My first results
Hopefully I have got this right.I am not expecting too much help here as the info is very limited so far,only used it twice.Had to have a break for a couple of days as we went away and so could not be doing with it in all honesty.So from what I have here,is there anything anyone can point out that I should or should not be concerned with? I did wake up early hours and took the mask off again,this I think was due to me just being uncomfortably hot last night,been using the spare room which is in the middle floor and is thereofre the warmest compared to upstairs,besides I forgot to open the window.I had the humidifier on but am considering not using that tonight,I guess the use of a humidifier is down to personal preference if you suffer with a dry throat? I imagine we all handle air temeratures differently,or will it be a case that when it gets cold it will be needed?



Day 1
[Image: G30BQ2Tl.png]


Day 2
[Image: 8ooRwiSl.png]
Post Reply Post Reply
#2
RE: My first results
Mask leaks need to be fixed.
Download SleepyHead
Organize your Sleepyhead Charts
Posting Charts
Beginner's Guide to SleepyHead
Mask Primer
5
Advisory Members serve as an "Advisory Committee" to help shape Apnea Board's rules & policies.

Membership in the Advisory Members group does not imply medical expertise or qualification for advising Sleep Apnea patients concerning their treatment.



Post Reply Post Reply
#3
RE: My first results
Your results are satisfactory considering you are using wide-open settings with EPR at 3. If you find that pressure relief comfortable, I'd like to encourage you to change your minimum pressure so the EPR remains constant, and you have a faster response to obstructive events. Setting the minimum CPAP pressure at 7.0 with EPR at 3 results in a consistent 7/4 pressure. If pressure needs to rise it continues to have the same EPR of 3. With your current settings, you start at 4/4, and proceed to 5/4, 6/4 and so forth.

It is the exhale pressure that needs to be elevated to prevent OA, and it is the pressure differential between inhale and exhale that helps relieve hypopnea. Changing minimum pressure to 7 should help with both of those events, and should remain very comfortable. The mask leaks are physical movement of the mask, and will probably diminish as you adapt to this. Get comfortable, open the window, consider raising minimum pressure.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#4
RE: My first results
(08-20-2017, 11:28 AM)Sleeprider Wrote: Your results are satisfactory considering you are using wide-open settings with EPR at 3.  If you find that pressure relief comfortable, I'd like to encourage you to change your minimum pressure so the EPR remains constant, and you have a faster response to obstructive events.  Setting the minimum CPAP pressure at 7.0 with EPR at 3 results in a consistent 7/4 pressure.  If pressure needs to rise it continues to have the same EPR of 3.  With your current settings, you start at 4/4, and proceed to 5/4, 6/4 and so forth.  

It is the exhale pressure that needs to be elevated to prevent OA, and it is the pressure differential between inhale and exhale that helps relieve hypopnea.  Changing minimum pressure to 7 should help with both of those events, and should remain very comfortable.  The mask leaks are physical movement of the mask, and will probably diminish as you adapt to this.  Get comfortable, open the window, consider raising minimum pressure.

Thank you much appreciated,Yeh I do realise about the mask leaks and as you say more as likely due to movement.I am also looking at changing out the mask shortly either to another FFM or a nasal one.I will change those settings tonight and see how a I fair out.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Oscar CPAP Optimization Results-Awesome!? SeePak 17 321 2 hours ago
Last Post: PeaceLoveAndPizza
  [Treatment] Help with interpreting OSCAR results 6bez8dF5lf 7 179 Yesterday, 10:13 AM
Last Post: 6bez8dF5lf
  Test results Kmzizzle 18 1,520 04-12-2024, 10:36 AM
Last Post: Crimson Nape
  [Treatment] New to CPAP but keen to learn - Please could you help with results and Q's? JustinWoods 19 1,948 04-10-2024, 10:13 AM
Last Post: Jay51
  Help interpreting my results (and hello!) mluck94 4 144 04-08-2024, 10:18 AM
Last Post: mluck94
  Having fluctuating results 3rd week in. Data attached Jscott321 5 204 04-03-2024, 10:29 PM
Last Post: Jscott321
  Do I not need a CPAP anymore? In-Lab Study Results brehski 6 203 04-03-2024, 12:38 AM
Last Post: Deborah K.


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.