Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

[split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
#1
[split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
I have a RELATED PROBLEM.  SOMEHOW the EPR PATIENT ENABLE got turned off - I have gone into the CLINICAL MENU, but do not see how to turn it back on again.  Already in the ADVANCED - patient settings.   PLEASE HELP. 
Device Settings
Therapy Mode: AutoSet EPR: FULL_TIME EPR Setting: 3.0 cmH2O
EPR Enable: ON EPR Patient Enable: OFF Ramp Enable: RAMP_ON
Ramp Time: 20.0 Minutes PatientView: ADVANCED Response: STANDARD
Minimum Pressure: 4.2 cmH2O Maximum Pressure: 15.0 cmH2O

Best   Steve
              Sad
Post Reply Post Reply
#2
RE: How to access Airsense 10 Expiratory pressure relief (EPR™) settings
Your min pressure is set 4.2. The absolute lowest it can go is 4. No matter what you set the EPR to it can only reduce 0.2 cm. 

You need to set min to 7. EPR 3

Min7 - EPR3 = 4
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#3
RE: How to access Airsense 10 Expiratory pressure relief (EPR™) settings
NO,  THAT IS NOT THE PROBLEM.  Please Read above note again!  Problem is "the EPR patient enable is now turned off"  Was on a few days ago.
Need to get THAT turned back on...   Max pressure is 15...     Steve   Sad
Post Reply Post Reply
#4
RE: How to access Airsense 10 Expiratory pressure relief (EPR™) settings
Maybe that is not the only problem but it surely is a problem!
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#5
RE: How to access Airsense 10 Expiratory pressure relief (EPR™) settings
Stevekuk, we always access EPR On and the settings that by going through the clinical menu. Click the home button and control knob at the same time and enter the clinical menu. Scroll down to comfort settings and verify EPR is set On Full Time at the setting you choose. It is possible the patient menu for EPR is turned off, but you can still access it in the clinical menu.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#6
RE: [split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
stevekuk,
Since you are using the new A11, I have split your post from the 2017 thread discussing the A10 that you posted in.
OpalRose
Apnea Board Administrator
www.apneaboard.com

_______________________
OSCAR Chart Organization
How to Attach Images and Files.
OSCAR - The Guide
Soft Cervical Collar
Optimizing therapy
OSCAR supported machines
Mask Primer



INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE.  ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA.  INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#7
RE: [split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
Here is the link to instructions to access the Airsense 11 clinical settings.
https://www.apneaboard.com/resmed-airsen...setup-info

Sleeprider gave you instruction for the airsense 10
Download OSCAR

Organize Charts
Attaching Charts

Mask Primer
Soft Cervical Collar

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.

Post Reply Post Reply
#8
RE: [split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
I apologize for the post ---  I FIGURED IT OUT - SIMPLE STUPID MISTAKE -   In the patient menu is "pressure reduce"  It is NOT labeled EPR so I missed it.    The problem is fixed.  Sorry for all the fuss.  I even sent it to Univ Sleep center - but no reply...
(My post got moved - so apologize if 2 replies)
         Sleep Well!   I am not sleeping well - bad knee - APAP does not help that...     Best, Steve   [Image: thanks.gif] [Image: biggrin.gif]
Post Reply Post Reply
#9
RE: [split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
Thank you for the information. I know we will have many AS11 users looking for the same thing. Until the release of the AS11, it was listed as, "EPR".
Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#10
RE: [split] How to access Airsense 11 Expiratory pressure relief (EPR™) settings
(01-13-2022, 10:52 AM)Crimson Nape Wrote: Until the release of the AS11, it was listed as, "EPR".

To clarify, the AS11 still lists it as "EPR" on the Clinician menu.  On the Patient menu (where ON/OFF are only selections), it is listed as "Pressure Relief".   I guess Resmed thought patients would be confused by the abbreviation and didn't have room for all three words.  :-)

Clinician Menu:
   

Patient Menu:
   
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  nasal congestion relief sleepingbetterinFl 2 160 03-23-2024, 01:54 PM
Last Post: sleepingbetterinFl
  From DreamStation to AirSense 10 or AirSense 11 xaid 7 621 03-20-2024, 04:57 PM
Last Post: labrat
  Split night sleep study Bwanajohn51 0 68 03-20-2024, 01:07 PM
Last Post: Bwanajohn51
  [Pressure] Understanding Pressure Settings on ASV Suzee 85 3,574 03-19-2024, 03:17 PM
Last Post: Suzee
  AirSense 10 Max Pressure setting Tomt 13 859 03-04-2024, 07:11 PM
Last Post: Tomt
  [split] Doctor won't help improve Vauto settings JohnCondor 7 527 02-19-2024, 04:20 PM
Last Post: SarcasticDave94
  [Pressure] unable to find pressure settings on my cpap timoteo666 7 450 02-17-2024, 09:56 PM
Last Post: OpalRose


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.