Post: #21
RE: This is so frustrating....
Hi, Jen. Thanks for posting the graph. Yes, your machine still looks like something is wrong with it (or it has a new "feature" that we don't know about).

The machine is supposed to start at the pressure shown in the "minimum pressure" value in the machine's settings. It should only go lower than that value if you hit the "Ramp" button. I can see where the pressure dropped to 6.0 when you hit the ramp.

If the pressure starts off at 9.5, the auto-adjusting action of the machine should not cause it to wander down to 4.0 at any time.

To say that in the reverse way, if the minimum pressure is 4.0 as shown in SleepyHead, the pressure should start off at 4.0 and not at 9.5.

Do you have a copy of EncoreBasic, the program that your DME uses to look at your machine information? It might be good to look at a daily report for your machine with that program.

Alternately, you could print out the three daily graphs from SleepyHead and show them to the respiratory therapist at your DME. Mention the three pressure values: starts at 9.5, ramp set at 6.0, but drops as low as 4.0 during the night.


Other stuff:
If you are awake a lot with the machine on, then yes, that will make your
AHI lower.

Your leak rate doesn't look high except for the period between 3:30-4:30. The top line on the leak rate graph is the total air leaving the CPAP-human system. It includes air that goes out your mask vent plus any air that leaks out your mouth or between your mask/nose.

There is an equation for the amount of air that's supposed to go out the mask vent depending on what the machine pressure is. I'm not clear on all the details, but the machine uses that to figure out the unintentional leak rate (the bottom number on the Leaks graph). The bottom number is the one you want to look at.

Your machine has some liters/min value for leaks that it considers to be too high. SleepyHead also has a value. I'm not clear if it gets that value from the machine or if it uses a default value. My SleepyHead program has 24 liters/min set as the threshold value for "Large Leak". You can see the value SH is using if you look under "File - Preferences - CPAP".

Sorry to hear that your chin strap was uncomfortable.
06-27-2016 12:28 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
chill Offline

Advisory Members

Posts: 856
Joined: Mar 2016

Machine: DreamStation X500
Mask Type: Full face mask
Mask Make & Model: Walk-in Mask Closet
Humidifier: DreamStation
CPAP Pressure: 13-16
CPAP Software: SleepyHead

Other Comments: CPAP True Believer

Sex: Male
Location: Vancouver, Canada

Post: #22
RE: This is so frustrating....
I agree. I'd definitely take it in and ask. I've used three DreamStation machines, none behaved like this.
06-27-2016 01:08 PM
Find all posts by this user Post Reply Quote this message in a reply
jen61 Offline

Preferred Members

Posts: 26
Joined: Jun 2016

Machine: DreamStation Auto CPAP
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10
Humidifier: Built In
CPAP Pressure: 4-20
CPAP Software: SleepyHead

Other Comments:

Sex: Female
Location: Missouri USA

Post: #23
RE: This is so frustrating....
I do know that it is the newest model. When I got it a few weeks ago, she said they had just come out with it. As to whether it has a new feature that allows a pressure variation like that I don't know. I am definitely going to either call or take it in tomorrow to make sure it's doing what it's supposed to be doing.
06-27-2016 05:43 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
chill Offline

Advisory Members

Posts: 856
Joined: Mar 2016

Machine: DreamStation X500
Mask Type: Full face mask
Mask Make & Model: Walk-in Mask Closet
Humidifier: DreamStation
CPAP Pressure: 13-16
CPAP Software: SleepyHead

Other Comments: CPAP True Believer

Sex: Male
Location: Vancouver, Canada

Post: #24
RE: This is so frustrating....
I am pretty sure there is only one version/revision so far. I got mine in March. I had it at the hospital recently and the RT there had not even seen one before!
06-27-2016 06:02 PM
Find all posts by this user Post Reply Quote this message in a reply
green wings Offline

Advisory Members

Posts: 547
Joined: Feb 2016

Machine: Respironics System One RemStar Pro (460)
Mask Type: Nasal pillows
Mask Make & Model: AirFit P10
Humidifier: integrated humidifier
CPAP Pressure: 13.0 cm
CPAP Software: SleepyHead

Other Comments: using CPAP since Jan. 2016

Sex: Undisclosed
Location: US

Post: #25
RE: This is so frustrating....
Hi Jen. I just read in another thread about a feature called Opti-Start which exactly describes what your machine was doing. I'm going to link to that thread here so that you can read the explanation of the Opti-Start feature. It's in post #2 in the thread.
07-01-2016 09:00 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  Very frustrating scenario, any ideas? vig1116 19 2,644 08-06-2014 06:06 PM
Last Post: vig1116
  My ongoing (and frustrating) gripe - ImRichard 11 2,477 05-15-2014 10:21 PM
Last Post: Gabby
  Sort of Frustrating Paul Canaris 14 3,197 03-01-2013 08:20 PM
Last Post: cincysinner

Forum Jump:

Who's Online (Complete List)