Support Apnea Board & OSCAR  

Difference between revisions of "REMstar error codes"

From Apnea Board Wiki
Jump to: navigation, search

(See Also)
(E-61 - E-80)
Line 161: Line 161:
 
|}
 
|}
 
==E-61 - E-80==
 
==E-61 - E-80==
 +
{| class="wikitable"
 +
|-
 +
! scope="col" width="75" align="center" | Error Code !! scope="col" width="325" | Description !! scope="col" width="225" | Probable Cause !! scope="col" width="225" | Corrective Action
 +
|-
 +
| E-61 || ERR_UNUSED_61 || Reserved || Reserved
 +
|-
 +
| E-62 || ERR_UNUSED_62 || Reserved || Reserved
 +
|-
 +
| E-63 || ERR_WATCHDOG_FAILURE || Defective microprocessor || Replace Main PCA
 +
|-
 +
| E-64 || ERR_UNEXPECTED_WATCHDOG || Defective FLASH chip<br />Defective RAM chip || Replace Main PCA<br />Replace Main PCA
 +
|-
 +
| E-65 || ERR_UNUSED_65 || Reserved || Reserved
 +
|-
 +
| E-66 || ERR_RTOS_INIT_FAILURE || Defective PCA || Replace Main PCA
 +
|-
 +
| E-67 || ERR_SERIAL_QUEUE_FULL || Defective PCA || Replace Main PCA
 +
|-
 +
| E-68 || ERR_SNORE_SAMPLING_THREAD_LOCKED || Defective PCA || Replace Main PCA
 +
|-
 +
| E-69 || ERR_MAIN_EXECUTION_THREAD_LOCKED || Defective PCA || Replace Main PCA
 +
|-
 +
| E-70 || ERR_INTERNAL_RAM_FAILURE || Defective microprocessor || Replace Main PCA
 +
|}
 +
 +
{| class="wikitable"
 +
|-
 +
! scope="col" width="75" align="center" | Error Code !! scope="col" width="325" | Description !! scope="col" width="225" | Probable Cause !! scope="col" width="225" | Corrective Action
 +
|-
 +
| E-71 || ERR_EVENT_LOG_QUEUE_FULL || Defective PCA || Replace Main PCA
 +
|-
 +
| E-72 || ERR_TREND_DIV_BY_ZERO || Defective RAM chip || Replace Main PCA
 +
|-
 +
| E-73 || ERR_ROLLOVER_CASE || Defective RAM chip || Replace Main PCA
 +
|-
 +
| E-74 || ERR_WRONG_DISPLAY_TYPE || Defective PCA || Replace Main PCA
 +
|-
 +
| E-75 || ERR_ALARM_CAPACITOR_FAILURE || Defective PCA || Replace Main PCA
 +
|-
 +
| E-76 || ERR_OVER_PRESSURE || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
 +
|-
 +
| E-77 || ERR_UNDER_PRESSURE || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
 +
|-
 +
| E-78 || ERR_OVER_PRESSURE_REG || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
 +
|-
 +
| E-79 || ERR_UNDER_PRESSURE_REG || Occluded or restricted tube<br />Defective PCA || Replace Pressure tubing<br />Replace Main PCA
 +
|-
 +
| E-80 || ERR_ROTOR_LOCKED || Locked Motor<br />Defective PCA || Replace Motor<br />Replace PCA
 +
|}
  
 
==E-80 - E-86==
 
==E-80 - E-86==

Revision as of 20:52, 14 April 2012

Cleanup.png


Here are the Error Codes for REMstar Machines:

E-0 - E-20

Error Code Description Probable Cause Corrective Action
E-0 NO_ERROR Reserved N/A
E-1 ERR_SOFTWARE Reserved Replace Main PCA
E-2 ERR_PROGRAM_CRC_FAILURE Defective FLASH chip Replace Main PCA
E-3 ERR_EXTERNAL_RAM_FAILURE Defective RAM chip Replace Main PCA
E-4 ERR_FIQ_STACK_OVERFLOW Defective RAM chip Replace Main PCA
E-5 ERR_FIQ_STACK_UNDERFLOW Defective RAM chip Replace Main PCA
E-6 ERR_NEST_STACK_OVERFLOW Defective RAM chip Replace Main PCA
E-7 ERR_NEST_STACK_UNDERFLOW Defective RAM chip Replace Main PCA
E-8 ERR_IRQ_STACK_OVERFLOW Defective RAM chip Replace Main PCA
E-9 ERR_IRQ_STACK_UNDERFLOW Defective RAM chip Replace Main PCA
E-10 ERR_TMR_STACK_OVERFLOW Defective RAM chip Replace Main PCA


