Difference between revisions of "S9 Data Format"
m (→S9 Data Format) |
(format change for new banner) |
||
(19 intermediate revisions by one other user not shown) | |||
Line 1: | Line 1: | ||
{{OSCAR}} | {{OSCAR}} | ||
− | ='''ResMed Data Format | + | <br /> |
+ | |||
+ | ='''ResMed Data Format (excluding S8 models)'''= | ||
This page was originally written by Mark Watkins, during the initial build of the SleepyHead program. It needs change in approach, or should be taken as "history". | This page was originally written by Mark Watkins, during the initial build of the SleepyHead program. It needs change in approach, or should be taken as "history". | ||
− | == | + | |
+ | ==Data format analysis== | ||
Grade: C (extra points awarded for being open) | Grade: C (extra points awarded for being open) | ||
'''Negatives''' | '''Negatives''' | ||
− | * Only keeps basic summary data when card is left out. NO graphs. | + | * Only keeps basic summary data when the card is left out. NO graphs. |
* Unnecessarily deletes high-res flow waveform data after 7 days, all other graphs after 30 days. | * Unnecessarily deletes high-res flow waveform data after 7 days, all other graphs after 30 days. | ||
* Summary does not contain settings info for each session, only once per day, according to ResMed split, which is not suitable for night-owls. | * Summary does not contain settings info for each session, only once per day, according to ResMed split, which is not suitable for night-owls. | ||
− | * Evil SDcard "drift" problem, which | + | * Evil SDcard "drift" problem, which causes Events to NOT sync well over flow waveform. Needs a regular card format to work around this. |
* Bloaty data format. | * Bloaty data format. | ||
'''Positives''' | '''Positives''' | ||
− | * Manufacturer chose to use an industry standard data format. | + | * Manufacturer chose to use an industry-standard data format. |
* Reasonable assortment of data available, and at a decent resolution. | * Reasonable assortment of data available, and at a decent resolution. | ||
− | = | + | = Data Format = |
− | ResMed chose an intelligent and sensible approach when designing their machines. | + | ResMed chose an intelligent and sensible approach when designing their machines, they used an industry standard. (Thankyou ResMed for showing common sense!) |
− | + | '''EDF+ Format'''<br /> | |
+ | ResMed uses the European Data Format Plus format. This format is documented [http://www.edfplus.info here]. It is rather verbose, but sensibly designed.. The Event Annotations are a little tricky. | ||
− | == | + | = ResMed SDCard Folder Structure = |
− | + | <nowiki> | |
− | + | [DATALOG] Folder containing all the session data | |
− | + | [SETTINGS} Folder with machine settings | |
+ | Identification.crc Checksum info | ||
+ | Identification.tgt Machine identification data | ||
+ | Journal.dat tbd | ||
+ | STR.crc Checksum info | ||
+ | STR.edf All overall summary information</nowiki> | ||
− | + | = Analysis of folders and files = | |
+ | ==Identification.tgt== | ||
+ | This file holds an extensive number of data and parameters of the machine. These are shown here in the same fixed order as in the file. | ||
− | + | {| class="wikitable" | |
+ | |- | ||
+ | ! Code !! Remark !! Explanation !! Example | ||
+ | |- | ||
+ | | #IMF|| All later models || TBD || #IMF 0001 | ||
+ | |- | ||
+ | | #VIR|| All later models|| TBD || #VIR 0064 | ||
+ | |- | ||
+ | | #RIR|| All later models|| TBD || #RIR 0064 | ||
+ | |- | ||
+ | | #PVR|| All later models|| TBD || #PVR 0064 | ||
+ | |- | ||
+ | | #PVD|| All later models|| TBD || #PVD 000F | ||
+ | |- | ||
+ | | #CID|| All later models|| TBD || #CID CX040-044-018-015-100-100-100 | ||
+ | |- | ||
+ | | #RID|| All later models|| TBD || #RID 0012 | ||
+ | |- | ||
+ | | #VRN|| Only S9 Models || TBD || #VRN 20142455118 | ||
+ | |- | ||
+ | | #VPC|| Only S9 Models|| TBD || #VPC 3688P | ||
+ | |- | ||
+ | | #NID|| Only S9 Models|| TBD || #NID 0001 | ||
+ | |- | ||
+ | | #VID || All Models || TBD || #VID 002C | ||
+ | |- | ||
+ | | #SRN|| All Models|| Serial number || #SRN 22182060837 | ||
+ | |- | ||
+ | | #SID|| All Models || Serial ID?|| #SID SX584-0201 | ||
+ | |- | ||
+ | | #RID|| Only S9 Models || TBD || #RID 0064 | ||
+ | |- | ||
+ | | #PNA || All Models || Part Name|| #PNA Lumis_VPAP_150_ST | ||
+ | |- | ||
+ | | #PCD|| All Models || TBD || #PCD 28328 | ||
+ | |- | ||
+ | | #PCB|| All Models || Printed Circuit Board number? || #PCB (90)R280-776(91)T3(21)89180702 | ||
+ | |- | ||
+ | | #MID|| All Models || TBD || #MID 0028 | ||
+ | |- | ||
+ | | #FGT|| All Models || TBD || #FGT 28_M40_V44 | ||
+ | |- | ||
