New to CPAP. on it (with Auto-SV) for a little over a week. Doing well with it, sleeping 7 hours a night or so, and feeling better during the day.

I am coughing more though, and having mucus in my lungs. Can't figure out if it's too much moisture or not enough. I do have to wake up and take the mask off a few times a night to drink water, my mouth gets dried out.

I have my humidifier set at 2. Higher than that and it splashes water in my mouth at first.

Also, is there a difference from how much lower than me on the bed the machine is? I have it a little below my head level on a side table. Tried it much lower one night, like six inches above ground, that night I was more dried out, but could have been any of the medications I'm on too.

What are the variables here? And if I'm suddenly coughing a little bit during the day and have a bit of mucus in my lungs, is that too much humidity or not enough?

I live a mile above sea level, in a town that's very dry in the winter. Though the relativity humidity on weather.com says 69% (we had a lot of snow this week and it's very cold).

Any experience on any of this is appreciated.

Thank you!
Michael
12-16-2015 10:54 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
trish6hundred Offline

Advisory Members

Posts: 6,449
Joined: May 2012

Machine: Resmed S9 AutoSet for Her
Mask Type: Full face mask
Mask Make & Model: Fisher & Paykel Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 10 - 7-20 Cm H2O
CPAP Software: Not using software

Other Comments: I started CPAP in 2008. Totally blind since birth.

Sex: Female
Location: Missouri, USA

Post: #2
RE: Too much humidity or too little?
Hi MichaelWD64,

You might try turning your humidity up a notch and see if that helps. Hopefully, just turning it up one notch won’t splash you too bad.
You shouldn’t need to have your machine too low, just an inch or 2 below your bed should work fine.
Hang in there for more ideas.

trish6hundred
12-16-2015 11:54 PM
Find all posts by this user Post Reply Quote this message in a reply
MichaelWD64 Offline

Preferred Members

Posts: 24
Joined: Dec 2015

Machine: ResMed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: ResMed H51
CPAP Pressure: 14
CPAP Software: SleepyHead

Other Comments: Pre-CPAP AHI = 145. Sleep Apnea & Central Sleep Apnea. GERD. Severe Allergies- IgE over 2000.

Sex: Male
Location: Wyoming (5,200 feet above sea level)

Post: #3
RE: Too much humidity or too little?
(12-16-2015 11:54 PM)trish6hundred Wrote:  Hi MichaelWD64,

You might try turning your humidity up a notch and see if that helps. Hopefully, just turning it up one notch won’t splash you too bad.
You shouldn’t need to have your machine too low, just an inch or 2 below your bed should work fine.
Hang in there for more ideas.


Thank you Trish!
MWD
12-16-2015 11:55 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,995
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #4
RE: Too much humidity or too little?
(12-16-2015 10:54 PM)MichaelWD64 Wrote:  I have my humidifier set at 2. Higher than that and it splashes water in my mouth at first.

That's weird. The humidity setting has nothing to do with what happens at first. That setting just raises the temperature of the warming plate below the tank. Are you using the warm up feature? If so, that could cause it. Of it could be too much water in the tank or water left in the hose or mask from washing or from the night before.

Quote:What are the variables here? And if I'm suddenly coughing a little bit during the day and have a bit of mucus in my lungs, is that too much humidity or not enough?

That sounds like an upper respiratory issue, like from a cold. That's best treated with humid air. I would raise it to the highest you can go without getting rain out (condensation).

Another option is a heated hose. You can get one for your machine. Alternatively you can insulate the hose with a wrap.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-17-2015 12:07 AM
Find all posts by this user Post Reply Quote this message in a reply
MichaelWD64 Offline

Preferred Members

Posts: 24
Joined: Dec 2015

Machine: ResMed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: ResMed H51
CPAP Pressure: 14
CPAP Software: SleepyHead

Other Comments: Pre-CPAP AHI = 145. Sleep Apnea & Central Sleep Apnea. GERD. Severe Allergies- IgE over 2000.

