I received your PM requesting help, and suggesting that you had used my recommendations to help setup your machine. As Bonjour said, I have never recommended that someone setup an BiPAP ST with EPAP 4, IPAP 15 (PS 11.0). The problem with your current settings is that the large pressure support (difference between IPAP and EPAP) will cause central apnea, and make you reliant on the timed trigger to IPAP. It is a very very poor choice for therapy! I actually think you should have received an auto CPAP or Auto bilevel (Vauto) rather than the ST. We can help you get the best therapy from your current machine which has many features of an ASV. Please clarify if your diagnosis included central apnea or just obstructive, and be sure to download a copy of the Resmed Aircurve 10 ST-A manual (I have included instructions below).
Go into your settings, and change the mode from ST to iVAPS. Use the following settings:
Mode: iVAPS
EPAP: 5.0
PS Min 4.0
PS Max 15.0
Enable iBR (intellegent breath rate)
Set Target PT Rate: 12.0
Target Va: 500 or use the Learning Mode below
Ti Min: 0.7
Ti Max 2.0
Rise Time: 300
Trigger: High
Cycle: Medium
Height: Set your actual height in cm
Mask: Full
Ramp: Off
Instead of setting the Target pt Rate and Target Va (breath rate and alveolar minute vent), you can use the following Learning Procedure. (This is from the clinical menu that you can obtain by sending an email to
: apneaboard@gmail.com and put
Setup Manual in the subject line. Then, write Resmed Aircurve 10 ST-A in the body of the email.)
This settting will provide dynamic pressure support to resolve your central apnea.
This is the titration protocol we will follow: