RE: Using heated coil or not in Summer?
Thanks for the observations and comments. I read somewhere to set the humidifier on auto and it would adjust temperature based on ambient air conditions. We don't use AC nor Heat where I live and April and May can get close to 90 degrees. Add that to being at 5200 ft and it feels hotter. I just didn't want heated air but bet all I have to do is take it off Auto and adjust the temp to the minimum. Problem solved. Thanks
Barbara
Ajijic, Chapala, Jalisco, Mexico
Post Reply Post Reply
#12
RE: Using heated coil or not in Summer?
You can see the sensor in the end of the climateline hose. It protrudes into the airflow.
[Image: climate.line.heatedtube.forresmed.s9.jpg]
Admin Note:
JustMongo passed away in August 2017
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#13
RE: Using heated coil or not in Summer?
(04-08-2015, 08:07 PM)justMongo Wrote: You can see the sensor in the end of the climateline hose. It protrudes into the airflow.
[Image: climate.line.heatedtube.forresmed.s9.jpg]

Cool! Now we all know!
Barbara
Ajijic, Chapala, Jalisco, Mexico
Post Reply Post Reply
#14
RE: Using heated coil or not in Summer?
The purpose of the heated hose is not so much about a comfort factor or an air temp factor as it is about being a method to prevent humidified air from precipitating against a colder tube wall, which is what occurs in rainout. That is an apt name, as this is exactly what happens; it "rains" inside the tube (the temp differential converts suspended molecules to actual water drops) if there is a temp differential.

So whether you have a rainout, or whether it becomes important or not to have a heated hose, is basically dependent on the temp of the ambient air in your bedroom. Regardless of the season, if you have cooler air there, then there is probably more of a need for the heated hose. But most folks keep the room air about the same all year around.

Another hedge against rainout is a nice fleece hose cozy, which is also aesthetically welcome. IOW, the cozy provides some insulation meaning the heated hose may not be as necessary. Best 19 bucks I ever spent.
Post Reply Post Reply
#15
RE: Using heated coil or not in Summer?
(04-08-2015, 08:35 PM)TyroneShoes Wrote: The purpose of the heated hose is not so much about a comfort factor or an air temp factor as it is about being a method to prevent humidified air from precipitating against a colder tube wall, which is what occurs in rainout. That is an apt name, as this is exactly what happens; it "rains" inside the tube (the temp differential converts suspended molecules to actual water drops) if there is a temp differential.

So whether you have a rainout, or whether it becomes important or not to have a heated hose, is basically dependent on the temp of the ambient air in your bedroom. Regardless of the season, if you have cooler air there, then there is probably more of a need for the heated hose. But most folks keep the room air about the same all year around.

Another hedge against rainout is a nice fleece hose cozy, which is also aesthetically welcome. IOW, the cozy provides some insulation meaning the heated hose may not be as necessary. Best 19 bucks I ever spent.

Sounds like win win. I'll look into getting the hose cozy.
Barbara
Ajijic, Chapala, Jalisco, Mexico
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Using the heated tube on Dream Station 2 arpegio 4 950 02-16-2024, 06:52 PM
Last Post: arpegio
  Using cleaning brushes in heated tubing TiredTom111 9 604 01-29-2024, 08:30 PM
Last Post: TechieHippie
  ResMed AirSense 11 + ClimateLine 11 Heated Tubing + Oxygen YankeeLaker 2 355 01-22-2024, 08:17 PM
Last Post: YankeeLaker
  Do I have to "clean" my heated CPAP hose? Labromaniac 13 1,287 12-28-2023, 04:23 PM
Last Post: Palmtrees
  [Equipment] Airsense 11 Heated Hose MJLeveille 3 638 12-12-2023, 05:18 PM
Last Post: robbob2112
  GRRR. DME supplier shipped ClimateLineAir Heated Tubing in a BAG DaveInDenver 4 545 11-12-2023, 10:32 AM
Last Post: OpalRose
  Heated hose setting Uzupan 15 4,093 09-21-2023, 11:02 AM
Last Post: Lucky7


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.