Ok, i took 6 screen shots from last nights sleep data, i saved them in word, here are the first 2


Machine Settings

ModeCPAP Mode APAP
Pressure MinMinimum Therapy Pressure 6.00 cmH2O
Pressure MaxMaximum Therapy Pressure 20.00 cmH2O
Antibacterial FilterAntibacterial Filter No
Climate ControlClimate Control Auto
EPRResMed Exhale Pressure Relief Off
Humidifier StatusHumidifier Enabled Status On
Humidity LevelHumidity Level 3
MaskResMed Mask Setting Full Face
Patient AccessPatient Access No
RampRamp Enable Off
Smart StartMachine auto starts by breathing On
TemperatureClimateLine Temperature 27 ºC
Temperature EnableClimateLine Temperature Enable 1

Post Reply Post Reply
#12
RE: Waking up/Air pressure
I'm no sure if what I just sent is of any help, it is incomplete and I have reached my max upload quota.
Post Reply Post Reply
#13
RE: Waking up/Air pressure
there is a ton of data in pie charts, graphs and percentages that i do not understand. i am not sure how i could get someone to view and decipher what is going on during my sleeps and direct me.
Post Reply Post Reply
#14
RE: Waking up/Air pressure
The reason we use links is because of the upload quota, hence steps 6 and 7

6) upload this to Imgur, or put it in a Google Drive drawing
7) copy the link to a post on the forum
                                                                                                                          
Note: I'm an epidemiologist, not a medical provider. 
Post Reply Post Reply
#15
RE: Waking up/Air pressure
What I can see your pressure rising but no reason here, possibly snoring or flow limitation.
Then rapidly rising after a period of large leaks, a Hyponea, and a CA event to about 15/16 then coming down
Later preceding your wakeup event your pressure rose again, this time in response to a series of events strating with a hyponea, 3 CAs and 2 UAs, and possibily snores and flow limits.

I would like to see the snore and flow limit graphs, but based on this i would consider setting the range to 7 - 15.
15 will stop the rise above that of which you have very little.
7 because I think you could use a little more support on the low end.
Post Reply Post Reply
#16
RE: Waking up/Air pressure
Here is some useful info:
Taking screenshots in SleepyHead can be achieved by pressing the F12 key. The screenshot will be stored in the SleepyHeadData/Profiles/YourName/Screenshot directory. The image file will have a ".png" extension.

Beginner's Guide To SleepyHead:
http://www.apneaboard.com/wiki/index.php...SleepyHead
Note: See Item #14 for more info on posting images.

Ordering your charts for posting:
http://www.apneaboard.com/wiki/index.php...ganization

Using Imgur to post charts:
http://www.apneaboard.com/wiki/index.php...pnea_Board


Crimson Nape
Apnea Board Moderator
www.ApneaBoard.com
___________________________________
Useful Links -or- When All Else Fails:
The Guide to Understanding OSCAR
OSCAR Chart Organization
Attaching Images and Files on Apnea Board
Apnea Helpful Tips

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#17
RE: Waking up/Air pressure
Lets see if these work. There are 2 charts per file..

http://imgur.com/a/i3iRk
http://imgur.com/a/qGqxL

Thanks,
Ed

p.s.- when I look at the other nights sleeps that I uploaded to sleepyhead, no two nights are the same.
Post Reply Post Reply
#18
RE: Waking up/Air pressure
Ed, your latest graphs are working pretty well. Decent results with a few hypopnea, but not a problem. How about we add 2.0 to your minimum pressure from 6.0 to 8.0, and also put in 2.0 for EPR. What that is going to do is give some pressure difference between inhale and exhale. It won't really change your pressures, but sometimes pressure support is good for hypopnea resolution. It might be worth a try. Keep it if it's comfortable and doesn't cause any problems; go back to what you have if you don't like it.


Tom
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#19
RE: Waking up/Air pressure
eddieo,
The minimum pressure of 6 is too low, and your pressure range of 6-13 is too wide.

The pressure will rise when responding to snores. Snores indicate the minimum pressure needs adjusted. Most likely, you are waking up due to the machine ramping up due to snores and also the few leaks you have. It could be your mask moving or mouth breathing. Leaks are not too bad, but try to be aware of that..

I would adjust the pressure to a minimum of 9cm and maximum of 14cm.
You can also try EPR. This can help when exhaling.

And it would be rare to see any two nights the same. AHI numbers can vary and different factors play into that.
OpalRose
Apnea Board Administrator
www.apneaboard.com

_______________________
OSCAR Chart Organization
How to Attach Images and Files.
OSCAR - The Guide
Soft Cervical Collar
Optimizing therapy
OSCAR supported machines
Mask Primer



INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE.  ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA.  INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#20
RE: Waking up/Air pressure
Looking good,
This chart shows your base support level is 8 so set you min to 8 as suggested above.
Did you wake up again? If so why do you think that happened?
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  SleepHQ O2 Ring Alarm waking me up MurrayApnea 1 184 04-19-2024, 08:53 AM
Last Post: G. Szabo
  Flow Limit and frequent waking issues Premier1991 3 227 04-15-2024, 08:07 AM
Last Post: Sleeprider
Gross Waking up anxious Gailofiowa 3 128 04-09-2024, 08:10 AM
Last Post: HalfAsleep
Angry Waking Up with CPAP, Not Getting Full Nights Rest. What Settings to Change? xgrea40 3 355 04-01-2024, 11:40 AM
Last Post: jcp519
Sad [Treatment] 10 Months In and Still Waking Up After 4 to 5 Hours Burnsy 3 407 03-14-2024, 11:37 AM
Last Post: Burnsy
  [CPAP] Persistant night waking - Oscar results help please philjones 0 213 03-07-2024, 05:39 PM
Last Post: philjones
Sad [Symptoms] Traumatic life event. Now I'm waking up every night suffocating. BostonEngineer 9 466 03-03-2024, 12:44 PM
Last Post: HalfAsleep


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.