Expiratory Pressure Relief setting for ResMed AirSense 10 Autoset manual says "When EPR is enabled, the delivered pressure will not drop below a minimum pressure of 4 cm H 2 O, regardless of the settings." so what I don't understand is the EPR levels of 1, 2, or 3 cm H2O? If my regular pressure range is 5 to 12 what will be the effect of changing the EPR to 1, 2 or 3 ?

I'm asking because the sales guy set it to 3 and I think the sample unit I tested on was set to 1. The new machine seems like it's pushing the air back in my lungs when I breathe out but the test machine was much more relaxed.

any help?

thanks


------------------------manual says-------------------------
EPR
On—EPR is enabled.
Off—EPR is disabled.
The following settings are only available if EPR is On:
EPR Type Full Time—If set to Full Time, EPR is enabled for the whole therapy session.
EPR Level 1, 2, 3 cm H 2 O
Ramp Only—If set to Ramp Only, EPR is only enabled during ramp time.
When EPR is enabled, the delivered pressure will not drop below a minimum pressure of 4 cm H 2 O,
regardless of the settings.
Post Reply Post Reply
#2
RE: Expiratory pressure setting for AirSense 10 Autoset
(01-19-2016, 12:10 AM)bazder Wrote: so what I don't understand is the EPR levels of 1, 2, or 3 cm H2O? If my regular pressure range is 5 to 12 what will be the effect of changing the EPR to 1, 2 or 3 ?
At low pressure, some would feel not getting enough air and the fresh air entering the mask is not sufficient to flush out exhaled air and can result in re-breathing own exhaled air

Whether EPR 1, 2, or 3 , exhale pressure does not drop below 4
With auto range 5-12 ... EPR level changes as pressure fluctuate
e.g ... EPR level 2, when pressure at 5, exhale pressure 4 (drop by 1) but as pressure increase to 6, exhale pressure 4 (drop by 2), again as pressure increase further, EPR level becomes the difference between inhale pressure and exhale pressure which is always 2

Personally, I don't use EPR, find it annoying but for some is a godsend
You can change it to whatever level feel most comfortable to you





Post Reply Post Reply
#3
RE: Expiratory pressure setting for AirSense 10 Autoset
thanks for the answer ... strangely enough the pressure seemed less last night so I'll leave it alone for a while - I think I'm still getting used to the whole thing.
Post Reply Post Reply
#4
RE: Expiratory pressure setting for AirSense 10 Autoset
(01-19-2016, 06:15 AM)zonk Wrote: At low pressure, some would feel not getting enough air and the fresh air entering the mask is not sufficient to flush out exhaled air and can result in re-breathing own exhaled air

Whether EPR 1, 2, or 3 , exhale pressure does not drop below 4
With auto range 5-12 ... EPR level changes as pressure fluctuate
e.g ... EPR level 2, when pressure at 5, exhale pressure 4 (drop by 1) but as pressure increase to 6, exhale pressure 4 (drop by 2), again as pressure increase further, EPR level becomes the difference between inhale pressure and exhale pressure which is always 2

Personally, I don't use EPR, find it annoying but for some is a godsend
You can change it to whatever level feel most comfortable to you

Zonk,

I've always wondered why many people seem to be against using the EPR setting, since I always thought it could only make therapy more comfortable by making it easier to exhale. Are you saying that if you don't use it, you can maintain fresher air (more oxygen-rich and CO2-poor) in the mask, by forcing exhaled air to be pushed out of the exhale ports due to the higher pressure at exhale? If so, that might be the answer to my lingering issue with blood oxygen desaturation events..

Thanks,
Possum
Post Reply Post Reply
#5
RE: Expiratory pressure setting for AirSense 10 Autoset
EPR is kind of like using a bilevel machine with a pressure support of 3, except that instead of setting the EPAP minimum pressure, you begin with the CPAP minimum pressure, and EPR is subtracted from that.

With bilevel an EPAPmin is specified that is high enough to prevent airway collapse. Pressure support is added to obtain the IPAP pressure which should resolve hypopnea and RERA. For Auto Bilevel, the machine can increase EPAP and IPAP as needed above the minimum. CPAP with EPR is based on the CPAP pressure settings, but exhale relief is subtracted from that pressure. It is easy to imagine many people having EPAP pressure that allows airway collapse leading to increased OA.

For any CPAP the minimum pressure is 4.0, so EPR does not take effect until pressure is higher than 4.0. As a result, with low pressure settings up to 7.0, the possible IPAP/EPAP pressures are: 4/4, 5/4, 6/4, 7/4, 8/5 etc. Think in terms of bilevel therapy; someone who needs a minimum EPAP pressure of 8.0 needs to set the CPAP min pressure to 11.0.

EPR is great, but I think a lot of people miss how much it potentially reduces therapy pressures, and can allow OA to occur following expiration, before inspiration begins.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#6
RE: Expiratory pressure setting for AirSense 10 Autoset
Hi bazder,
WELCOME! to the forum.!
Much success to you with your CPAP therapy.
trish6hundred
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
Question Suggested Settings Switching AirSense 10 CPAP to AirSense 11 APAP Rickyricardo 2 81 Yesterday, 11:58 PM
Last Post: SarcasticDave94
Question New CPAP User - Help Please - ResMed Airsense 11 Autoset Pawpri 53 1,464 Yesterday, 05:25 PM
Last Post: Pawpri
  Resmed Airsense 10 time to return to low pressure setting jfchap 3 138 04-13-2024, 03:54 PM
Last Post: SarcasticDave94
  Tube Temp and Humidity Setting To Avoid Rainout tomkelly 10 278 04-11-2024, 11:11 PM
Last Post: tomkelly
  Trigger Setting Switch - Aircurve 10 MrIvanDrago 4 155 04-09-2024, 12:35 PM
Last Post: MrIvanDrago
  CPAP to ASV setting translation? Moemah123 2 522 04-03-2024, 10:30 AM
Last Post: Yujiyun
  Need help with Airsense 10 AutoSet JonUK 2 202 03-25-2024, 06:11 PM
Last Post: JonUK


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.