Well after 6 months of therapy with the CPAP and input from wonderful individuals on this board, a recent visit with my sleep doctor has resulted in him replacing my cpap machine with a bipap machine. Apparently, the cpap machine was not doing the trick for me. After trying different masks, and narrowing pressure ranges I still found myself with AHI's of 8-10, and centrals of 6-9. He has suggested that I try the bipap machine for at least 3 months. I was just getting to know the airsense 10 machine now I have to start all over with another machine.

Does anyone have any comments or opinions regarding the resmed bipap machine??? anything I should be aware of.
12-20-2015 11:10 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Gru. Offline

Members

Posts: 9
Joined: Jul 2015

Machine: S9 Vpap auto
Mask Type: Nasal mask
Mask Make & Model: resmed mirage FX
Humidifier: H5i
CPAP Pressure: E. 13.5 / I. 15.5
CPAP Software: SleepyHead

Other Comments:

Sex: Undisclosed
Location:

Post: #2
RE: going from cpap to bipap
Quote:
Does anyone have any comments or opinions regarding the resmed bipap machine??? anything I should be aware of.

I found the bipap way more comfortable than cpap with the relief pressure on exhalation. Only thing I've found that is sometimes bothersome is the mask will sometimes pulse with the pressure going from exhalation to inhalation. Not a problem with nasal pillows and small nasal masks. Hope this helps a little.
12-21-2015 12:20 AM
Find all posts by this user Post Reply Quote this message in a reply
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,385
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #3
RE: going from cpap to bipap
ckingzzzs, be aware that you are currently in an insurance mandated progression, that first tried CPAP/APAP, and will now try the next level, BiPAP to resolve your AHI that is predominately central or complex in nature. Bilevel will be more comfortable for you, and it may help, but it's unlikely to resolve the problem because it still relies on your spontaneous initiation of a breath. If you continue to see high levels of central apnea, you should shortcut your doctor's recommendation of a 3-month trial, and get to the ASV therapy you probably need.

Best of luck, and hope it works for you. Don't hesitate to ask questions, or to question your doctor if you fail to see positive results from bilevel alone. BiPAP is a trademark of Respironics, and VPAP is the term used for bilevel at Resmed. You will get either an Aircurve 10 Vauto or the 10S version which is not auto adjusting. You will probably end up with the Aircurve 10 ASV unless you for some reason respond to bilevel. Also, with an AHI less than 15, insurance may not allow the ASV level since it is about twice as expensive as VAuto.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
(This post was last modified: 12-21-2015 11:20 AM by Sleeprider.)
12-21-2015 11:10 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
ckingzzzs Offline

Preferred Members

Posts: 57
Joined: Aug 2015

Machine: resmed aircurve 10 VAUTO
Mask Type: Nasal pillows
Mask Make & Model: swift fx nasal pillows
Humidifier: resmed aircurve10 vauto
CPAP Pressure: min epap 7, max ipap 20, PS 4
CPAP Software: ResScan SleepyHead

Other Comments: CPAP user June-Dec 2015, VPAP starting Jan 2016 and struggling with proper adjustments

Sex: Male
Location: maine

Post: #4
RE: going from cpap to bipap
(12-21-2015 11:10 AM)Sleeprider Wrote:  ckingzzzs, be aware that you are currently in an insurance mandated progression, that first tried CPAP/APAP, and will now try the next level, BiPAP to resolve your AHI that is predominately central or complex in nature. Bilevel will be more comfortable for you, and it may help, but it's unlikely to resolve the problem because it still relies on your spontaneous initiation of a breath. If you continue to see high levels of central apnea, you should shortcut your doctor's recommendation of a 3-month trial, and get to the ASV therapy you probably need.

Best of luck, and hope it works for you. Don't hesitate to ask questions, or to question your doctor if you fail to see positive results from bilevel alone. BiPAP is a trademark of Respironics, and VPAP is the term used for bilevel at Resmed. You will get either an Aircurve 10 Vauto or the 10S version which is not auto adjusting. You will probably end up with the Aircurve 10 ASV unless you for some reason respond to bilevel. Also, with an AHI less than 15, insurance may not allow the ASV level since it is about twice as expensive as VAuto.

sleeprider, thanks for your input I greatly appreciate it and your time. My doctor and I did discuss ASV as a possible resort, but he indicated mortality rate issues using ASV, and the fact that it is a very difficult device to get use to. Do you know what and how ASV works because I have a very limited amount of info and knowledge as to how ASV works. You are correct in that it appears I am having issues with central sleep apnea. Though I am really not sure I wish to pursue ASV as my centrals are 5-7.
12-22-2015 09:16 AM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,853
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #5
RE: going from cpap to bipap
In simple terms, the ASV keeps track of your respiration. It realizes when a breath is missing and provides pressure to inflate the lungs much in the same manner as mouth-to-mouth. I would say they are a little trickier to set up (titrate); but they are smart machines.

