Hello,

I am inputting the IPAP/EPAP/PS, etc settings on my new ResMed AirCurve 10 VAuto (purchased through a private seller, 0 hours), and I am not sure what my settings should be set to. The original prescription from last year's sleep study called for EPAP of 10 and IPAP of 15, and nothing else on the prescription. (I will be getting another copy of my sleep study tomorrow to help ascertain what my settings should be, AHI, and all that). 

So far I have the settings as:

Mode - VAuto
Max IPAP - 15
Min EPAP - 10
PS 4.0
Ti Max - 2.5 s
Ti Min - 1.0 s
Trigger - Med
Cycle - Med
Mask - Full Face

This is my first time using an auto bipap. Should I set the Max IPAP and Min EPAP to different (higher IPAP/lower EPAP) values to allow the bipap to increase or decrease within a larger threshold as needed (I'm assuming from what I've read that the device will monitor my apneas and adjust settings to reduce apneas). What is PS and what should it be at? I will adjust Ti Max and min on the fly, and what are trigger and cycle? 

Finally, I know this isn't the place for medical diagnoses or advice in place of a doctor, but I have Kaiser insurance and they're pretty much leaving me to fend for myself in terms of treating my sleep apnea, even with what I thought was a good plan (Platinum from ACA). I have tried setting an appointment with a sleep doctor or therapist the last few days with no updates or calls back. Besides I'd like to educate myself on how to properly monitor my own therapy. 

Thanks, Jio


RE: How to properly set VAuto clinician settings? - justMongo - 02-08-2017

The original Rx was for fixed bilevel with IPAP=15 and EPAP=10. That implies a PS=5 (IPAP - EPAP = PS)
Consider keeping the PS of 5. Give you machine a little upper room to adjust. Perhaps Max IPAP =17.

You could consider backing off on the Min EPAP a bit since part of the night you may be good with a lower pressure.
Your choice -- but, consider Min EPAP of 9. Run a couple of nights and adjust.


RE: How to properly set VAuto clinician settings? - Jioaa1988 - 02-08-2017

I'll give it a go. Thanks JM

If I post sleepyhead data in the future, is it possible to seek help for advice on adjustments later on this forum?


RE: How to properly set VAuto clinician settings? - justMongo - 02-08-2017

(02-08-2017, 01:04 AM)Jioaa1988 Wrote: I'll give it a go. Thanks JM

If I post sleepyhead data in the future, is it possible to seek help for advice on adjustments later on this forum?

Certainly.  We encourage it.

Kind Regards,  Mongo


RE: How to properly set VAuto clinician settings? - Sleeprider - 02-08-2017

The original prescription was applicable to a fixed bilevel. You can get some additional flexibility, and still use the PS of 5 by setting EPAP min to 8 or 9, IPAP max to 18, and setting PS to 5. The machine will self-titrate EPAP to prevent obstructive events. Presumably the pressure support is for suppression of hypopnea and ventilation support, and is fixed on the Resmed Vauto. If you experience excessive CA then reducing the pressure support will be a logical response.