Data format/capabilities analysis

Grade: F

Negatives

  • Manufacturer failed to openly document this format.
  • No Flow Waveform, unless hacked out through serial port. (awarded an instant F)
  • Very Low resolution event/graph data..
  • Events wound not be able to sync even with flow waveform data.
  • Only the last two settings changes are kept.. :(
  • Needs a separate data module for SD logging.

Positives

  • Non-complicated data structures
  • Machine is firmware upgradable, so there is hope for this machine.

Intellipap Data Format

Intellipap support is still very young, but functional.

I've still got to finish updating this..

Most of this is crud.. Check the SleepyHead source code for more accurate information.

"U" Files

This is the session Index.

It contains repeating 9 byte codes, containing two 32bit (big endian) timestamps, and a single checksum byte.

Timestamps are in seconds since midnight 1 January 2002.

Points to the start and stop time of each session.

A few lines of hex data

12,07,CB,EF,12,08,2F,98,A3
12,09,27,15,12,09,34,14,D8
12,09,35,0C,12,09,35,C4,F1
12,09,35,C6,12,09,90,CF,5F
12,0A,08,4F,12,0A,08,50,86
12,0A,08,51,12,0A,09,76,62
12,0A,09,78,12,0A,09,95,55
12,0A,09,98,12,0A,09,BA,ED

"L" Files

This contains the session data.

Repeating 26 byte codes.

Start with a 32bit time-stamp, a sequence of graph/etc data, and end with a 8bit checksum.

Code          Details               Notes
0x00-0x03     32bit TimeStamp       Big Endian, seconds since 1st January, 2002
0x04          ??? Snore
0x05          ??? Snore             Seems the more likely of the two.
0x06          Leak Average
0x07          Leak Max
0x08-0x09     16bit Tidal Volume    Big Endian
0x0a          Respiratory Rate
0x0b          ??? Unknown           Always 0x0a. Could be Standard Deviation?
0x0c          ??? Unknown graph     (Mapped to Ti) in SleepyHead
0x0d          Pressure
0x0e          ??? Unknown           Always seems to be 0x90
0x0f          ??? Unknown
0x10          >0 Apnea
0x11          >0 Hypopnea
0x12          >0 NRI
0x13          Low Pressure
0x14          Hi Pressure
0x15-0x18     ??? Unused
0x19          Checksum

A few lines of hex data

12,07,CB,EF,00,00,43,20,04,23,10,0A,0E,59,90,00,00,00,00,5A,78,00,00,00,00,80
12,07,CD,57,00,00,39,21,02,BC,10,0A,14,5B,90,00,00,02,01,5A,78,00,00,00,00,EA
12,07,CE,BF,00,00,37,22,02,5A,13,0A,13,63,90,00,00,02,00,5A,78,00,00,00,00,5C
12,07,D0,27,00,00,34,22,02,55,13,0A,14,6A,90,00,00,01,00,5A,78,00,00,00,00,77
12,07,D1,8F,00,01,3D,23,02,9B,13,0A,12,6B,90,00,00,02,00,5A,78,00,00,00,00,CA
12,07,D2,F7,00,01,4D,25,03,53,10,0A,0D,6E,90,00,00,00,00,5A,78,00,00,00,00,4F
12,07,D4,5F,00,00,3D,25,02,70,11,0A,0D,69,90,00,00,01,00,5A,78,00,00,00,00,BA
12,07,D5,C7,00,00,36,25,02,1F,15,0A,0F,68,90,00,00,00,00,5A,78,00,00,00,00,39