As for mortality, your doctor may be referring to this article:
http://www.resmed.com/us/en/consumer/new...erapy.html

excerpt:
"San Diego, Calif. – May 13, 2015 – ResMed (NYSE: RMD) today announced that SERVE-HF, a multinational, multicenter, randomized controlled Phase IV trial did not meet its primary endpoint. SERVE-HF was designed to assess whether the treatment of moderate to severe predominant central sleep apnea with Adaptive Servo-Ventilation (ASV) therapy could reduce mortality and morbidity in patients with symptomatic chronic heart failure in addition to optimized medical care.

The study did not show a statistically significant difference between patients randomized to ASV therapy and those in the control group in the primary endpoint of time to all-cause mortality or unplanned hospitalization for worsening heart failure (based on a hazard ratio HR = 1.136, 95 percent confidence interval [95% CI] = (0.974, 1.325), p-value = 0.104). The results from SERVE-HF are preliminary and will be submitted for future publication after further analysis.

A preliminary analysis of the data identified a statistically significant 2.5 percent absolute increased risk of cardiovascular mortality for those patients in the trial who received ASV therapy per year compared to those in the control group. In the study, the cardiovascular mortality rate in the ASV group was 10 percent per year compared to 7.5 percent per year in the control group. There were no issues associated with the performance of the ASV therapy device in the trial."
Fair use, educational non-profit. 17 USC 107

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
(This post was last modified: 12-22-2015 12:01 PM by justMongo.)
12-22-2015 11:59 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
Sleeprider Online
Wiki Editor
Advisory Members

Posts: 3,385
Joined: Dec 2014

Machine: Resmed Aircurve 10 Vauto
Mask Type: Nasal pillows
Mask Make & Model: Resmed Airfit P10
Humidifier: Resmed Climateline
CPAP Pressure: Auto Bilevel 18/9, PS 3
CPAP Software: ResScan SleepyHead EncoreBasic

Other Comments:

Sex: Male
Location: Where they make Respironics

Post: #6
RE: going from cpap to bipap
I know of a number of members that use ASV, including some with no central apnea problems...they just like it. There was a recent study by Resmed that indicated an increased mortality rate for severe chronic heart failure patients with ASV. Here is a link to the the study, and the most updated study conclusions and ongoing research status: STUDY Ongoing Research

The take-away is that while a specific population with severe CHF displayed a possible (and unexpected) adverse affect, this conclusion would not be applicable outside of that cohort of CHF patients. Adaptive Servo Ventilators serve as non invasive, positive pressure bilevel devices with an added algorithm that analyzes a patient's respiratory rate and volume, and can respond on a breath by breath basis, when needed, to provide additional inspiratory pressure to induce a breath. They have gotten so good with the auto algorithm that it is actually pretty easy to provide a machine to a person, and it will reliably provide good and comfortable therapy with the default settings. I have personally seen very severe complex apnea resolved with ASV, and have also seen people with mild occurrence of centrals and OA have near-zero AHI using these machines. About 15% of the population using CPAP would benefit from ASV. So it may not be as unusual as some think.

______________________________________________
Organize your SleepyHead Data
Post your SleepyHead Data from Imgur
Robysue's Beginner's Guide to Sleepyhead
12-22-2015 12:06 PM
Find all posts by this user Post Reply Quote this message in a reply
justMongo Offline

Monitors

Posts: 3,853
Joined: Sep 2013

Machine: ResMed VPAP Auto (S9)
Mask Type: Full face mask
Mask Make & Model: ResMed Mirage Quattro
Humidifier: ResMed H5i
CPAP Pressure: 18 IPAPmax - 11 EPAPmin, PS=5
CPAP Software: Other Software

