Hi vsheline, thanks for the warm welcome.

It's too late! Have the 960 on its way to me now.

A few other things have changed since then too, like my methods to get a nasal mask working effectively and I will be a bit more bold with settings experimentation. Of course I would not fight against futility if the machine was that bad, but hopefully I'll 'adapt' Wink

Looking back through some of goodonya's experiences, it also seems the Respironics algorithm doesn't really suit his disposition. Not sure if my arousals were just more sudden PS related which might be an issue on Resmed too. My memory of the period isn't too sharp so will have to get back to basics.

Might just leave settings wide open for a bit and see what the machine thinks, I know setting the EPAP near my 95% the last time isn't necessarily a golden ticket as the 95% is just that pressure or below on Resmed I think, so could spend a lot of the night below that.
08-29-2014 11:29 AM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
zoophaugus Offline

Members

Posts: 16
Joined: Aug 2014

Machine: BiPAPautoSVADVSysOne60SrsCA
Mask Type: Full face mask
Mask Make & Model: Resmed Mirrage Quatro
Humidifier: System One HT Humidifier
CPAP Pressure: Epap 8-12 PS 4.5-10
CPAP Software: ResScan EncoreBasic

Other Comments: Started with cpap S9 elite 16 CmH20 , S9 Auto set 5-17CmH20, VPAP auto Epap 8-16, PS 4 CmH20, & now

Sex: Male
Location: (GTA) Greater Toronto Area, Ontario, Canada

Post: #12
RE: servo ventilation not working
Hey guys/gals ..thank you so much for the replies as they are a fountain of info to be absorbed and applied for all who read and their individual situations.

I did go ahead with the purchase of the Philips Respironics System One BiPap AUTO SV ADVANCE machine - as stated before it had more functionality or fine adjustments than the Resmed VPAP Adapt . Its ability the set a minimum Back up rate proved indispensable to me.

I will admit I am not "cured" but rather I appear to be stabilized from the downward spiral I was in.

As stated before i did tighten and continue to tweak the parameters of the machine.

My RX is Min/Max Epap 8-16 - Min/Max PS 3-15 - Backup rate Auto

I now have the machine set to epap = 8-12 And PS 4.5-10 with BR 16

I did notice some thing very interesting - leaving the epap at 8-12 and leaving the PS wide open I have the lowest AHI (below 1 average)

Leaving epap 8-12 and dropping PS 4.5-10 my AHI goes to around 3.

Leaving alone epap and further dropping PS to 4.5 - 7 or 8 my AHI goes to around 5ish or just a little over .

I actually feel better when the AHI rises a bit - having it too low means the machine is going ape sh*t trying to "fix" me causing more severe fatigue the next day - most likely disrupting slow wave sleep with the higher pressures.

I have a few more experiments I would like to try with the machine in the near future.
I think I may also be suffering from hypercapnia as well (high C02 levels in the blood stream) which explain the horrible way I feel a lot of times in the morning and trough out the day on top of the tiredness. I will try and set up the machine more as a basic ventilator with hopes of possibly blowing off more C02 during the inhale and exhale process - I am going to raise the PS to 8-10 and lower the Back up rate to 14 and increase the TI (inspiratory time) to allow for the larger breaths with hopes that it may make me feel better.

As Vsheline sated previously perhaps I may just need straight ventilation with minimum target ventilation.
08-29-2014 08:24 PM
Find all posts by this user Post Reply Quote this message in a reply
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #13
RE: servo ventilation not working
(08-29-2014 08:24 PM)zoophaugus Wrote:  I am going to raise the PS to 8-10 and lower the Back up rate to 14 and increase the TI (inspiratory time) to allow for the larger breaths with hopes that it may make me feel better.

Hi zoophaugus,

A minimum PS of 8 would be very high.

I would recommend buying a Pulse Oximeter to verify your average SpO2 (not counting the occasional dips during apnea events) will not be higher than around 96%. If using at least 8 for PS all the time, your average SpO2 might be dangerously high most of the night.

Until you are monitoring your SpO2, I suggest leaving Min PS at 5 or lower.

