock the option? Or is there some super secret way of turning off the lock.

The "system lock" option is already turned off and I can change other clinician options like pressure.
01-04-2015 05:05 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
DocWils Offline

Members-b

Posts: 1,271
Joined: Jul 2012

Machine: Philips Remstar series 60
Mask Type: Nasal pillows
Mask Make & Model: ResMed P10
Humidifier: same as machine
CPAP Pressure: 5-15
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Switzerland

Post: #2
RE: System One no longer heating hose
Sounds like it doesn't recognise the hose - are you sure you have the correct heated hose, and not a standard hose? If you have been using the correct hoses all along, then the problem lies in the humidifier's connector to the hose, in all probability, and needs to be swapped out.
01-04-2015 05:36 AM
Find all posts by this user Post Reply Quote this message in a reply
Ghost1958 Offline

Advisory Members

Posts: 957
Joined: Aug 2014

Machine: PRS1 Series Sixty Auto Aflex
Mask Type: Nasal mask
Mask Make & Model: F&P Eson
Humidifier: PRS1 with Heated Hose
CPAP Pressure: 10.5 min 16 max
CPAP Software: SleepyHead EncoreBasic

Other Comments: Moderate to Severe SA Obstructive. Rapid O2 desat off the machine.

Sex: Male
Location: KY

Post: #3
RE: System One no longer heating hose
If your auto aflex is sensing the heated hose the little humidifier wave icons will be orange if not they will be blue.

If there are NO humidifier lights as in the little waves then your humidifier may not be functioning at all.

You can turn the heated hose off in clinical. If those little wave icons are blue that is likely what has been done either by accident or design. Otherwise the machine will automatically go into heated hose mode when the heated hose is connected.


Check the settings to see that heated hose is not turned off in clinical. If its on try separating the humidifier and blower units and putting them back together.
01-04-2015 08:40 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,389
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #4
RE: System One no longer heating hose
In the clinical menu, Heated tube humidifier is set to ON? Tubing type 15H?
If those options are grayed out the machine is not detecting the heated tube setup. You should pull the lid off and ensure the pins are engaging in the humidifier. If you are unsure how to do this, search for the heated tube upgrade kit installation instructions, or let me know and I can link you to that.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
01-04-2015 10:42 AM
Find all posts by this user Post Reply Quote this message in a reply
flibbidyfloo Offline

Members

Posts: 8
Joined: Jun 2013

Machine: REMStar Auto A-Flex
Mask Type: Nasal mask
Mask Make & Model: resmed mirage fx
Humidifier: system one
CPAP Pressure: 8-12
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #5
RE: System One no longer heating hose
Thanks for all the feedback!

When therapy is turned started, the humidifier wavy lines blink yellow for about 10 seconds then turn solid blue.

They also blink when I first plug in the house, then they turn off.

Clinician menu:
"Heated tube humidification" is OFF and shows lock symbol
"System one humidification" is ON
"Tubing type lock" is OFF
"Tubing type" says 15 and shows lock symbol
"Lock system" is OFF

The tube has raised letters that say SYS1HT15, which I'm guessing stands for heated 15?

I don't know if it makes a difference, but in the info menu the humidifier stats shows 60w, but the power supply says 80 watt.

I remember the heating was working last winter but I don't recall when it stopped because it wasn't cold until recently (California climate).

I went ahead and detached the humidifier unit from the main unit and checked the connectors. They look normal, but it's hard to tell much from that little plug. I put it back together, no change.
(This post was last modified: 01-04-2015 12:14 PM by flibbidyfloo.)
01-04-2015 12:13 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
bill-e Online

Advisory Members

Posts: 115
Joined: Mar 2014

Machine: Z1, Z1A, PR System One REMstar Auto CPAP Machine with A-Flex
Mask Type: Nasal mask
Mask Make & Model: F&P Eson
Humidifier: Respironics, HME, and F&P
CPAP Pressure: 14-20
CPAP Software: Other Software

Other Comments:

Sex: Male
Location: New Hampshire

Post: #6
RE: System One no longer heating hose
My tubing type is 15H with the lock symbol but when I remove the tubing it reverts to type 22 with no lock symbol.

See if it changes when you do that. you have to move the dial off Tube Type and back again for it to show the change.
(This post was last modified: 01-04-2015 01:12 PM by bill-e.)
01-04-2015 01:11 PM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,389
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #7
RE: System One no longer heating hose
The power for the heated hose actually routes through the back of the hinged lid of the humidifier. If you pull out the water tray, you can lift the lid up and forward and the lid will detach. A wire and plug from the left-rear of the lid, connects to pins at the top left of the humidifier to provide power for the heated hose. Be sure that is properly connected.