+ | | #BID|| All Models || TBD || #BID SX585-0100 | ||
+ | |} | ||
− | + | == DATALOG Folder == | |
− | |||
− | |||
− | |||
− | |||
− | |||
− | |||
− | |||
− | |||
− | |||
The main data is stored in 4 EDF files per session, with each file name consisting of a date, time, and type, followed by .edf | The main data is stored in 4 EDF files per session, with each file name consisting of a date, time, and type, followed by .edf | ||
− | There is also a .crc checksum file for each of the above, but they are not needed to parse the data. | + | There is also a .crc (checksum) file for each of the above, but they are not needed to parse the data. |
==== Example ==== | ==== Example ==== | ||
Line 56: | Line 101: | ||
PLD Low resolution graph data</nowiki> | PLD Low resolution graph data</nowiki> | ||
− | = File Type Notes = | + | === File Type Notes === |
'''Note:''' Some foreign language machines use a different set of signal codes in the EDF data. | '''Note:''' Some foreign language machines use a different set of signal codes in the EDF data. | ||
− | == EVE Event Annotation Data == | + | ==== EVE Event Annotation Data ==== |
The S9 series flags the following events: | The S9 series flags the following events: | ||
*Hypopnea | *Hypopnea | ||
Line 67: | Line 112: | ||
*Central Apnea | *Central Apnea | ||
− | == BRP High resolution graph data == | + | ==== BRP High resolution graph data ==== |
<nowiki> | <nowiki> | ||
Code Details Units Notes | Code Details Units Notes | ||
Line 74: | Line 119: | ||
"Resp Event" Respiratory Event ? VPAP</nowiki> | "Resp Event" Respiratory Event ? VPAP</nowiki> | ||
− | == SAD Pulse Oximetry data == | + | ==== SAD Pulse Oximetry data ==== |
<nowiki> | <nowiki> | ||
Code Details Units Notes | Code Details Units Notes | ||
Line 82: | Line 127: | ||
The oximetry device that records this information is hideously expensive, notoriously unreliable, and ResMed should be ashamed for releasing it. | The oximetry device that records this information is hideously expensive, notoriously unreliable, and ResMed should be ashamed for releasing it. | ||
− | == PLD Low resolution graph data == | + | ==== PLD Low resolution graph data ==== |
'''Signal List''' | '''Signal List''' | ||
Line 108: | Line 153: | ||
The two empty signals always appear at the end. not entirely sure of their purpose. | The two empty signals always appear at the end. not entirely sure of their purpose. | ||
− | == STR.edf == | + | == SETTINGS Folder == |
+ | The Settings folder holds a number of log (.log) files and settings (.tgt) files, each with its own .crc (checksum) file. | ||
+ | |||
+ | == STR.edf File == | ||
This file contains a summary database in EDF format. It contains things like AHI indexes, 95% values, etc... | This file contains a summary database in EDF format. It contains things like AHI indexes, 95% values, etc... | ||
Line 140: | Line 188: | ||
It is implemented in SleepLib/loader_plugins/resmed_loader.cpp / .h | It is implemented in SleepLib/loader_plugins/resmed_loader.cpp / .h | ||
+ | |||
+ | _________________________________________________________________ |
Latest revision as of 06:10, 12 December 2022
Afrikaans /
في الصفحة /
български /
中国 /
臺灣 /
čeština/
Dansk /
Deutsch /
Ελληνικά /
Español /
Filipino /
Français /
עברית /
Magyar /
Italiano /
日本語 /
한국어 /
Nederlands /
Norsk /
Polskie /
Português /
Română /
Pусский /
Suomalainen /
Svenska /
ภาษาไทย /
Türkçe |
Contents
ResMed Data Format (excluding S8 models)
This page was originally written by Mark Watkins, during the initial build of the SleepyHead program. It needs change in approach, or should be taken as "history".
Data format analysis
Grade: C (extra points awarded for being open)
Negatives
- Only keeps basic summary data when the card is left out. NO graphs.
- Unnecessarily deletes high-res flow waveform data after 7 days, all other graphs after 30 days.
- Summary does not contain settings info for each session, only once per day, according to ResMed split, which is not suitable for night-owls.
- Evil SDcard "drift" problem, which causes Events to NOT sync well over flow waveform. Needs a regular card format to work around this.
- Bloaty data format.
Positives
- Manufacturer chose to use an industry-standard data format.
- Reasonable assortment of data available, and at a decent resolution.
Data Format
ResMed chose an intelligent and sensible approach when designing their machines, they used an industry standard. (Thankyou ResMed for showing common sense!)
EDF+ Format
ResMed uses the European Data Format Plus format. This format is documented here. It is rather verbose, but sensibly designed.. The Event Annotations are a little tricky.