Sex: Male
Location: Wyoming (5,200 feet above sea level)

Post: #5
RE: Too much humidity or too little?
(12-17-2015 12:07 AM)Sleepster Wrote:  
(12-16-2015 10:54 PM)MichaelWD64 Wrote:  I have my humidifier set at 2. Higher than that and it splashes water in my mouth at first.

That's weird. The humidity setting has nothing to do with what happens at first. That setting just raises the temperature of the warming plate below the tank. Are you using the warm up feature? If so, that could cause it. Of it could be too much water in the tank or water left in the hose or mask from washing or from the night before.

Quote:What are the variables here? And if I'm suddenly coughing a little bit during the day and have a bit of mucus in my lungs, is that too much humidity or not enough?

That sounds like an upper respiratory issue, like from a cold. That's best treated with humid air. I would raise it to the highest you can go without getting rain out (condensation).

Another option is a heated hose. You can get one for your machine. Alternatively you can insulate the hose with a wrap.

Not using the warm up feature. How do I acess it and what does it do?

I get rainout when I fill the humidifier to the "max" line. When I didn't fill it quite full one night, I ran out of water before I was done sleeping 7 hours, with the humidifier set on 4.

My doctor has prescribed a heated hose, at my request. I'm still waiting on insurance and delivery for it.

thank you,

MWD
12-17-2015 12:18 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleepster Offline
Wiki Editor
Moderators

Posts: 4,995
Joined: Feb 2012

Machine: ResMed AirCurve10 VAuto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: HumidAir and SlimLine Hose
CPAP Pressure: MaxI 13.6 | MinE 5.2 | PS 4.4
CPAP Software: ResScan SleepyHead

Other Comments: Diagnosed Nov 2011. Conquered aerophagia.

Sex: Male
Location: Houston, Texas

Post: #6
RE: Too much humidity or too little?
(12-17-2015 12:18 AM)MichaelWD64 Wrote:  I get rainout when I fill the humidifier to the "max" line.

Only when you fill it to the max line? If so, I can't understand how that could be the cause of rain out.

Quote:When I didn't fill it quite full one night, I ran out of water before I was done sleeping 7 hours, with the humidifier set on 4.

If you get a room humidifier for your bedroom then your CPAP humidifier won't have to work as hard and it will use less water.

Sleepster
Apnea Board Moderator
www.ApneaBoard.com


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-17-2015 01:52 AM
Find all posts by this user Post Reply Quote this message in a reply
MichaelWD64 Offline

Preferred Members

Posts: 24
Joined: Dec 2015

Machine: ResMed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: ResMed H51
CPAP Pressure: 14
CPAP Software: SleepyHead

Other Comments: Pre-CPAP AHI = 145. Sleep Apnea & Central Sleep Apnea. GERD. Severe Allergies- IgE over 2000.

Sex: Male
Location: Wyoming (5,200 feet above sea level)

Post: #7
RE: Too much humidity or too little?
(12-17-2015 01:52 AM)Sleepster Wrote:  
(12-17-2015 12:18 AM)MichaelWD64 Wrote:  I get rainout when I fill the humidifier to the "max" line.

Only when you fill it to the max line? If so, I can't understand how that could be the cause of rain out.

Quote:When I didn't fill it quite full one night, I ran out of water before I was done sleeping 7 hours, with the humidifier set on 4.

If you get a room humidifier for your bedroom then your CPAP humidifier won't have to work as hard and it will use less water.

