Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Woofer777 - OSCAR Analysis
#51
RE: Woofer777 - OSCAR Analysis
I've completed the first phase of testing at Fixed Pressure = 9 and EPR = 2.  The 4 day averages by events are:

CA = 6.6  OB = 2.8  HY = 1.7  RE = .1  AHI = 11.1   Results varied from AHI = 9 to 14.  

In general, sleep has been far more comfortable with the lower pressure level (compared to 12 and higher upper limits).   Unfortunately, overall numbers are not better -  with slightly higher OB.  Centrals remain high.  As a further note, I am unable to perceive any differences in quality of sleep night to night, even though numbers have varied. 

Tonight  (and for 4 nights) , Fixed Pressure = 7 and EPR = 0.  Unless, based on above results, a different test settings would be recommended?
Post Reply Post Reply






Donate to Apnea Board  
#52
RE: Woofer777 - OSCAR Analysis
Yeah try 0 EPR now.
Post Reply Post Reply
#53
RE: Woofer777 - OSCAR Analysis
I've completed the second test with fixed pressure 7 and EPR 0.   Centrals and obstructive events improved some while Hypopneas and RERA increased significantly.

Below is the average results from each test so far.

Fixed 9 EPR 2: CA = 6.6  OB = 2.8  HY = 1.7  RE = .1  AHI = 11.1   Results varied for AHI = 9 to 14.  4 nights

Fixed 7 EPR 0: CA = 4.4  OB = 1.8  HY = 6.7  RE = 1.7  AHI = 13.1   Results for AHI varied = 11 to 16.  3 nights

If screenshots for a typical night on either test above would be helpful let me know.

Sleeprider/Geer1 - Thoughts on next steps?
Post Reply Post Reply
#54
RE: Woofer777 - OSCAR Analysis
I would think the fixed pressure 9 with EPR 2 was much more comfortable, in spite of higher CA, than the higher hypopnea and RERA. Going back to my reply where we started this fixed pressure, I think the next step is to try lower pressure with EPR 2. Let's start dropping the fixed pressure in 1-cm increments to 8.0, then 7.0, but keep the EPR. http://www.apneaboard.com/forums/Thread-...#pid438288
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply






Donate to Apnea Board  
#55
RE: Woofer777 - OSCAR Analysis
Yeah so mild improvement of CA with no EPR but restrictions worsened as would be expected. As Sleeprider said now we pick a setting to use (I agree with EPR 2 as long as that felt better to you) then try different pressures to see what effects they have. Can try both lower and higher pressures. Lower may help with CA but if you go too low obstruction/restriction will get worse. Higher might help the obstruction and restriction present but might make CA worse.
Post Reply Post Reply
#56
RE: Woofer777 - OSCAR Analysis
I will next test Fixed Pressure 8 and EPR 2.

Yes, the higher levels of hypopnea and RERA were more uncomfortable with EPR 0.  (Although, not nearly the discomfort as when I had sustained high pressure levels when using a variable range with higher upper limits).

This is a slow process, however,  it does allow a more systematic approach for the data collection. 

I'll report back here with the next test results.
Post Reply Post Reply
#57
RE: Woofer777 - OSCAR Analysis
I have tried different settings during the last few weeks, with the summarized results below.


       Fixed 9 EPR 2: CA = 6.6  OB = 2.8  HY = 1.7   RE = .1    AHI = 11.1   Results varied for AHI = 9 to 14.  4 nights


       Fixed 7 EPR 0: CA = 4.4  OB = 1.8  HY = 6.7  RE = 1.7    AHI = 13.1   Results for AHI varied = 11 to 16.  3 nights

       Fixed 8 EPR 2: CA = 7.9  OB = 5.8  HY = 1.9    RE = .9    AHI = 15.6

     Fixed 10 EPR 2: CA = 7.9  OB = 1.5  HY = 1.4    RE = .0    AHI = 10.9   

Var 10 to 11 EPR 3: CA = 6.3  OB = 3.3  HY = 1.3    RE = .0    AHI = 11.1

   Fixed 10 EPR 3 CA = 5.1  OB = 3.0  HY = 1.0  RE = .0     AHI = 8.9   4 night average


From what I can observe, the fixed 10 EPR 3 has provided the best results so far, and is the setting I have used for the last 4 nights. When I increased the pressure to  variable 10 to 11, it did not have any effect on the obstructive events.    The pressure when set at 10 to 11 stayed closer to 11 during the night, but did drop down below at various times. 