ResMed SDCard Folder Structure
[DATALOG] Folder containing all the session data [SETTINGS} Folder with machine settings Identification.crc Checksum info Identification.tgt Machine identification data Journal.dat tbd STR.crc Checksum info STR.edf All overall summary information
Analysis of folders and files
Identification.tgt
This file holds an extensive number of data and parameters of the machine. These are shown here in the same fixed order as in the file.
Code | Remark | Explanation | Example |
---|---|---|---|
#IMF | All later models | TBD | #IMF 0001 |
#VIR | All later models | TBD | #VIR 0064 |
#RIR | All later models | TBD | #RIR 0064 |
#PVR | All later models | TBD | #PVR 0064 |
#PVD | All later models | TBD | #PVD 000F |
#CID | All later models | TBD | #CID CX040-044-018-015-100-100-100 |
#RID | All later models | TBD | #RID 0012 |
#VRN | Only S9 Models | TBD | #VRN 20142455118 |
#VPC | Only S9 Models | TBD | #VPC 3688P |
#NID | Only S9 Models | TBD | #NID 0001 |
#VID | All Models | TBD | #VID 002C |
#SRN | All Models | Serial number | #SRN 22182060837 |
#SID | All Models | Serial ID? | #SID SX584-0201 |
#RID | Only S9 Models | TBD | #RID 0064 |
#PNA | All Models | Part Name | #PNA Lumis_VPAP_150_ST |
#PCD | All Models | TBD | #PCD 28328 |
#PCB | All Models | Printed Circuit Board number? | #PCB (90)R280-776(91)T3(21)89180702 |
#MID | All Models | TBD | #MID 0028 |
#FGT | All Models | TBD | #FGT 28_M40_V44 |
#BID | All Models | TBD | #BID SX585-0100 |
DATALOG Folder
The main data is stored in 4 EDF files per session, with each file name consisting of a date, time, and type, followed by .edf
There is also a .crc (checksum) file for each of the above, but they are not needed to parse the data.
Example
20110601_021922_EVE.edf (date)_(time)_(type).edf
The 4 File Types
Type Details EVE Event annotation data BRP High resolution graph data SAD Pulse oximetry data PLD Low resolution graph data
File Type Notes
Note: Some foreign language machines use a different set of signal codes in the EDF data.
EVE Event Annotation Data
The S9 series flags the following events:
- Hypopnea
- Apnea
- Obstructive Apnea
- Central Apnea
BRP High resolution graph data
Code Details Units Notes "Flow" Flow Rate ? "Mask Pres" High-res Mask Pressure cmH2O "Resp Event" Respiratory Event ? VPAP
SAD Pulse Oximetry data
Code Details Units Notes "Pulse" Pulse Rate Beats per minute "SpO2" SpO2 (Oxygen Saturation) % Saturation
The oximetry device that records this information is hideously expensive, notoriously unreliable, and ResMed should be ashamed for releasing it.
PLD Low resolution graph data
Signal List
Code Details Units Notes "Snore Index" Snore Index ?? "Therapy Pres" Therapy Pressure cmH2O "Insp Pres" IPAP / Inspiratory Pressure cmH2O "MV" Minute Ventilation ? "RR" Respiratory Rate Breaths per Minute "Vt" Tidal Volume ? "Leak" Leak Rate L/Second (I convert to L/Min) "FFL Index" Flow Limitation Index ? "Mask Pres" Low-res Mask Pressure cmH2O "Exp Pres" EPAP / Expiratory Pressure cmH2O "I:E" Inspiratory : Expiratory Ratio ? VPAP "Ti" Ti?? ? VPAP "Te" Te?? ? VPAP "TgMV" Target Minute Ventialation ? ASV Machines "" Unknown Empty ? "" Unknown Empty ?
most ? fields are known, I just have to look them up again..?? fields really are unknown (to me).
The two empty signals always appear at the end. not entirely sure of their purpose.
SETTINGS Folder
The Settings folder holds a number of log (.log) files and settings (.tgt) files, each with its own .crc (checksum) file.
STR.edf File
This file contains a summary database in EDF format. It contains things like AHI indexes, 95% values, etc...
There is only one copy, in the root directory of the SD-card, and it is updated at the end of a session.
Most records contain one entry per day.
SleepyHead doesn't need much of it, as it provides it's own indexing..
It does contain some data not available elsewhere, most notably EPR data.
ResMed S9 Storage Gotchas
The S9 continuously erases older data to keep the card from overflowing:
- BRP data older than 7 days.
- PLD data older than 30 days.
- Everything older than 365 days.
Regularly backup copies of your SD card to work around this!
SleepyHeads Implementation
SleepyHeads EDF parser and S9 importer has been written from scratch
It consists of two parts
- A generic EDF parser class named EDFParser
- and the custom S9 importer class named ResMedLoader
It is implemented in SleepLib/loader_plugins/resmed_loader.cpp / .h