Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
#1
Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
Like many of you have a Dream Station (Auto-bipap) - and I recently got a Resmed Air Curve 10 Vauto
I would like to duplicate my settings as closely as possible - but there are a few settings that the Dreamstation has that are not on the Resmed
Namely the BiFlex and Flex settings (the flex settings make it easier to exhale by temporarily lowering the pressure on an exhale)

here are my Dreamstation (Auto-Bipap) settings:
Ipap max : 19
Epap Min : 9
Ps min : 3
Ps max : 4
Flex - BiFlex
Flex : 3

Here are the Resmed AirCurve 10 Vauto settings (As I have them currently set)
Mode : Vauto
Max Ipap : 19
Min Epap : 7
Ps : 4

I really like the flex settings that make it easier to exhale on the Dreamstation.
I am trying to duplicate this on the Resmed with the lower Epap setting

could someone make any recommendations if I should change any of the settings 
to more closely approximate what I have on the Dreamstation.

also .. I am not quite sure I understand the role of PS  (Pressure support)
the instructions say to adjust it for patient comfort.
for an easier exhale .. should I raise it or lower the number?


thanks
MRC
Post Reply Post Reply
#2
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
It would help to look at a daily details chart from Oscar to help guide you. In general, a greater pressure support gives the feeling of easier exhale, regardless of the pressure settings.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
Sure .. let me share my Oscar data.
Let me know if you need any additional data.

thanks again for your time in helping me with my settings

MRC


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#4
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
I don’t see the ResMed charts.
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#5
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
I haven't used the Resmed yet, not until I get the settings figured out.
the data attached is for the Dream station.

thanks
MRC
Post Reply Post Reply
#6
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
The ResMed the min + ps is the inhale pressure 
Exhale is the min pressure

Max is the maximum the machine will go to stop events. 

The only other thing is the trigger - set to high it helps with centrals. You can go to extra high if need be.
Apnea (80-100%) 10 seconds, Hypopnea (50-80%) 10 seconds, Flow Limits (0-50%) not timed  Cervical Collar - Dealing w DME - Chart Organizing
Post Reply Post Reply
#7
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
Moving to the Resmed, let's try these settings. Your chart did not have any OA, so we will start with a lower EPAP min, but push the PS to 4.0 to deal with hypopnea. We may reduce this if centrals are persistent as we see in your Dreamstation chart:

Mode: Vauto
EPAP min 8.0
PS 4.0
Max IPAP 15.0
Trigger Sensitivity HIGH
Cycle Sensitivity Medium
Ti Min 0.3
Ti Max 2.8

Give it a go and if I got something wrong, we will pick up the pieces tomorrow.
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#8
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
Looking for any more recommendations or suggestions.
let me explain my 1st night with the resmed.

Last night went fairly well, but i had to make a few changes.
it felt uncomfortable to exhale at a starting pressure of 8 - like it took too much effort
(I am used to using an epap pressure of 9 on the dreamstation (which is really 6 with the biflex flex-3 option turned on)

so I turned the Epap Min down to 7 
and increased the pressure support to 5
with the Max pressure set to 19 just to make sure I wasn't hitting a pressure limit.

breathing felt natural with the other settings.
I am posting the resmed data below.

here is a synopsis of the night
3am - went to bed, a bit anxious about the new machine
3:45am - took an ambien 
4:00am asleep
7:15am bathroom
7:30am asleep 
10:20am awake 
11:00am up and out of bed.

thanks
MRC


Attached Files Thumbnail(s)
   
Post Reply Post Reply
#9
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
What is your trigger sensitivity?
Try setting it to high. That could help with your centrals
Most of your OA is likely caused by chin tucking but not enough here to worry about.
Post Reply Post Reply
#10
RE: Moving from a Dreamstation Bipap to Resmed Aircurve 10 Vauto
the other settings were (left those the same as was suggested in the earlier post)

Trigger Sensitivity HIGH
Cycle Sensitivity Medium
Ti Min 0.3
Ti Max 2.8

thanks
MRC
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  [Equipment] Bag for Aircurve 10 without humidifier? melbell 2 90 Yesterday, 09:09 AM
Last Post: MrIvanDrago
Smile Alan's New ResMed AirCurve 10 Vauto Came Today! Deborah K. 20 964 04-14-2024, 08:15 PM
Last Post: Sleeprider
  Aircurve 10 Vauto ANTIKATE 5 138 04-14-2024, 12:53 PM
Last Post: HalfAsleep
  Rant and AirCurve 10 vs AirCurve 11 question Whacked 17 669 04-14-2024, 05:29 AM
Last Post: stevew168
  Bipap VAuto Settings Help Please jonoc99 5 141 04-12-2024, 04:16 PM
Last Post: Sleeprider
  [Treatment] Valtor Therapy Thread - from Autoset to Vauto Valtor 50 1,329 04-12-2024, 09:24 AM
Last Post: Sleeprider
  Aircurve 10 or 11 replacement Shamus54 8 211 04-11-2024, 09:12 PM
Last Post: SakimaStorm


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.