Thank you! Will do.
MWD
12-17-2015 01:58 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
OMyMyOHellYes Offline

Advisory Members

Posts: 915
Joined: Mar 2013

Machine: Respironics 560 Auto
Mask Type: Nasal mask
Mask Make & Model: Resmed Mirage FX
Humidifier: Respironics System One (no heated hose)
CPAP Pressure: 08.0-15.0 cm/H2O
CPAP Software: EncoreBasic

Other Comments:

Sex: Male
Location: Texas

Post: #8
RE: Too much humidity or too little?
I don't understand the "splashes water" part. Humidity settings have nothing to do with mechanical movement of liquid water. As Sleepster said, all it does is raise the temperature of the water, making it a little easier for the water to evaporate. If there is water in the tube/mask at start up, it is likely either condensation from the night before or water left from cleaning. Water level in in the tank, in and of itself, cannot cause condensation in the tube. And should not blow through. Just curious - do you put the mask on and then turn the machine on or do you turn the machine on and then put the mask on? But if it were water blown through, it would not matter what the humidification is. Another thought - during the day, is there a low spot where any condensation would form (ie, not drain back into the unit)? A humidifier temp too high that causes condensation that would collect in the tube from the night before?

What temperature do you keep the room at while you sleep? That will have a definite impact on condensation.

When you get the climate line tubing, you might try auto humidity. That doesn't set a humidity level so much as a temperature. The relative humidity delivered is a function of the temperature differential. Pick a temp that is a couple degrees higher than the overnight room temp. See if the water in the mask is not resolved then. Meanwhile, you could try wrapping the tube in a towel like a tube cozy to prevent condensation.

Position of the bed relative to the unit should not have any impact on the humidification. The only thing that does is keep condensation from running down into the mask. If you have condensation, the tube should be set so that it has a relatively straight run downhill from the mask to the unit, without dips where condensation can form.

OK on to the other part of the question regarding stuffiness. Try something one night. To some, it is counter intuitive. More humidity is not always the answer. I find that I have more sinus stuffiness and mucous most of the time if I increase humidity settings. If I set the humidifier settings on the PR 560 to "1", I am generally comfortable and use very, very little water (a tank can last me 3-4 days). If I set it at "2" I use more water and I wake up with a snoot full of snot. I once tired a "3" setting and it was uncomfortable to the point I was startign to feel suffocated.

Using a FFM I assume you are a mouth breather at night and I don't think the humidity settings are ever going to be enough to keep your mouth from drying out to a certain extent. And if could, it may be too much for your respiratory system comfort. I have a much easier time breathing outside on a cold, clear morning than I do in Houston in the summer.

Note that the RH reported by weather.com is outside at the cold temp, not inside heated air which is far drier.

Probably too much rambling before coffee.

Oh, is WD64 24 points better than WD40?

OMMOHY
(This post was last modified: 12-17-2015 06:21 AM by OMyMyOHellYes.)
12-17-2015 06:20 AM
Find all posts by this user Post Reply Quote this message in a reply
MichaelWD64 Offline

Preferred Members

Posts: 24
Joined: Dec 2015

Machine: ResMed S9 VPAP Adapt
Mask Type: Full face mask
Mask Make & Model: ResMed Quattro Air
Humidifier: ResMed H51
CPAP Pressure: 14
CPAP Software: SleepyHead

Other Comments: Pre-CPAP AHI = 145. Sleep Apnea & Central Sleep Apnea. GERD. Severe Allergies- IgE over 2000.

Sex: Male
Location: Wyoming (5,200 feet above sea level)

Post: #9
RE: Too much humidity or too little?
(12-17-2015 06:20 AM)OMyMyOHellYes Wrote:  I don't understand the "splashes water" part. Humidity settings have nothing to do with mechanical movement of liquid water. As Sleepster said, all it does is raise the temperature of the water, making it a little easier for the water to evaporate. If there is water in the tube/mask at start up, it is likely either condensation from the night before or water left from cleaning. Water level in in the tank, in and of itself, cannot cause condensation in the tube. And should not blow through. Just curious - do you put the mask on and then turn the machine on or do you turn the machine on and then put the mask on? But if it were water blown through, it would not matter what the humidification is. Another thought - during the day, is there a low spot where any condensation would form (ie, not drain back into the unit)? A humidifier temp too high that causes condensation that would collect in the tube from the night before?