I have felt the best during the day with current settings and seem to have decent sleep experience during the night. (i.e. breathing is comfortable, and no issues with mask).

Ob events are occurring together with CA events typically, and still mostly clustered in 2 to 3 main clusters. 

I'm inclined to stay at current setting for another week before possibly trying Fixed 10 EPR 2 again,  for more extended time period as that has the lowest OB and low HY.  

I would be interested in comments and interpretations of the data from above test settings.  If screenshots for any of above are desired, let me know.
Post Reply Post Reply






Donate to Apnea Board  
#58
RE: Woofer777 - OSCAR Analysis
Instead of trying 10/2 again I would try 11/3 for multiple nights. It will give you the same EPAP as 10/2 and help draw more conclusions on whether higher pressure helps. Then depending on results could try 12/3 to further figure out if higher pressure helps.
Post Reply Post Reply
#59
RE: Woofer777 - OSCAR Analysis
I will stick with 10/3 for the next week.   Do you feel that 11/3 might be different than 10-11 /3 ?  I've tested that level and it was less effective than 10/3, and more difficult to tolerate while sleeping even though only 1 higher.

The other thought I had was on the 10/2 results.   If centrals from my data are not controlled much by various settings, and not treated with my current machine, wouldn't the low O and H combined ( less than 3) , be considered a success?   At least until I get overnight lab study in a couple months? 

Stated slightly different, should my goal be to arrive at a setting where O and H combined are at lowest levels independent of centrals?   And then see if centrals gradually go away?
Post Reply Post Reply
#60
RE: Woofer777 - OSCAR Analysis
I wasn't sure how many days you tried that setting (sounded like only one) and it is possible you react to pressure changes so a fixed pressure is more interesting to me to be able to compare easily. What I usually do is vary EPR to find which seems to be best. Vary pressure to see effects of it and find best range. Then try a range vs fixed pressure to see if it is advantageous. Sometimes you have to revisit EPR if the pressure has a noticeable effect.

Overall there isn't an obvious trend in that data but part of the reason is because of multiple variables being adjusted and because you have two complications known to easily affect results (central apnea and positional apnea). The best indication on pressure effects is if you look at EPR 2 results (since it is the only EPR you really tried at multiple pressures) and there was a reduction each time you increased the pressure. Your worse results were at the lowest EPAP (6 cm when using 8/2 setting) in which both your obstructive apnea and central apnea were worse.

Obstructive apnea can be the cause of central apnea in the same way that central apnea can be repetitive. The obstruction causes large amplitude recovery breaths which wash out CO2 which triggers a central apnea and then the central apnea might cause a few more before breathing returns to normal.
Post Reply Post Reply






Donate to Apnea Board  


Possibly Related Threads...
Thread Author Replies Views Last Post
  ASV user request for analysis Labromaniac 40 1,041 06-15-2022, 06:50 PM
Last Post: Labromaniac
  [split] New member - Analysis Request PLM, limb movement arousals Eddi3p 9 431 04-13-2022, 01:12 PM
Last Post: Geer1
  Timmo - Therapy Analysis Timmo 9 556 04-09-2022, 07:45 PM
Last Post: Timmo
  sfr79's Treatment Thread // Analysis, Opinions, Suggestions Welcomed sfr79 7 509 02-14-2022, 01:54 PM
Last Post: Dormeo
  koy23 - Flow Rate Analysis koy23 55 2,748 02-08-2022, 06:02 PM
Last Post: koy23
  [CPAP] Need help with OSCAR analysis neostarsky97 0 232 02-06-2022, 12:41 AM
Last Post: neostarsky97
  Data Analysis brownrabbit34 9 525 01-28-2022, 07:21 PM
Last Post: Dormeo


New Posts   Today's Posts






About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.