#21
RE: Wish to have zero pressure at exhaling
Hi Guy,


The seller is willing to give me a Resmed or a bi-level machine for a few days trail (not sure if one or both…). which one is preferred? Which setup should I use with the Resmed? The bi-level?
 
Meanwhile I proceed using the Prisma, playing with the setup parameters to see what happens, will update you.
 
Thank,
Arik
 
Post Reply Post Reply






Donate to Apnea Board  
#22
RE: Wish to have zero pressure at exhaling
ResMed AutoSet (Auto CPAP) 7-15 EPR 3
BiLevel EPAP Min 4, IPAP Max 15, PS 3

Both of these are similar to the current settings on your machine. Run these settings then post detailed results (Sleepyhead) and adjust from there.

Fred
Fred Bonjour - Project Manager and Lead Tester for OSCAR - Open Source CPAP Analysis Reporter 
OSCAR

Download OSCAR
New to Apnea? Helpful tips to ensure success
Soft Cervical Collar
Mask Primer
Dealing with a DME
Organize Charts
Attaching Charts

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#23
RE: Wish to have zero pressure at exhaling
(01-09-2019, 04:53 PM)bonjour Wrote: ResMed AutoSet (Auto CPAP) 7-15 EPR 3
BiLevel EPAP Min 4, IPAP Max 15, PS 3

Both of these are similar to the current settings on your machine.  Run these settings then post detailed results (Sleepyhead) and adjust from there.

Fred

Hi Fred,

Apparently, what I got from the seller is Resmed VPAP III ST with RS232 as an interface...

I found the user manual in the Internet. I can change EPAP and IPAP pressure and mode of operations (CPAP, Spontaneous, Spontaneous/timed and Timed). Also I have to tell the machine respiratory rate and some other factors that I'm not familiar with. PS3 is not found in the menus.
Anyway, I'll give it a try. The disadvantage is that I have no idea how to get the clinical data out of it. I'll judge it according to my feelings the day after.
If you have any idea, please let me know. I'll keep you updated on the results.
Thanks,
Arie
Post Reply Post Reply
#24
RE: Wish to have zero pressure at exhaling
That Resmed S8 VPAP III ST will not produce useful data. It is a fixed bilevel machine that has a timed backup mode if you use ST mode, or normal bilevel in S mode. To enter the clinical menu, press the right and down arrows simultaneously for about 3 seconds until the word "clinical" appears. Select Menu, then use the left and right arrows to cycle through the settings. Use the up and down arrows to make adjustments. The clinical manual is available from Apnea Board by requesting the ResMed VPAP III & III-ST. Send an email to: apneaboard@gmail.com and put "Setup Manual" (without the quotes) in the subject line. Then, in the body of your email put ResMed VPAP III & III-ST. The manual will be emailed.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply






Donate to Apnea Board  


Possibly Related Threads...
Thread Author Replies Views Last Post
  Is expiratory pressure relief similar to BiPAP on lower pressure settings? drmaestro 11 251 05-18-2020, 01:35 PM
Last Post: bonjour
  [Pressure] Difference between EPAP set pressure and operational pressure brugerard 8 467 01-02-2020, 10:03 PM
Last Post: brugerard
Question 90% Pressure meaning & Setting Correct Pressure for Auto-CPAP dk2011 2 1,318 12-28-2018, 05:37 PM
Last Post: bonjour
  Pressure vs Mask Pressure paulag1955 8 1,972 08-21-2018, 09:31 AM
Last Post: Sleeprider
  Isn't Resmed's Expiratory Pressure Relief really Inspiratory Pressure Support?? Reznik 116 8,043 11-27-2017, 07:28 PM
Last Post: Reznik
  With a face mask pressure is about 8.5. With a nasal pillow pressure was 20 at one po brojohn1611 10 2,277 11-04-2017, 03:05 AM
Last Post: James2017
  [Equipment] exhaling thorugh pillows confusedcpap 13 1,561 11-01-2017, 09:21 AM
Last Post: sonicboom


New Posts   Today's Posts






About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.