What temperature do you keep the room at while you sleep? That will have a definite impact on condensation.

When you get the climate line tubing, you might try auto humidity. That doesn't set a humidity level so much as a temperature. The relative humidity delivered is a function of the temperature differential. Pick a temp that is a couple degrees higher than the overnight room temp. See if the water in the mask is not resolved then. Meanwhile, you could try wrapping the tube in a towel like a tube cozy to prevent condensation.

Position of the bed relative to the unit should not have any impact on the humidification. The only thing that does is keep condensation from running down into the mask. If you have condensation, the tube should be set so that it has a relatively straight run downhill from the mask to the unit, without dips where condensation can form.

OK on to the other part of the question regarding stuffiness. Try something one night. To some, it is counter intuitive. More humidity is not always the answer. I find that I have more sinus stuffiness and mucous most of the time if I increase humidity settings. If I set the humidifier settings on the PR 560 to "1", I am generally comfortable and use very, very little water (a tank can last me 3-4 days). If I set it at "2" I use more water and I wake up with a snoot full of snot. I once tired a "3" setting and it was uncomfortable to the point I was startign to feel suffocated.

Using a FFM I assume you are a mouth breather at night and I don't think the humidity settings are ever going to be enough to keep your mouth from drying out to a certain extent. And if could, it may be too much for your respiratory system comfort. I have a much easier time breathing outside on a cold, clear morning than I do in Houston in the summer.

Note that the RH reported by weather.com is outside at the cold temp, not inside heated air which is far drier.

Probably too much rambling before coffee.

Oh, is WD64 24 points better than WD40?

OMMOHY

Hi!

I turn the unit on, then put the mask on.

I keep the room coldish, like 65 when I go to bed, and it's probably 59 or 60 when I get up.

I'll try a humidity setting of 1 tonight. Thanks! I've tried 2, 3, and 4 before. 3 and 4 use up all the water before I get up, which ends up sort of cooking my lungs.

>Oh, is WD64 24 points better than WD40?

yes. Yes it is.

Thank you!

MWD
(This post was last modified: 12-17-2015 06:44 AM by MichaelWD64.)
12-17-2015 06:29 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
PoolQ Offline

Advisory Members

Posts: 986
Joined: Jul 2015

Machine: AirCurve 10 VAUTO
Mask Type: Full face mask
Mask Make & Model: Amara View
Humidifier: Integrated
CPAP Pressure: 6.2, 15, PS4.0
CPAP Software: SleepyHead

Other Comments:

Sex: Male
Location: Sillicon Valley

Post: #10
RE: Too much humidity or too little?
make sure the CPAP is level when in use so the water level is still at max and not above that by the bed.
Rainout takes a while. it's caused when warm, moist air goes through a colder tube and has to condense out, build up, and then get to the end of the hose.
I also need low to NO humidity or I get congested. I also set the temp at 63, cold for me is better than hot.
12-17-2015 03:00 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
Question Change Humidity Level Seasonally txpap 9 519 11-21-2016 06:58 PM
Last Post: Sleepster
  pressure vs EPR vs hypopnea vs humidity vs congestion Rick76599 6 400 10-15-2016 06:08 PM
Last Post: Rick76599
  Has anyone developed a cough from humidity being set too low? NickNick 7 555 09-13-2016 04:31 PM
Last Post: BadGoodDeb
  (Another Newbie) Mask Humidity janeaverage 4 462 09-11-2016 07:20 PM
Last Post: JudgeMental
  Humidity button gone chuck6161 5 337 07-31-2016 08:41 PM
Last Post: justMongo
  Please help! Humidity/Rainout questions pitt412 10 712 07-28-2016 03:35 PM
Last Post: OMyMyOHellYes
  System One Humidity Too much or Too little chriswhite 2 421 06-16-2016 04:59 PM
Last Post: trish6hundred

Forum Jump:

Who's Online (Complete List)