For those Respironics' xPAP machines that are data capable and use a 'smart card' to store the data, the machine will display error messages if something goes wrong. Below are the error codes pertaining to the smart cards.
Contents
C-1 - C-10
Error Code |
Description |
Corrective Action
|
C-1 |
SMARTCARD_12C_WRITE_FAIL |
Card inserted upside-down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
|
C-2 |
SMARTCARD_BAD_CARD_TYPE |
Card is not supported by this device. Otherwise, card is corrupt and should be reprogrammed.
|
C-3 |
SMARTCARD_BAD_CIB_VERSION |
Card is corrupt and should be reprogrammed.
|
C-4 |
SMARTCARD_SCRIPT_DURING_PROVIDER_MODE |
Remove card. Exit Provider Mode screens. Then reinsert card.
|
C-5 |
SMARTCARD_PRESENT_IN_CAL_MODE |
Remove card. Exit Calibration Mode. Then reinsert card.
|
C-6 |
SMARTCARD_BAD_CIB_CHECKSUM |
Card is corrupt and should be reprogrammed.
|
C-7 |
SMARTCARD_NOT_PRESENT |
|
C-8 |
SMARTCARD_12C_READ_FAIL |
Card inserted upside down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
|
C-9 |
SMARTCARD_WRITE_CONFIRM_FAIL |
Card is corrupt and should be reprogrammed. Otherwise, replace card.
|
C-10 |
SMARTCARD_BAD_RI_BRAND |
Card is corrupt and should be reprogrammed.
|
C-100 - C-109
Error Code |
Description |
Corrective Action
|
C-100 |
SMARTCARD_BAD_CPB_CHECKSUM |
Card is corrupt and should be reprogrammed.
|
C-101 |
SMARTCARD_BAD_CPB_VERSION |
Card is corrupt and should be reprogrammed.
|
C-102 |
SMARTCARD_INVALID_UNIT |
|
C-103 |
SMARTCARD_UNKNOWN_CPB_FORMAT |
Card contains a prescription format that is not supported. Perhaps card is intended for another device.
|
C-104 |
SMARTCARD_UNKNOWN_CPB_PRESCRIPTION |
Card contains a prescription type that is not supported. Perhaps card is intended for another device.
|
C-105 |
SMARTCARD_SCRIPT_PARM_RANGE |
Card contains prescription settings that are outside of this device’s range. Perhaps card is intended for another device.
|
C-106 |
SMARTCARD_BAD_CPB_SCRIPT_READY |
Card is corrupt and should be reprogrammed.
|
C-107 |
SMARTCARD_BAD_CPB_START_ADDR |
Card is corrupt and should be reprogrammed.
|
C-108 |
SMARTCARD_BAD_CPB_END_ADDR |
Card is corrupt and should be reprogrammed.
|
C-109 |
SMARTCARD_BAD_CPB_ERASE_PERMISSION |
Card is corrupt and should be reprogrammed.
|
C-200 - C-207
Error Code |
Description |
Corrective Action
|
C-200 |
SMARTCARD_BAD_CDCB_VERSION |
Card is corrupt and should be reprogrammed.
|
C-201 |
SMARTCARD_BAD_CDB_VERSION |
Card contains a storage format that is not supported. Perhaps card is intended for another device.
|
C-202 |
SMARTCARD_BAD_CDCB_CHECKSUM |
Card is corrupt and should be reprogrammed.
|
C-203 |
SMARTCARD_BAD_CDCB_CONTROL_BLOCKS |
Card is corrupt and should be reprogrammed.
|
C-204 |
SMARTCARD_BAD_CDCB_START_ADDR |
Card is corrupt and should be reprogrammed.
|
C-205 |
SMARTCARD_BAD_CDCB_END_ADDR |
Card is corrupt and should be reprogrammed.
|
C-206 |
SMARTCARD_BAD_CDB_START_ADDR |
Card is corrupt and should be reprogrammed.
|
C-207 |
SMARTCARD_BAD_CDB_END_ADDR |
Card is corrupt and should be reprogrammed.
|
C-300 - C-303
Error Code |
Description |
Corrective Action
|
C-300 |
SMARTCARD_BAD_UIDB_VERSION |
Card is corrupt and should be reprogrammed.
|
C-301 |
SMARTCARD_BAD_UIDB_CHECKSUM |
Card is corrupt and should be reprogrammed.
|
C-302 |
SMARTCARD_BAD_UIDB_START_ADDR |
Card is corrupt and should be reprogrammed.
|
C-303 |
SMARTCARD_BAD_UIDB_END_ADDR |
Card is corrupt and should be reprogrammed.
|
C-400 - C-404
Error Code |
Description |
Corrective Action
|
C-400 |
SMARTCARD_BAD_DDCB_VERSION |
Card is corrupt and should be reprogrammed.
|
C-401 |
SMARTCARD_BAD_DDCB_CHECKSUM |
Card is corrupt and should be reprogrammed.
|
C-402 |
SMARTCARD_BAD_DDCB_PROFILE_CONTROL_BLOCKS |
Card is corrupt and should be reprogrammed.
|
C-403 |
SMARTCARD_BAD_DDCB_HISTOGRAM_CONTROL_BLOCKS |
Card is corrupt and should be reprogrammed.
|
C-404 |
SMARTCARD_BAD_DDCB_FOSQ_COMPL_CONTROL_BLOCKS |
Card is corrupt and should be reprogrammed.
|
C-500 - C-506
Error Code |
Description |
Corrective Action
|
C-500 |
SMARTCARD_FILE_OPEN_FAIL |
Card is corrupt and should be reprogrammed.
|
C-501 |
SMARTCARD_FILE_TELL_FAIL |
Card is corrupt and should be reprogrammed.
|
C-502 |
SMARTCARD_FILE_SEEK_FAIL |
Card is corrupt and should be reprogrammed.
|
C-503 |
SMARTCARD_FILE_READ_FAIL |
Card is corrupt and should be reprogrammed.
|
C-504 |
SMARTCARD_FILE_WRITE_FAIL |
Card is corrupt and should be reprogrammed.
|
C-505 |
SMARTCARD_FILE_GET_PTR_FAIL |
Card is corrupt and should be reprogrammed.
|
C-506 |
SMARTCARD_FILE_SET_PTR_FAIL |
Card is corrupt and should be reprogrammed.
|
See Also
Donate to Apnea Board