Take care,
--- Vaughn

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
08-29-2014 08:44 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
zoophaugus Offline

Members

Posts: 16
Joined: Aug 2014

Machine: BiPAPautoSVADVSysOne60SrsCA
Mask Type: Full face mask
Mask Make & Model: Resmed Mirrage Quatro
Humidifier: System One HT Humidifier
CPAP Pressure: Epap 8-12 PS 4.5-10
CPAP Software: ResScan EncoreBasic

Other Comments: Started with cpap S9 elite 16 CmH20 , S9 Auto set 5-17CmH20, VPAP auto Epap 8-16, PS 4 CmH20, & now

Sex: Male
Location: (GTA) Greater Toronto Area, Ontario, Canada

Post: #14
RE: servo ventilation not working
(08-29-2014 08:44 PM)vsheline Wrote:  
(08-29-2014 08:24 PM)zoophaugus Wrote:  I am going to raise the PS to 8-10 and lower the Back up rate to 14 and increase the TI (inspiratory time) to allow for the larger breaths with hopes that it may make me feel better.

Hi zoophaugus,

A minimum PS of 8 would be very high.

I would recommend buying a Pulse Oximeter to verify your average SpO2 (not counting the occasional dips during apnea events) will not be higher than around 96%. If using at least 8 for PS all the time, your average SpO2 might be dangerously high most of the night.

Until you are monitoring your SpO2, I suggest leaving Min PS at 5 or lower.

Take care,
--- Vaughn

You are right, caution is advisable and i will head your advise. I will leave things as they are a dig up my pulse oximeter before going ahead - I have a CMS50E.
Resmed makes a device that attaches inside of you full face mask and actually gives you and approximate reading of how much C02 is being exhaled. My "GUY" is
looking into it for me.
Accordinng to the machine data last night's average PS was 5.6 .
Min/max was 4.5 - 10.
08-29-2014 09:00 PM
Find all posts by this user Post Reply Quote this message in a reply
zoophaugus Offline

Members

Posts: 16
Joined: Aug 2014

Machine: BiPAPautoSVADVSysOne60SrsCA
Mask Type: Full face mask
Mask Make & Model: Resmed Mirrage Quatro
Humidifier: System One HT Humidifier
CPAP Pressure: Epap 8-12 PS 4.5-10
CPAP Software: ResScan EncoreBasic

Other Comments: Started with cpap S9 elite 16 CmH20 , S9 Auto set 5-17CmH20, VPAP auto Epap 8-16, PS 4 CmH20, & now

Sex: Male
Location: (GTA) Greater Toronto Area, Ontario, Canada

Post: #15
RE: servo ventilation not working
(08-29-2014 08:44 PM)vsheline Wrote:  
(08-29-2014 08:24 PM)zoophaugus Wrote:  I am going to raise the PS to 8-10 and lower the Back up rate to 14 and increase the TI (inspiratory time) to allow for the larger breaths with hopes that it may make me feel better.

Hi zoophaugus,

A minimum PS of 8 would be very high.

I would recommend buying a Pulse Oximeter to verify your average SpO2 (not counting the occasional dips during apnea events) will not be higher than around 96%. If using at least 8 for PS all the time, your average SpO2 might be dangerously high most of the night.

Until you are monitoring your SpO2, I suggest leaving Min PS at 5 or lower.

Take care,
--- Vaughn

Vsheline,

Im not sure what exactly happened but I seemed to have found a "sweet" spot with the machine - I Feel 10 times better than before - All I did was increase the PS from 4.5-10 to 4.5 to 11 - .
I stil get tired but have not used my modafonil for 2 whole days now.

I am some what confused as the "BiPAP Auto SV Advanced summary"
is stating
My AVERAGE PS is 6cmH20
90% of the time device PS was 9.1cmH20

If the average PS is 6, then how can 90% of time be 9.1 ?

I am almost certain that the higher pressure support is the direct cause of me feeling better.