Error Code Description Probable Cause Corrective Action
E-11 ERR_TMR_STACK_UNDERFLOW Defective RAM chip Replace Main PCA
E-12 ERR_SVC_STACK_OVERFLOW Defective RAM chip Replace Main PCA
E-13 ERR_SVC_STACK_UNDERFLOW Defective RAM chip Replace Main PCA
E-14 ERR_THREAD_STACK_OVERFLOW Defective RAM chip Replace Main PCA
E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION Defective FLASH chip Replace Main PCA
E-16 ERR_ARM_SWI_EXCEPTION Defective FLASH chip Replace Main PCA
E-17 ERR_ARM_PREFETCH_EXCEPTION Defective FLASH chip Replace Main PCA
E-18 ERR_ARM_ABORT_EXCEPTION Defective RAM chip Replace Main PCA
E-19 ERR_ARM_RESERVED_EXCEPTION Defective FLASH chip
Defective microprocessor
Replace Main PCA
E-20 ERR_AIC_DEFAULT_HANDLER Defective FLASH chip Replace Main PCA

E-21 - E-40

Error Code Description Probable Cause Corrective Action
E-21 ERR_AIC_SPURIOUS_HANDLER Defective FLASH or RAM chip Replace Main PCA
E-22 ERR_CAL_DATA_CRC_FAILURE Improper calibration
Defective FLASH chip
Recalibrate
Replace Main PCA
E-23 ERR_INVALID_DRIFT_INDEX_AT_CAL Improper calibration
Occluded or restricted pressure tube
Defective flow sensor
Recalibrate
Replace pressure tube
Replace Main PCA
E-24 ERR_UNRECOGNIZED_CAL_VERSION_BY_SW Improper calibration
Defective FLASH chip
Recalibrate
Replace Main PCA
E-25 ERR_DRIFT_FLOW_TOO_HIGH Occluded or restricted pressure tube
Defective flow sensor
Replace Flow Tube
Replace Main PCA
E-26 ERR_UI_DISPLAY_FAILURE Defective LCD driver chip Replace Main PCA
E-27 ERR_DRIFT_PRESSURE_TOO_HIGH Occluded or restricted pressure tube
Defective pressure sensors
Replace pressure tubing
Replace Main PCA
E-28 ERR_CAL_DATA_BLANK Improper calibration
Defective FLASH chip
Recalibrate
Replace Main PCA
E-29 ERR_DRIFT_SECONDARY_PRESSURE_TOO_HIGH Occluded or restricted pressure tube
Defective pressure sensors
Replace pressure tubing
Replace Main PCA
E-30 ERR_INVALID_NVDATA_SCHEMA Normal for first time power up
Defective EEPROM
Remove & restore power
Replace Main PCA
Error Code Description Probable Cause Corrective Action
E-31 ERR_INVALID_NVDATA_LOCK Defective EEPROM Replace Main PCA
E-32 ERR_INVALID_NVDATA_CHECKSUM Normal for first time power up
Defective EEPROM
Remove & restore power
Replace Main PCA
E-33 ERR_NVDATA_STORAGE_ERROR Defective EEPROM Replace Main PCA
E-34 ERR_NVDATA_QUEUE_FULL Defective EEPROM Replace Main PCA
E-35 ERR_NVDATA_READBACK_FAILURE Defective EEPROM Replace Main PCA
E-36 ERR_NVDATA_RANGE_FAILURE Defective EEPROM Replace Main PCA
E-37 ERR_CORRUPT_RTC_VALUE Defective microprocessor Replace Main PCA
E-38 ERR_RTC_NOT_TICKING Defective microprocessor
Defective crystal
Replace Main PCA
Replace Main PCA
E-39 ERR_UI_QUEUE_FULL Defective PCA Replace Main PCA
E-40 ERR_INVALID_BIST_TEST_CALL Invalid RASP command Reissue RASP command

