Heated tubing: Solo?
#1
Question 
Heated tubing: Solo?
I was looking at the clinical manual for the AirSense 11, which mentions that the control of the heated tubing is separate from the humidifier controls. I've never used a humidifier, but I'm intrigued by the idea of trying the heated tubing to lessen the occurrence of "froze nose" in the winter.

I mean, at the velocity of air movement through the tube, I can't imagine it having that great an impact, but... this will not harm anything, will it? 

-Jeff
Post Reply Post Reply
#2
RE: Heated tubing: Solo?
When using the ClimateLine 11 tubing, it is detected automatically, and the Climate Control gives you an Automatic and a Manual setting. In the Automatic mode, you can control the tube temperature between 16 and 30 C. The machine controls the water temperature to give approximately 80% RH I believe. In the Manual mode, you can select both the tube temperature and the humidity level from off to 1 to 8 But I have no idea what that amounts to in RH levels.

I have always had the ClimateLine 10 and always used it, and always put water in the tank, so I have no experience with a dry machine.
Admin Note:

pholynyk passed away in October 2024
Click HERE to read his Memorial Thread

~ Rest in Peace ~
Post Reply Post Reply
#3
RE: Heated tubing: Solo?
The heated hose is used to make the interior of the hose warmer than ambient so that the moisture in the incoming air doesn't drop out of the airstream and condense on the hose. Think glass of ice water in the summer, condensation.

In used of my ResMed AirCurve 10 ASV, I've run the humidifier in manual mode, between 4-8 and heated hose at 80-86°F. I know the heat from the hose isn't really much warmer, but it felt more comfortable than if the hose weren't heated. I try to keep my sleeping space at 70-73°F at night. Some like living in a deep freeze bedroom and others live in a tropical bedroom. So all those variables means YMMV/your mileage may vary.
Mask Primer

Positional Apnea

Attach OSCAR, etc.

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#4
RE: Heated tubing: Solo?
Thanks. I understand the "rain out" issue, and how the heated tubing assists with that. We tend to keep the winter temps inside no warmer than 65F, and my room is the most-exposed to temperature extremes. Plus, I have the PAP sitting on a box that raises it to ~6" off the floor. So, the air is colder, there.

-Jeff
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  Custom Solo Pillows for my fat columella? bcolgate 3 248 05-15-2025, 06:08 PM
Last Post: Deborah K.
  12mm Non-Heated Tubing? StratCat48 1 272 05-07-2025, 09:17 PM
Last Post: jsmit86
  Climateline tubing causing machine problems Billmt 1 351 03-21-2025, 06:59 PM
Last Post: super7pilot
  [Equipment] ResMed heated humidifier tubing doesn't stay warm Dr.Bob 11 3,479 03-13-2025, 01:02 AM
Last Post: gotaug
  Evora FFM tubing connector leak, normal? GreenAvocado 8 579 03-12-2025, 04:57 PM
Last Post: SarcasticDave94
  new F&P Solo Nasal mask martyedb 3 402 03-07-2025, 01:52 PM
Last Post: martyedb
  F&P Solo tightening during the night acordeon 10 973 01-27-2025, 08:47 PM
Last Post: Deborah K.


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.