I have yet to sleep with the pulse oximeter and check saturation levels.
09-04-2014 09:11 PM
Find all posts by this user Post Reply Quote this message in a reply

Donate to Apnea Board
vsheline Offline

Advisory Members

Posts: 1,908
Joined: Jul 2012

Machine: S9 VPAP Adapt (USA Model# 36007, not better 36037 or 36067)
Mask Type: Full face mask
Mask Make & Model: F10 or SimPlus w/ 2Liners. MirageQuatro & Gecko gel pad
Humidifier: H5i
CPAP Pressure: 15 EPAP, PS 5-10
CPAP Software: ResScan

Other Comments: Marfan Syndrome, chronic bradycardia, occasional Cheyne-Stokes Respiration

Sex: Male
Location: California, USA

Post: #16
RE: servo ventilation not working
(09-04-2014 09:11 PM)zoophaugus Wrote:  I Feel 10 times better than before - All I did was increase the PS from 4.5-10 to 4.5 to 11 - .
I stil get tired but have not used my modafonil for 2 whole days now.

I am some what confused as the "BiPAP Auto SV Advanced summary"
is stating
My AVERAGE PS is 6cmH20
90% of the time device PS was 9.1cmH20

If the average PS is 6, then how can 90% of time be 9.1 ?

I am almost certain that the higher pressure support is the direct cause of me feeling better.

You are also adapting to the therapy, and changing PS back to 4.5 to 10 might not make a lot of difference now. If swallowing too much air is a problem for you, lowering the Max PS might actually help, but I would recommend against lowering Max PS any lower than around 10 unless you are have problems (pain) from swallowing too much air.

The summary is reporting that the median Pressure Support is 6, which means that AT LEAST 50% of the time PS was 6 OR LOWER, and AT LEAST 50% of the time PS was 6 OR HIGHER.

The "90%" value is the 90 percentile value. It means PS was 9.1 OR LOWER at least 90% of the time. This also means PS was 9.1 OR HIGHER at least 10% of the time.

As an extreme example of medians and 90 percentiles, if PS was exactly 6 for 51% of the night and was exactly 9.1 for 49% of the night, the median PS would have been 6 (even though the average would have been higher than 6) and the 90% PS would have been 9.1 (even though PS would have been 9.1 longer than 10% of the time).

And if PS was exactly 6 for 49% of the night and was exactly 9.1 for 51% of the night, the median PS and the 90% PS would both have been 9.1.

Very glad things are going well for you.

Take care,
Vaughn

Membership in the Advisory Member group should not be understood as in any way implying medical expertise or qualification for advising Sleep Apnea patients concerning their treatment. The Advisory Member group provides advice and suggestions to Apnea Board administrators and staff on matters concerning Apnea Board operation and administrative policies - not on matters concerning treatment for Sleep Apnea. I think it is now too late to change the name of the group but I think Voting Member group would perhaps have been a more descriptive name for the group.
(This post was last modified: 09-05-2014 03:12 AM by vsheline.)
09-05-2014 01:44 AM
Find all posts by this user Post Reply Quote this message in a reply
Post Reply 


Possibly Related Threads...
Thread: Author Replies: Views: Last Post
  [Equipment] ResMed Airsense 10 Quit Working SherrieR 4 249 11-26-2016 07:51 PM
Last Post: trish6hundred
  [CPAP] Changes in how the machine has been working? Xinectyl 9 318 11-07-2016 12:02 PM
Last Post: Xinectyl
Gross [Equipment] Unit not working Monty6894 3 371 09-18-2016 06:52 AM
Last Post: cate1898
  [Equipment] Dreamstation Auto ON/OFF set but not working newb 11 2,198 09-14-2016 02:39 PM
Last Post: rkl122
  Remstar quit working ropdve 4 387 09-06-2016 05:11 PM
Last Post: cate1898
  [Treatment] Auto CPAP, high/erratic AHI, high Clear Apneas - it's not working jmortensenjr 10 768 08-30-2016 09:07 PM
Last Post: jmortensenjr
  DreamStation Humidifier working/testing? NorthernGuy 6 544 08-23-2016 09:43 PM
Last Post: justMongo

Forum Jump:

Who's Online (Complete List)