Hang in there, you are doing great. I know how hard it is trying to help somebody else with their PAP therapy, interpreting for another person does complicate things, but we will work through this.
A few questions that will guide my thought process.
Overweight, what BMI are we dealing with?
EPAP At 5.0 cmH2O seems low for stenting a "normal BMI adult" throat open, if her BMI is high it may take a few more cmH2O to stent her throat. as higher BMI usually carry extra demands to stent the throat. My own personal experience.
She has been on PAP therapy for over a year, She has acclimated to the PAP Therapy pressures, her tolerance to PAP therapy has built up, her body may require more pressure for IPAP to be effective..
Rather than run as a BiLevel, take advantage of the auto feature, to provide a limited range of IPAP increase.
You could try Mode: VPAPAuto
Min EPAP: 7.0 cmH2O
Max IPAP: 13.0 cmH2O
PS: 4.0 cmH2O
Cycle: Medium
Ramp: OFF
Smart Start: OFF
Trigger: Medium
Her pressure would start low and never exceed 13 cmH2O. If that is too much for her, you can adjust IPAP to a lesser value.
Or...You can actually clamp the values of Min EPAP & Max IPAP in VPAPauto Mode so tight, you effectively are running the same BiLevel pressures she is on right now. Then incrementally increase Max IPAP by 0.20 cmH2O at a time, to allow her to acclimate to a higher value. The ResMed will only provide the Max IPAP you set,
if needed, otherwise it will remain at the lower values.
Let me know how you would like to approach Machine settings.
Thank you for the Zoomed in view, I can see some breath wave shapes that are distorted, indicative of a Hypopnea Event.
There are some sleep disturbed breathing conditions that are not recorded as "Events" by the ResMed algorithms. My previous request for a zoomed in view, was for a suspected
Non Flagged Event. I want to know not only what the machine is telling us is a notable event, but also what their algorithms do not catch.
So from my previous post:
"
In your OSCAR Report of 10 March could you post a zoomed view centered around the breathing disturbance at ~0215.
It is possible she is having events not scored by her machine, such as UARS."
Your ResMed did not flag it, but it is there.
We can utilize such non flagged events to fine tune her PAP Therapy for optimized performance.
Could you post the requested view from the 10 March OSCAR Report?
As always any questions, concerns or comments, please post back.