VPAP vs VPAP Auto
#1
Just got a S9 VPAP Auto after using a System One BiPAP. With the BiPAP I was getting AHI of between .11 and 1. The first day with the Auto I get 2.61. I also get a ton more info with the S9. Is it unusual to start off higher? And is that a bad first day? This machine is silent and the pressure more comfortable. I was getting leaks around my eyes every night before. This one nothing.
Post Reply Post Reply



Donate to Apnea Board  
#2
Usually the Auto machines do start with a lower pressure and increase as needed. This is part of why you are seeing less leaks and find it more comfortable. Also each manufacturer has a different way of responding to inhale and exhale cycles. Your 2.61 AHI is within the generally acceptable range. The important thing is how you feel in the morning.
Post Reply Post Reply
#3
(08-16-2013, 09:28 PM)tomko44 Wrote: Just got a S9 VPAP Auto after using a System One BiPAP. With the BiPAP I was getting AHI of between .11 and 1. The first day with the Auto I get 2.61. I also get a ton more info with the S9. Is it unusual to start off higher? And is that a bad first day?

Hi tomko44, welcome to the forum!

Your minimum pressure setting is just 4. I would suggest raising that to your median EPAP (exhale) pressure, or perhaps up to a few cmH2O below the 95 percentile EPAP pressure.

The problem with a "wide open" pressure range is the pressure will occasionally slowly drift way too low, perhaps resulting in additional obstructive events. This may be most of the reason for your higher AHI with the VPAP Auto. Another reason may be differences in criteria for when/how ResMed versus Philips Respironics score events.

Take care,
--- Vaughn
Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
Post Reply Post Reply
#4
(08-17-2013, 06:01 PM)vsheline Wrote:
(08-16-2013, 09:28 PM)tomko44 Wrote: Just got a S9 VPAP Auto after using a System One BiPAP. With the BiPAP I was getting AHI of between .11 and 1. The first day with the Auto I get 2.61. I also get a ton more info with the S9. Is it unusual to start off higher? And is that a bad first day?

Hi tomko44, welcome to the forum!

Your minimum pressure setting is just 4. I would suggest raising that to your median EPAP (exhale) pressure, or perhaps up to a few cmH2O below the 95 percentile EPAP pressure.

The problem with a "wide open" pressure range is the pressure will occasionally slowly drift way too low, perhaps resulting in additional obstructive events. This may be most of the reason for your higher AHI with the VPAP Auto. Another reason may be differences in criteria for when/how ResMed versus Philips Respironics score events.

Take care,
--- Vaughn
I set the EPAP min to 9 and will see what happens. Had another 2+ night, but a bunch were false as I was not asleep, I was in bed but didn't shut off the machine when I awoke.
Post Reply Post Reply
#5
(08-17-2013, 07:02 PM)tomko44 Wrote:
(08-17-2013, 06:01 PM)vsheline Wrote:
(08-16-2013, 09:28 PM)tomko44 Wrote: Just got a S9 VPAP Auto after using a System One BiPAP. With the BiPAP I was getting AHI of between .11 and 1. The first day with the Auto I get 2.61. I also get a ton more info with the S9. Is it unusual to start off higher? And is that a bad first day?

Hi tomko44, welcome to the forum!

Your minimum pressure setting is just 4. I would suggest raising that to your median EPAP (exhale) pressure, or perhaps up to a few cmH2O below the 95 percentile EPAP pressure.

The problem with a "wide open" pressure range is the pressure will occasionally slowly drift way too low, perhaps resulting in additional obstructive events. This may be most of the reason for your higher AHI with the VPAP Auto. Another reason may be differences in criteria for when/how ResMed versus Philips Respironics score events.

Take care,
--- Vaughn
I set the EPAP min to 9 and will see what happens. Had another 2+ night, but a bunch were false as I was not asleep, I was in bed but didn't shut off the machine when I awoke.

After raising the EPAP to 9 the AHI dropped down to .98. So your suggestion worked perfectly!
Post Reply Post Reply



Donate to Apnea Board  


Possibly Related Threads...
Thread Author Replies Views Last Post
  ResMed Vpap Auto Question... Please PositiveMe123 6 197 07-05-2018, 03:13 PM
Last Post: PositiveMe123
  VPAP - Trigger function can prevent CA's. Kosoku 16 591 06-21-2018, 11:57 AM
Last Post: alps
  [Equipment] ResMed AirCurve 10 VPAP help johnathonm 49 1,177 06-17-2018, 08:05 AM
Last Post: Sleeprider
Question Used S9 VPAP ST as a backup APAP DaveY 5 230 05-10-2018, 02:24 PM
Last Post: DaveY
  Help with Resmed S9 VPAP Adapt - it wont let me fall asleep! LuckyD 4 297 04-26-2018, 12:08 AM
Last Post: jaswilliams
  S-10 Auto Set Stop Auto Shutoff duyos 9 431 04-22-2018, 08:32 AM
Last Post: gwc2795
  Resmed S9 VPAP with COPD support for Mom silversnore 2 194 04-17-2018, 05:22 PM
Last Post: silversnore

Forum Jump:

New Posts   Today's Posts




About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.