Can anyone tell me what the difference is between the settings for Min and Max EPAP  vs  Min and Max PS in the Clinician Setting menu on the ResMed AirCurve 10 ASV?  I notice that the Min PS setting only goes to 6, whereas the Min EPAP setting goes to 9.  

What's the difference between the two kinds of settings?
Post Reply Post Reply
#2
RE: ResMed AirCurve 10 ASV
EPAP min and max is the range of pressure settings for the expiratory positive air pressure. The minimum can be set from 4.0 to 15.0. We use EPAP to hold the airway patent against obstruction during exhale. Minimum PS or pressure support on the Resmed can be set from 0 to 15 and the maximum pressure support can be set to the minimum PS plus 5.0. Resmed does not allow minimum and maximum PS to be equal, so if we set PS min to 3.0, PS max must be at least 8.0. EPAP plus pressure support is IPAP. the minimum PS is used for comfort and easier breathing, and the maximum PS is used to help supplement ventilation in the event of hypopnea, periodic breathing or a central apnea. The Resmed ASV can change the amount of pressure support applied between the minimum and maximum setting on a breath by breath basis. Default settings in ASV Auto mode are EPAP min 4.0, EPAP max 15.0, PS min 3.0 and PS max 15.0. This will result in a potential range of 7.0/4.0 (inhale exhale) to 25.0/15.0. The machine generally will not exceed 23 cm in auto ASV mode.

Your profile indicates you have a Philips Respironics 60 Series Auto BiPAP. ASV settings are not relevant to that machine. Can you clarify what you need to know?
Sleeprider
Apnea Board Moderator
www.ApneaBoard.com

____________________________________________
Download OSCAR Software
Soft Cervical Collar
Optimizing Therapy
Organize your OSCAR Charts
Attaching Files
Mask Primer
How To Deal With Equipment Supplier


INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#3
RE: ResMed AirCurve 10 ASV
EPAP is the airflow that keeps your airway open while pressure support is the difference between EPAP and IPAP. PS in the case of an ASV tends to be used to combat the Central component in conjunction with the unique ASV algorithm controlling it.
INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEBSITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  New ResMed AirCurve 10 VAuto - Oscar Chart Raven87 12 229 Yesterday, 08:11 AM
Last Post: Sleeprider
  [Equipment] Bag for Aircurve 10 without humidifier? melbell 3 157 04-20-2024, 02:07 AM
Last Post: melbell
  Help with new AirCurve 10 VAuto Settings aubreyjs 6 180 04-19-2024, 09:40 AM
Last Post: aubreyjs
  Does Aircurve 10 Vauto record events and flow limits during ramp mode? SingleH 3 103 04-19-2024, 09:12 AM
Last Post: SingleH
Smile Alan's New ResMed AirCurve 10 Vauto Came Today! Deborah K. 20 1,018 04-14-2024, 08:15 PM
Last Post: Sleeprider
  Aircurve 10 Vauto ANTIKATE 5 183 04-14-2024, 12:53 PM
Last Post: HalfAsleep
  Rant and AirCurve 10 vs AirCurve 11 question Whacked 17 746 04-14-2024, 05:29 AM
Last Post: stevew168


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.