(08-30-2019, 09:36 AM)slowriter Wrote:
(08-30-2019, 09:26 AM)Sleeprider Wrote: You seem to benefit from pressure support, or in your case, EPR.  If you had an Aircurve 10 machine, you would have considerably more control over trigger and cycle sensitivity, minimum and maximum time of inspiration for IPAP, and of course a finer control of pressure support at 0.2 cm increments, rather than EPR settings of 1, 2 or 3.  I think if you have found a pressure that feels most comfortable to you, it might be worth trying an EPR setting of 2 to evaluate if flow limitation increases or CA events are decreased.

Thanks for the explanation.

Key point: I now (as of two days ago; see above) have an Aircurve 10 VAuto!

Here's the settings I have, with the ones marked with asterisk changes I made based on reading threads here.

Mode - S
EPAP - 7
IPAP - 12
Ti Max - 3.0*
Ti Min - 0.3
Trigger - Medium
Cycle - Very High*
Post Reply Post Reply
#62
RE: UARS and APAP
Titration study results (to eliminate the RERAs):

EPAP - 6
IPAP -12

I forgot to ask, but the tech didn't mention any centrals, so am pretty much assuming what the machine's been reporting there is not worth worrying about.

They did it on a Respironics System One which, interestingly to me, felt very different (much less "natural" transitions between IPAP and EPAP?) than the VAuto.
Post Reply Post Reply
#63
RE: UARS and APAP
I'll be curious how the new settings work for you. I'm not completely sure that the optimal settings on the PR machine will translate directly into the optimal settings on your ResMed machine, but it's good they actually titrated with RERAs in view. About the different feels of the brands: I had home titration on a PR machine and always felt it was trying to hustle into my next breath, but my keeper ResMed machine just followed my lead. Different and, for me, much superior algorithm. I look forward to hearing how things go with 6/12.
Post Reply Post Reply
#64
RE: UARS and APAP
(09-04-2019, 09:46 AM)Dormeo Wrote: I'll be curious how the new settings work for you.  I'm not completely sure that the optimal settings on the PR machine will translate directly into the optimal settings on your ResMed machine, but it's good they actually titrated with RERAs in view.  About the different feels of the brands: I had home titration on a PR machine and always felt it was trying to hustle into my next breath, but my keeper ResMed machine just followed my lead.  Different and, for me, much superior algorithm.  I look forward to hearing how things go with 6/12.

Exactly my impression as well.

And yeah, I was wondering how well the titrations transfer across machines.

I think the key point that I appreciated confirming is that I need a fair bit of PS to resolve the RERAs. I think I'll start (well, continue, as I've had success with this the previous couple of nights) at a PS of 5 tonight, because 6 felt a bit odd.

The tech that did this study was also the one that did the initial sleep study, and the first person to introduce me to UARS. So once I saw she was doing it, I figured she'd get it right.
Post Reply Post Reply
#65
RE: UARS and APAP
Once you post some charts, you ca fine-tune. I think the PS of 6.0 looks pretty good, and I don't have any problem with using S-mode. Be sure the option for Easybreathe™ is turned ON.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#66
RE: UARS and APAP
(09-04-2019, 04:00 AM)slowriter Wrote: Titration study results (to eliminate the RERAs):

EPAP - 6
IPAP -12

I forgot to ask, but the tech didn't mention any centrals, so am pretty much assuming what the machine's been reporting there is not worth worrying about.

They did it on a Respironics System One which, interestingly to me, felt very different (much less "natural" transitions between IPAP and EPAP?) than the VAuto.

Probably because Biflex was turned off on the Dreamstation.
Post Reply Post Reply
#67
RE: UARS and APAP
Yes. When I complained, she mentioned something about turning on biflex.
Post Reply Post Reply
#68
RE: UARS and APAP
Is S mode (6/12) w/Easy Breathe functionally equivalent to tightly-controlled VAuto (6 min EPAP with 6 PS and max IPAP of 12)?

Does appear that flow limitations aren't recorded in S, so that's one difference.

But beyond that?
Post Reply Post Reply
#69
RE: UARS and APAP
Below was the night before the study; an example where I slept pretty well. 

Following is last night, with settings derived from the titration study, where I didn't sleep very well.

Only significant difference is the PS (4 vs 6).

Just wondering on strategy; should I:
  1. stay with the higher number and expect to adjust over next few weeks, or ...
  2. ... work up to that gradually over a similar time-frame?
[attachment=15239]
[attachment=15240]
Post Reply Post Reply
#70
RE: UARS and APAP
BTW, I still think the CAs the machine is reporting are not significant. When I sleep through the night, they mostly go away (as with the first screenshot). When I don't, they come back, regardless of the pressure settings.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Sad [Treatment] Struggling to treat UARS with BIPAP Humancyclone7 6 353 28 minutes ago
Last Post: Humancyclone7
  Flow Limits How to Upgrade CPAP to Bilevel Deborah K. 110 8,701 6 hours ago
Last Post: ewriter
  New to BiPAP, hoping for titration assistance? (Probable UARS) Easing5319 15 1,112 Yesterday, 05:43 PM
Last Post: jkossis
  PSG Results: Could this be UARS? deebob 260 18,332 Yesterday, 01:48 PM
Last Post: Crimson Nape
  BiPAP Pressure for Possible UARS bertchintus 50 2,146 04-16-2024, 02:57 AM
Last Post: SingleH
  New Bilevel user help with settings (oscar attached) opo218 6 131 04-15-2024, 02:56 PM
Last Post: opo218
  Treating low grade CA with ASV tossinandturnin 17 437 04-13-2024, 08:52 PM
Last Post: jcp519


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.