E-41 - E-60

Error Code Description Probable Cause Corrective Action
E-41 ERR_BIST_THREAD_LOCKED Defective PCA Replace Main PCA
E-42 ERR_ALARM_VOLTAGE_FAILURE Defective power supply Replace Main PCA
E-43 ERR_12VREF_FAILURE Defective power supply Replace Main PCA
E-44 ERR_5VREF_FAILURE Defective power supply Replace Main PCA
E-45 ERR_VBULK_FAILURE Defective power supply Replace Main PCA
E-46 ERR_NEG15VREF_FAILURE Defective power supply Replace Main PCA
E-47 ERR_FL12_SENSOR_RAILED Occluded or restricted pressure tube
Defective flow sensor
Replace flow tube
Replace Main PCA
E-48 ERR_PV2_SENSOR_RAILED Occluded or restricted pressure tube
Defective pressure sensor
Replace pressure tube
Replace Main PCA
E-49 ERR_FLASH_FAILURE Defective FLASH chip Replace Main PCA
E-50 ERR_UNRECOGNIZED_BOARD_REV Needs new software
Defective PCA
Replace Main PCA
Replace Main PCA
Error Code Description Probable Cause Corrective Action
E-51 ERR_MOTOR_FAILURE Defective motor drive chip Replace Main PCA
E-52 ERR_BLOWER_RPM_FAILED_TOLERANCE Stalled or locked blower Replace Blower
Replace Main PCA
E-53 ERR_BLOWER_OFF_CURRENT_HIGH Stalled or locked motor
Defective motor drive chip
Replace Blower
Replace Main PCA
E-54 ERR_BLOWER_ON_CURRENT_HIGH Stalled or locked motor
Defective motor drive chip
Replace Blower
Replace Main PCA
E-55 ERR_PRINTF_BUFFER_OVERFLOW Defective PCA Replace Main PCA
E-56 ERR_THERAPY_EVENT_QUEUE_FULL Defective PCA Replace Main PCA
E-57 ERR_KEYPAD_STUCK_KEY Verify all keys protrude through top
Defective Keypad
Defective pad on PCA
Align keys with top
Replace Keypad
Replace Main PCA
E-58 ERR_UNUSED_58 Reserved Reserved
E-59 ERR_UNUSED_59 Reserved Reserved
E-60 ERR_UNUSED_59 Reserved Reserved

E-61 - E-80

Error Code Description Probable Cause Corrective Action
E-61 ERR_UNUSED_61 Reserved Reserved
E-62 ERR_UNUSED_62 Reserved Reserved
E-63 ERR_WATCHDOG_FAILURE Defective microprocessor Replace Main PCA
E-64 ERR_UNEXPECTED_WATCHDOG Defective FLASH chip
Defective RAM chip
Replace Main PCA
Replace Main PCA
E-65 ERR_UNUSED_65 Reserved Reserved
E-66 ERR_RTOS_INIT_FAILURE Defective PCA Replace Main PCA
E-67 ERR_SERIAL_QUEUE_FULL Defective PCA Replace Main PCA
E-68 ERR_SNORE_SAMPLING_THREAD_LOCKED Defective PCA Replace Main PCA
E-69 ERR_MAIN_EXECUTION_THREAD_LOCKED Defective PCA Replace Main PCA
E-70 ERR_INTERNAL_RAM_FAILURE Defective microprocessor Replace Main PCA
Error Code Description Probable Cause Corrective Action
E-71 ERR_EVENT_LOG_QUEUE_FULL Defective PCA Replace Main PCA
E-72 ERR_TREND_DIV_BY_ZERO Defective RAM chip Replace Main PCA
E-73 ERR_ROLLOVER_CASE Defective RAM chip Replace Main PCA
E-74 ERR_WRONG_DISPLAY_TYPE Defective PCA Replace Main PCA
E-75 ERR_ALARM_CAPACITOR_FAILURE Defective PCA Replace Main PCA
E-76 ERR_OVER_PRESSURE Occluded or restricted tube
Defective PCA
Replace Pressure tubing
Replace Main PCA
E-77 ERR_UNDER_PRESSURE Occluded or restricted tube
Defective PCA
Replace Pressure tubing
Replace Main PCA
E-78 ERR_OVER_PRESSURE_REG Occluded or restricted tube
Defective PCA
Replace Pressure tubing
Replace Main PCA
E-79 ERR_UNDER_PRESSURE_REG Occluded or restricted tube
Defective PCA
Replace Pressure tubing
Replace Main PCA
E-80 ERR_ROTOR_LOCKED Locked Motor
Defective PCA
Replace Motor
Replace PCA

E-80 - E-86

See Also

Smart card error codes




Donate to Apnea Board