@ sleeprider and Sheepless
Really interesting discussion, worth pursuing I think both wrt to Trigger and Cycle.
I am really trying hard to understand this, to make sense of what I am feeling and seeing in my results.
If I am out of line, just shout, no problems, and will let the subject go.
Firstly, from the Vauto Clinician manual:
The Vauto should not
initiate a breath, (unlike the ASV type machine, or a machine with ST mode).
Page 5 from the Clinician Manual " ... the device triggers (initiates IPAP) and cycles (terminates IPAP) and changes to EPAP) as it
senses the change in patient flow." [my italics]
From this, as far as I understand it, the Vauto waits for
me to initiate a breath, at which point this then triggers the IPAP. Would this be a correct conclusion?
This is in fact, what I see on my OSCAR chart of the 14th Jan.
The second CA on the night of 14th January lasted 32 seconds. The Vauto then continues to produce the FOT waiting for me to breathe. It makes no attempt to make me breath. It waits for me to build up my CO2 at which point I make an inspiratory effort and then the IPAP triggers. Have I got that right? (Later on in the sleep session I can see I am getting cyclical breathing as I hover around my apnea threshold.)
Important for me to fully understand this because if I can bring the IPAP forward (to the left with higher sensitivity) as per diagram on page 5 am I not also potentially altering I:E ratios, use of ventilatory "dead space" and tinkering with other stuff I don't fully understand?
@sheepless
thank you, no you are not distracting at all, it all helps.
I have noticed that the Vauto shows % spontaneous cycled breaths "live" while you look at the screen: from the manual this is averaged over the last 20 breathes only.
When I switch the machine off, this figure is different and I assume the figure is for the whole night, and closely matches that shown in Statistics in ResScan.