RE: Help Interpreting Data
Based on this one example, your therapy looks pretty good. Your settings are at 5.0 to 10.0 with EPR at 3 and a ramp 4.0 on auto, so minimum pressure is too low for the EPR setting. We try to to set minimum pressure equal to the minimum pressure the machine is capable of plus the pressure reduction setting for EPR. For EPR 3, that means a minimum pressure of 7.0 which results in the bilevel pressure of 7.0/4.0 (inhale/exhale). By doing this, the auto CPAP algorithm can increase the minimum exhale pressure (EPAP) when it detects obstruction. In every case where your chart shows an obstructive event, your pressure is less than 7 and the machine takes time to increase pressure to increase EPAP which is how we prevent OA events. The recommended settings for you are minimum pressure 7.0, maximum pressure 10.0 with EPR 3 which gives a range of 7.0/4.0 to 10.0/7.0. That should reduce events. That is not particularly high pressure, but I don't know what your experience is with starting at 7.0 cm pressure. If you tolerate that comfortably, then turn off ramp and use that pressure. If not, then we should discuss further.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-07-2025, 09:44 PM
RE: Help Interpreting Data
jlwalker11a - Your post is related to your therapy. For this reason, I have moved your thread to the Main Forum. Please use this thread for all your therapy related posts. We can always change the title for a better fit. Just notify a moderator or administrator when you are ready for it to be changed.
- Red
Crimson Nape
Apnea Board Moderator Project Manager for OSCAR - Open Source CPAP Analysis Reporter www.ApneaBoard.com ___________________________________ Useful Links -or- When All Else Fails: The Guide to Understanding OSCAR OSCAR Chart Organization Attaching Images and Files on Apnea Board Apnea Helpful Tips INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-08-2025, 06:47 AM
RE: Help Interpreting Data
Ok. You can move this thread. My apologies for posting in incorrect forum.
Thanks for such a complete answer. I’ll use that setting for a few nights then post the new results.
03-08-2025, 09:34 AM
RE: Help Interpreting Data
Should I also leave Ramp Pressure at 4? Or change that to 7 also?
03-08-2025, 09:36 AM
RE: Help Interpreting Data
The use of ramp is for those that cannot tolerate CPAP pressure to start, and need lower pressure to go to sleep. Ideal is to turn off ramp, but if you need it, set to the highest pressure you tolerate.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-11-2025, 05:56 PM
RE: Help Interpreting Data
This is the data from last night.
Should I raise the pressure upper limit from 10 to 11?
03-11-2025, 08:05 PM
RE: Help Interpreting Data
Your event rate is about 1 every 2-hours. Pretty darn good! Current pressure is 7.0 to 10.0, and I think I would raise minimum to 8.0 and maximum to 11.0. Well done!
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
03-14-2025, 09:01 AM
RE: Help Interpreting Data
Thanks for that help!
These are the last 2 nights after changing that to 8-11. The first didn't look great, but last night looked really well I thought. Justin
03-14-2025, 04:35 PM
RE: Help Interpreting Data
The main difference was on the first night you had minor positional cluster of OA and that resolved the second night. Note the events occurred for no reason in pairs. That is usually positional or chin tucking. Not a big deal. With your excellent results, we’re mostly concerned with comfort and sleep quality.
Sleeprider
Apnea Board Moderator www.ApneaBoard.com ____________________________________________ Download OSCAR Software Soft Cervical Collar Optimizing Therapy Organize your OSCAR Charts Attaching Files Mask Primer How To Deal With Equipment Supplier INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT. |
« Next Oldest | Next Newest »
|
Possibly Related Threads... | |||||
Thread | Author | Replies | Views | Last Post | |
Help Interpreting Oscar Data | Kbarnes | 3 | 208 |
03-30-2025, 12:48 PM Last Post: ejbpesca |
|
Recent diagnosis- first APAP night - could do with some help interpreting | fieldssupsup | 2 | 376 |
03-09-2025, 05:31 PM Last Post: fieldssupsup |
|
OSCAR Data - Help Interpreting | learningtosleep | 2 | 279 |
02-28-2025, 03:33 AM Last Post: learningtosleep |
|
[Diagnosis] Could Use Help Interpreting Sleep Study | Penn Dickery | 2 | 247 |
02-26-2025, 07:33 AM Last Post: Penn Dickery |
|
[Treatment] Interpreting OSCAR data | Tired Young Man | 7 | 497 |
02-11-2025, 01:49 PM Last Post: DaveSkvn |
|
Help interpreting Oscar data | DanApnea | 3 | 440 |
02-02-2025, 12:48 AM Last Post: DanApnea |
|
[Diagnosis] Interpreting WatchPAT Results and Next Steps | fuyu | 0 | 218 |
01-30-2025, 02:37 PM Last Post: fuyu |