Instruction PDF

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
01-04-2015 01:34 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
flibbidyfloo Offline

Members

Posts: 8
Joined: Jun 2013

Machine: REMStar Auto A-Flex
Mask Type: Nasal mask
Mask Make & Model: resmed mirage fx
Humidifier: system one
CPAP Pressure: 8-12
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location:

Post: #8
RE: System One no longer heating hose
(01-04-2015 01:11 PM)bill-e Wrote:  My tubing type is 15H with the lock symbol but when I remove the tubing it reverts to type 22 with no lock symbol.

See if it changes when you do that. you have to move the dial off Tube Type and back again for it to show the change.

Mine behaves the same way. ANd if I'm in the clinician menu, when I unplug the tube the "heated tube humidification" option disappears. So it seems like it's detecting the hose correctly. It's just that darn heating option is locked when the hose is plugged in.

Same result with multiple tubes.

(01-04-2015 01:34 PM)Sleeprider Wrote:  The power for the heated hose actually routes through the back of the hinged lid of the humidifier. If you pull out the water tray, you can lift the lid up and forward and the lid will detach. A wire and plug from the left-rear of the lid, connects to pins at the top left of the humidifier to provide power for the heated hose. Be sure that is properly connected.

I did this and the plug seemed good, but I went ahead and pulled it out and put it back in. No change in behavior though Sad
01-04-2015 02:25 PM
Find all posts by this user Post Reply Quote this message in a reply
bill-e Online

Advisory Members

Posts: 115
Joined: Mar 2014

Machine: Z1, Z1A, PR System One REMstar Auto CPAP Machine with A-Flex
Mask Type: Nasal mask
Mask Make & Model: F&P Eson
Humidifier: Respironics, HME, and F&P
CPAP Pressure: 14-20
CPAP Software: Other Software

Other Comments:

Sex: Male
Location: New Hampshire

Post: #9
RE: System One no longer heating hose
(01-04-2015 02:25 PM)flibbidyfloo Wrote:  It's just that darn heating option is locked when the hose is plugged in.
My heating option shows on and unlocked so based on the behavior of the hose when you plug/unplug it, I'd say the unit is not detecting the heater in the Humidifier and therefor is turning OFF the heated option for both the hose and the humidifier...is that correct, is the water getting hot?
01-04-2015 03:02 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Galactus Offline

Advisory Members

Posts: 1,388
Joined: Jan 2014

Machine: Phillips Respironics System One PRS1 DS760P Auto BiPAP
Mask Type: Nasal pillows
Mask Make & Model: ResMed Airfit P10 Nasal Pillow
Humidifier: System One Heated Humidifier + Climate Line
CPAP Pressure: E/I 13/20 PS 4/7 BiFlex 3
CPAP Software: SleepyHead EncoreBasic

Other Comments: Sleep can be a good thing.

Sex: Male
Location: USA

Post: #10
RE: System One no longer heating hose
Are you sure you are in the clinician setup and not just in the setup? Because the lock you speak of sounds like it is locked in system one setting in the clinician setup and that what you see is the lock in the setup settings.

To get to the clinician menu you still dial to setup, but just clicking in setup doesn't take you to clinician mode it takes you to user setup.

If everyone thinks alike, then someone isn't thinking.
Everyone knows something, together we could know everything.
01-04-2015 07:48 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Climate line hose vs mask - adapter to connect? Beej 4 132 Yesterday 05:57 PM
Last Post: Beej
  [CPAP] hose valve noise ArtistEd 7 265 12-05-2016 11:05 PM
Last Post: ArtistEd
  PR System One: C-Check vs Auto? Ailu 8 3,023 11-29-2016 02:04 PM
Last Post: TonyJ
  Climateline hose duffer54 4 338 11-05-2016 03:24 PM
Last Post: 0rangebear
  Heated hose check ST1300 7 589 10-30-2016 09:15 AM
Last Post: Sleeprider
  Water in hose after cleaning... JustMac 3 375 10-30-2016 06:51 AM
Last Post: OpalRose
  Philips system one noisy jpsantell 2 266 10-29-2016 10:54 AM
Last Post: robysue

Forum Jump:

Who's Online (Complete List)