Hello Guest, Welcome to Apnea Board !
As a guest, you are limited to certain areas of the board and there are some features you can't use.
To post a message, you must create a free account using a valid email address.

or Create an Account


New Posts   Today's Posts

Respironics algorithm
#1
Respironics algorithm
I have been using a Resmed Auto SV for about a year. Last night I had to switch to my backup machine, Respironics System One Bi-pap autoSV. What  noticed right off the bat is the difference between inhaling on Resmed vs the Respironics. On the Resprionics there is a "delay" when  inhale. I inhale and it seems like I am "sucking" and a few msec later the pressure increases and then  I can take a complete breath. Is this normal with the Respironics algorithm? Or is my machine not working right? Thanks for any and all input.
Post Reply Post Reply
#2
RE: Respironics algorithm
Hi jallain,
WELCOME! to the forum.!
Hang in there for answers to your question and good luck to you as you continue CPAP therapy.
trish6hundred
Post Reply Post Reply
#3
RE: Respironics algorithm
jallain,
It's a bit hard to answer your question without more information.

Are you using the exact same pressure setting? Your profile reads 6-15. The minimum pressure may work with a ResMed machine, but you may want to raise it when using a respironics machine, as they are slower to respond.
OpalRose
Apnea Board Administrator
www.apneaboard.com

_______________________
OSCAR Chart Organization
How to Attach Images and Files.
OSCAR - The Guide
Soft Cervical Collar
Optimizing therapy
OSCAR supported machines
Mask Primer



INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE.  ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA.  INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#4
RE: Respironics algorithm
Thank you! I will raise the pressure on the Respironics and see how tonight goes.
Post Reply Post Reply
#5
RE: Respironics algorithm
G'day jallain. Welcome to Apnea Board.

I suspect the problem is not related to pressure per se, but to the way the machine transitions between inhalation and exhalation. From my limited understanding of your machine, I think there should be settings for Bi-flex and rise time. From something I found on-line:

Bi-Flex Comfort Feature
If enabled, the device provides a comfort feature called Bi-Flex. The Bi-Flex attribute adjusts therapy by inserting a
small amount of pressure relief during the latter stages of inspiration and during active exhalation (the beginning part
of exhalation). Bi-Flex levels of 1, 2, or 3 progressively reflect increased pressure relief that will take place at the end
of inspiration and at the beginning of expiration.

Rise Time Comfort Feature
If enabled, the device provides a comfort feature called Rise Time. Rise time is the amount of time it takes the device
to change from the expiratory pressure setting to the inspiratory pressure setting. Rise time levels of 0, 1, 2, or 3
progressively reflect slowed response of the pressure increase that will take place at the beginning of inspiration. A
setting of 0 is the fastest rise time while a setting of 3 is the slowest. Providers should adjust the rise time to find the
most comfortable setting for the patient. Rise time cannot be adjusted when Bi-Flex is enabled.


From the description of the problem you gave above, it sounds like Bi-flex should be off and rise time set to zero. Maybe one of the Respironics experts can confirm (or contradict) this.
Post Reply Post Reply
#6
RE: Respironics algorithm
Thanks for the input. I went into the setup menu because I wanted to change the rise time. My machine does not have a rise time parameter. I will turn on bi-flex and see if that helps. Thanks.
Post Reply Post Reply
#7
RE: Respironics algorithm
transitions from exhale to inhale seem not to be a priority for the PR BiPAP. flex settings center on the transition from inhale to exhale.

[Image: a566waAl.png]
from: Respironics Publication #1063524.

QAL
Dedicated to QALity sleep.
Post Reply Post Reply
#8
RE: Respironics algorithm
What you describe sounds to me like Bi-Flex is the cause. The default setting is 2. Try turning it down or off.

The transition to IPAP occurs when you start to inhale whereas with ResMed it occurs when you stop exhaling if I remember correctly.
Sleepster

INFORMATION ON APNEA BOARD FORUMS OR ON APNEABOARD.COM SHOULD NOT BE CONSIDERED AS MEDICAL ADVICE. ALWAYS SEEK THE ADVICE OF A PHYSICIAN BEFORE SEEKING TREATMENT FOR MEDICAL CONDITIONS, INCLUDING SLEEP APNEA. INFORMATION POSTED ON THE APNEA BOARD WEB SITE AND FORUMS ARE PERSONAL OPINION ONLY AND NOT NECESSARILY A STATEMENT OF FACT.
Post Reply Post Reply
#9
RE: Respironics algorithm
Try the bi-flex at 1 OR if adjusting the Rise Time the Bi-flex has to be off.  Probably why you couldn't adjust it.
From the clinician's manual:

• Flex type - This screen displays the comfort mode setting. You can select None or Bi-Flex.

• Bi-Flex - You can modify the Flex setting (1, 2 or 3) on this screen if you enabled Bi-Flex. The Bi-Flex comfort feature allows you to adjust the level of air pressure relief that the patient feels when exhaling during therapy. The setting of "1" provides a small amount of pressure relief, with higher numbers providing additional relief.

• Rise time - Rise time is the time it takes for the device to change from the expiratory pressure setting to the inspiratory pressure setting. This screen allows you to adjust the rise time so you can find the desired setting. A setting of "0" is the fastest rise time, while "3" is the slowest. This is only available if Flex type is set to None and PS max is greater than zero.
Post Reply Post Reply
#10
RE: Respironics algorithm
Yes, you are exactly correct.  I used the Phillips ASV machine for my sleep test, that was required by my doctor, before I could be approved by Medicare for any ASV machine.  Since my machine at that time was the ResMed AirCurve 10 VAuto, and I was used to it, I quickly noticed the ever so slight delay in the Phillips machine in responding to the inhale breath.  I found it so troublesome that I had  trouble sleeping for any length of time during the sleep test. Whereas the AirCurve was seamless in responding to the inhale breath, the Phillips had that ever so slight delay in it's response time.  It is not anything big, and over time I guess one would get used to it.
Post Reply Post Reply


Possibly Related Threads...
Thread Author Replies Views Last Post
  NOTICE: Philips Respironics Recall Replacement Tally RayBee 184 29,824 04-01-2024, 12:26 PM
Last Post: RayBee
  Complete Newb trying a Respironics Legacy CPAP for the first time. Dmickey68 5 267 02-28-2024, 03:20 PM
Last Post: Sleeprider
  Dreamwear Philips Respironics mask altrout 12 858 02-12-2024, 03:04 PM
Last Post: tbhausen
  Returning defective machines to Philips Respironics big_dave 5 867 02-04-2024, 02:15 PM
Last Post: big_dave
  Philips Respironics DreamStation check power Steve Foley 18 22,719 01-07-2024, 02:09 PM
Last Post: zebedee
  Philips Respironics DreamStation not reaching Max Pressure Xsoldier2000 15 2,397 01-05-2024, 04:08 PM
Last Post: Tom Buck
  Resmed Algorithm changes? BelgianWaffle 1 340 01-05-2024, 08:18 AM
Last Post: Sleepster


New Posts   Today's Posts


About Apnea Board

Apnea Board is an educational web site designed to empower Sleep Apnea patients.