Other Comments: μολὼν λαβέ

Sex: Male
Location: Калифорния

Post: #7
RE: going from cpap to bipap
^ What he said.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
12-22-2015 12:20 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
brianwood619 Offline

Advisory Members

Posts: 171
Joined: Jan 2014

Machine: ResMed S9 VPAP Auto
Mask Type: Full face mask
Mask Make & Model: F&P Simplus
Humidifier: H5i Heated Humidifier
CPAP Pressure: 16-22cm/h2o
CPAP Software: ResScan SleepyHead

Other Comments:

Sex: Male
Location: Peaceful Acres

Post: #8
RE: going from cpap to bipap
I was switched to BiPap and it was the best thing that's happened to me since I've been on therapy. So much more comfortable, and my AHI consistently stays under 1 each night. You may find it easier to adapt to BiPap than cpap. Good luck
12-23-2015 02:03 AM
Find all posts by this user Post Reply Quote this message in a reply
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #9
RE: going from cpap to bipap
Study results: http://www.resmed.com/us/dam/documents/serve-hf/Benelux-hcp.pdf

The devices used in the SERVE-HF trial used fixed EPAP, ResMed's older "ASV" therapy mode rather than their newer "ASV Auto" mode, but ResMed believes that automatic EPAP may also increase cardiovascular risk in patients with symptomatic chronic heart failure with reduced ejection fraction (LVEF<45%).

Although the small increase in risk (10% versus 7.5%) might have been a statistical fluke, the study was large enough (1,325 patients) that it is most likely correct.

But even the small increase in risk does not apply to us if we do not have symptomatic Chronic Heart Failure with Left Ventricular Ejection Fraction 45% or worse.

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 12-29-2015 03:17 PM by vsheline.)
12-29-2015 02:39 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #10
RE: going from cpap to bipap
(12-22-2015 12:06 PM)Sleeprider Wrote:  I have personally seen very severe complex apnea resolved with ASV, and have also seen people with mild occurrence of centrals and OA have near-zero AHI using these machines. About 15% of the population using CPAP would benefit from ASV. So it may not be as unusual as some think.

Before switching to an ASV machine I used a standard bilevel Auto, the S9 VPAP Auto. Although I much preferred the easiness of breathing with Pressure Support set to 4 or 5 (and the higher PS also beneficially boosted my SpO2), I found that PS higher than about 0.6 increased the number of Cental Apneas I would get. (Having a few short CA events, though, is no cause for concern.)

I no longer take my backup machine with me when traveling; instead I take my ASV with me, because I feel significantly better on ASV than when on APAP or standard bilevel. Every few months I'll have really huge Leak or while asleep will roll over onto my back, and will have some apneas or hypopneas. Other than that, week after week my AHI is always zero. And I can use all the Pressure Support I like.

When my present ResMed ASV machine wears out I will probably replace it with one from Phillips Respironics, because they are far more adjustable, far less one-size-fits-most. Unlike Philips Respironics ASV models, ResMed ASV models cannot be set up to work like an APAP or like a standard bilevel.

Also, in fixed-pressure CPAP therapy mode (their only non-ASV therapy mode) ResMed ASV machines cannot even distinguish between central apneas versus obstructive apneas. On an expensive machine specifically designed for patients with central apnea issues. Simply shameless.

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 12-29-2015 04:20 PM by vsheline.)
12-29-2015 03:44 PM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [Equipment] respironics bipap error code hanasdaddy 6 121 12-07-2016 01:48 AM
Last Post: hanasdaddy
  New User Problem with BiPAP AVAPS - Part 1 jmayer180 35 1,329 11-27-2016 09:50 AM
Last Post: jmayer180
  First time Bipap User, years of suffering SleepyMersim 18 668 11-16-2016 11:46 PM
Last Post: Galactus
Question BiPAP vs EPR/Flex etc? SleepyBee 7 275 11-13-2016 08:52 PM
Last Post: Galactus
  [Equipment] Annoying on/off beep on Respironics BIPAP S/T C series rcattell 8 350 11-06-2016 12:07 PM
Last Post: rcattell
  Blood sugar issues after treating apnea with BIPAP Gugliano 7 427 10-21-2016 01:59 PM
Last Post: southerndoc
  Can’t get to sleep with the BiPAP machine. StrugglinginOhio 26 1,545 10-20-2016 03:52 PM
Last Post: Sleeprider

Forum Jump:

Who's Online (Complete List)